ResMed AirSense 10: Won't let me set the time!
ResMed AirSense 10: Won't let me set the time!
Hi everyone,
I have a ResMed AirSense 10 Autoset. I was trying to sue the sleepyhead software but noticed that the time zones were messed up. In the maintenance menus I was able to find the time set option. The problem is, whenever I try to set the time to the correct time (earlier than the time it is set for) the machine gives me an error "cannot set time earlier than current time"! Oddly I can only change it to a time in the future! Has anyone else run into this? It is maddening!!!
I have a ResMed AirSense 10 Autoset. I was trying to sue the sleepyhead software but noticed that the time zones were messed up. In the maintenance menus I was able to find the time set option. The problem is, whenever I try to set the time to the correct time (earlier than the time it is set for) the machine gives me an error "cannot set time earlier than current time"! Oddly I can only change it to a time in the future! Has anyone else run into this? It is maddening!!!
Re: ResMed AirSense 10: Won't let me set the time!
you can't set the time backwards when there is already data in that days period, which is from noon to noon, machine time.
try setting it back just before you go to bed, maybe you'll have rolled over to a clean 'day' by then.
your other alternatives are to just live with it, or to erase all data, set the time, and then go forward.
try setting it back just before you go to bed, maybe you'll have rolled over to a clean 'day' by then.
your other alternatives are to just live with it, or to erase all data, set the time, and then go forward.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: ResMed AirSense 10: Won't let me set the time!
Thanks! I guess another option is to take the card out for a night and lose that data, but then I'd have a "clean" day? Or is the machine too smart for that?
- Okie bipap
- Posts: 3553
- Joined: Thu Oct 15, 2015 4:14 pm
- Location: Central Oklahoma
Re: ResMed AirSense 10: Won't let me set the time!
The machine is too smart for that. When I change the time on our machines, I make the change around 5:00 or 6:00 in the evening before the machines are started for the evening. The machine records data if the card is in or out. When the card is out, the amount of data that is saved is much more than what is saved on the internal memory.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Evora Full Face Mask - Fitpack |
Additional Comments: IPAP 20-25, ps 4, OSCAR software |
Growing old is mandatory, but growing up is optional.
-
- Posts: 1739
- Joined: Mon Mar 27, 2017 5:45 pm
- Location: Central Virginia
Re: ResMed AirSense 10: Won't let me set the time!
Deleted
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software |
Last edited by prodigyplace on Sat Apr 15, 2017 3:48 pm, edited 1 time in total.
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.
Re: ResMed AirSense 10: Won't let me set the time!
Why would you want to sue sleepyhead? That is maybe the most generous thing I have ever seen.robo2084 wrote:Hi everyone,
I have a ResMed AirSense 10 Autoset. I was trying to sue the sleepyhead software but noticed that the time zones were messed up. In the maintenance menus I was able to find the time set option. The problem is, whenever I try to set the time to the correct time (earlier than the time it is set for) the machine gives me an error "cannot set time earlier than current time"! Oddly I can only change it to a time in the future! Has anyone else run into this? It is maddening!!!
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Eson™ 2 Nasal CPAP Mask with Headgear |
All posts reflect my own opinion based on my experience and reading.
Your mileage may vary
Past performance is no guarantee of future results
Consult with your own physician as people very
Your mileage may vary
Past performance is no guarantee of future results
Consult with your own physician as people very
Re: ResMed AirSense 10: Won't let me set the time!
that is not an option, the machine has internal data storage.robo2084 wrote:Thanks! I guess another option is to take the card out for a night and lose that data, but then I'd have a "clean" day? Or is the machine too smart for that?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: ResMed AirSense 10: Won't let me set the time!
TASmart wrote:Why would you want to sue sleepyhead? That is maybe the most generous thing I have ever seen.robo2084 wrote:Hi everyone,
I have a ResMed AirSense 10 Autoset. I was trying to sue the sleepyhead software but noticed that the time zones were messed up. In the maintenance menus I was able to find the time set option. The problem is, whenever I try to set the time to the correct time (earlier than the time it is set for) the machine gives me an error "cannot set time earlier than current time"! Oddly I can only change it to a time in the future! Has anyone else run into this? It is maddening!!!
I really think he meant "setup" Sleepyhead. Don't you just love spell correction?
Old man, new machine. Better sleep![b]
Re: ResMed AirSense 10: Won't let me set the time!
The machine has no time zone settingsxxyzx wrote:which way do you need to move itrobo2084 wrote:Hi everyone,
I have a ResMed AirSense 10 Autoset. I was trying to sue the sleepyhead software but noticed that the time zones were messed up. In the maintenance menus I was able to find the time set option. The problem is, whenever I try to set the time to the correct time (earlier than the time it is set for) the machine gives me an error "cannot set time earlier than current time"! Oddly I can only change it to a time in the future! Has anyone else run into this? It is maddening!!!
what zone are you in
what zone is the machine set for
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
- Posts: 1739
- Joined: Mon Mar 27, 2017 5:45 pm
- Location: Central Virginia
Re: ResMed AirSense 10: Won't let me set the time!
I was thinking they meant use and their fingers got dyslexicjsielke wrote:TASmart wrote:Why would you want to sue sleepyhead? That is maybe the most generous thing I have ever seen.robo2084 wrote:Hi everyone,
I have a ResMed AirSense 10 Autoset. I was trying to sue the sleepyhead software but noticed that the time zones were messed up. In the maintenance menus I was able to find the time set option. The problem is, whenever I try to set the time to the correct time (earlier than the time it is set for) the machine gives me an error "cannot set time earlier than current time"! Oddly I can only change it to a time in the future! Has anyone else run into this? It is maddening!!!
I really think he meant "setup" Sleepyhead. Don't you just love spell correction?
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software |
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.
-
- Posts: 15
- Joined: Fri Apr 13, 2018 11:58 am
- Location: Maryland
Re: ResMed AirSense 10: Won't let me set the time!
Insert a different SD card that either has old or no data on it and change time. Then reinsert SD card that has all your historical data. No need to erase anything
Re: ResMed AirSense 10: Won't let me set the time!
Has nothing to do with SD card when the machine won't let a person go backwards with the time.wallanbrewer wrote: ↑Wed Dec 19, 2018 7:33 amInsert a different SD card that either has old or no data on it and change time. Then reinsert SD card that has all your historical data. No need to erase anything
It's because of the data stored in the internal memory of the machine..It will still give the same prompt with a new SD card or even no SD card.
It simply won't let you go back in time if there is already data recorded in the machines internal storage memory that is in the time frame for the machine's reporting sessions.
If you did this and it worked...it worked because by the time you tried a new blank SD card you were no longer trying to go back into time where there was already data stored in the machine. Had nothing to do with the SD card.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: ResMed AirSense 10: Won't let me set the time!
This is incorrect, the SD card has nothing to do with the issue. The issue is the compliance data internal to the machine.wallanbrewer wrote: ↑Wed Dec 19, 2018 7:33 amInsert a different SD card that either has old or no data on it and change time. Then reinsert SD card that has all your historical data. No need to erase anything