Sleephead graph settings suggestions

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
ephraimh
Posts: 114
Joined: Thu Jan 03, 2008 9:59 pm
Location: Seattle

Sleephead graph settings suggestions

Post by ephraimh » Mon Feb 20, 2017 10:13 pm

Could I have some input on possible settings changes to my AirCurve 10? Data for 2-8-2017 and 2-17 can be found at the link below

https://ephraimh.imgur.com/

thanks

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: min EPAP 10, max EPAP 14, PS min 6, PS max 12, ramp off
Satan ~~ Spawn of Apria Can't type well ith this stupid macular degeneration.

User avatar
ephraimh
Posts: 114
Joined: Thu Jan 03, 2008 9:59 pm
Location: Seattle

Re: Sleepyhead graph settings suggestions

Post by ephraimh » Thu Feb 23, 2017 12:58 am

Could I have some input on possible settings changes to my AirCurve 10? A corrected graph for 2-17 can be found at the link below

thank you

http://imgur.com/Xb7iYyM

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: min EPAP 10, max EPAP 14, PS min 6, PS max 12, ramp off
Satan ~~ Spawn of Apria Can't type well ith this stupid macular degeneration.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Sleepyhead graph settings suggestions

Post by palerider » Thu Feb 23, 2017 3:12 am

ephraimh wrote:Could I have some input on possible settings changes to my AirCurve 10? A corrected graph for 2-17 can be found at the link below

thank you

http://imgur.com/Xb7iYyM
a bit more epap.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
LSAT
Posts: 13232
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Sleepyhead graph settings suggestions

Post by LSAT » Thu Feb 23, 2017 5:51 am

ephraimh wrote:Could I have some input on possible settings changes to my AirCurve 10? A corrected graph for 2-17 can be found at the link below

thank you

http://imgur.com/Xb7iYyM
Need to control the leaks.....What mask are you using?

User avatar
ephraimh
Posts: 114
Joined: Thu Jan 03, 2008 9:59 pm
Location: Seattle

Re: Sleepyhead graph settings suggestions

Post by ephraimh » Fri Feb 24, 2017 2:32 pm

LSAT wrote:
ephraimh wrote:Could I have some input on possible settings changes to my AirCurve 10? A corrected graph for 2-17 can be found at the link below

thank you

http://imgur.com/Xb7iYyM
Need to control the leaks.....What mask are you using?
The mask is a Resmed Activa - LT large. The cushion and headgear are new and I have cinched the headgear as much as I can.

I am getting a new mask and but need to sample some so I can get a decent fit. My DME is Apria so I'm getting a fit from anopther DME, as well. My ASV S9 leaked a little bit but I was able to keep it under 10 almost every night I tried using an Airfit but I couldn't exhale trough it.

Someone said I needed more EPAP so I increased the EPAP Min on the AirCurve last night from 9 to 10. Perhaps turning it back to 9 and increasing the EPAP Max to 15 is what he meant. Last night's settings:
EPAP Min 10
EPAP Max 14
PS Min 6
PS Max 12

Here are the 2-17-2017 and 2-24-2017 nightly Sleepyhead graphs: https://imgur.com/a/kiBA9

Thanks very much

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: min EPAP 10, max EPAP 14, PS min 6, PS max 12, ramp off
Satan ~~ Spawn of Apria Can't type well ith this stupid macular degeneration.

User avatar
OkyDoky
Posts: 2870
Joined: Mon Aug 25, 2014 5:18 pm

Re: Sleephead graph settings suggestions

Post by OkyDoky » Fri Feb 24, 2017 2:42 pm

Double check your min Epap setting. The screenshot for the 24th shows it to still be 9 not 10. And yes the min was what palerider was suggesting.
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

User avatar
Pugsy
Posts: 64002
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sleephead graph settings suggestions

Post by Pugsy » Fri Feb 24, 2017 4:07 pm

Your leaks aren't horribly horrible unless you are seeing worse leaks than you show with this last report.
While it's not real pretty...there really wasn't a lot of prolonged time spent in large leak and it's not enough to hugely impact your therapy.
Less than 10 % of the night in large leak and unless they are disturbing your sleep and causing you to wake often it's not the end of the world.
ResMed won't even give you a Mr Frowny face for mask fit until you hit 30% or more of the night in large leak.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
bonjour
Posts: 322
Joined: Thu May 14, 2009 5:39 pm

Re: Sleepyhead graph settings suggestions

Post by bonjour » Fri Feb 24, 2017 5:25 pm

ephraimh wrote:
LSAT wrote:
ephraimh wrote: The mask is a Resmed Activa - LT large. The cushion and headgear are new and I have cinched the headgear as much as I can.
My mask for years was the original Aptiva nasal mask. The secret with it was a VERY loose fit and then let the PAP pressure actually seal it.

Try to hold it on lightly without straps, by hand and move it in and out . I haven't tried the LT but if it is anything like the orig that should work

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: VAuto 11-25cmH2O PS 3 --- OSCAR software, Many masks - Amara View FFM to P10 Pillows Several Nasal and FFM

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Sleepyhead graph settings suggestions

Post by palerider » Fri Feb 24, 2017 7:26 pm

ephraimh wrote: Someone said I needed more EPAP so I increased the EPAP Min on the AirCurve last night from 9 to 10. Perhaps turning it back to 9 and increasing the EPAP Max to 15 is what he meant
epap is what works to prevent obstructive apneas, if the epap is too low, your airway can close up and you can't inhale, so you can't trigger IPAP on your own...

maxepap is only good *after* you've had some breathing issues, so bumping up the minimum is what should be done to try and prevent them in the first place. hopefully that will help with the Apneas and hypopneas.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
ephraimh
Posts: 114
Joined: Thu Jan 03, 2008 9:59 pm
Location: Seattle

Re: Sleephead graph settings suggestions

Post by ephraimh » Tue Feb 28, 2017 5:18 pm

As Palerider suggested (thank you!), on Sunday night I upped the Min EPAP from 10 to 11; Max EPAP is 14.

These are Sunday night's results (2-26-2017)

http://imgur.com/VuWQ9xY

Monday night's (2-27-2017)results are here, but something went wrong. The graph shows this is for the 28th and ends on March 1. March 1 isn't here yet.The date on the Aircurve, right now, is 3-1-2017. It won't go reset to 2-28-2017

http://imgur.com/7wXlQJ3

Thank you for bearing with me.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: min EPAP 10, max EPAP 14, PS min 6, PS max 12, ramp off
Satan ~~ Spawn of Apria Can't type well ith this stupid macular degeneration.

User avatar
Pugsy
Posts: 64002
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sleephead graph settings suggestions

Post by Pugsy » Tue Feb 28, 2017 5:32 pm

Somehow or other the calendar date on the AirCurve has got moved forward a day and it won't let you go backwards because you already have data in the machine for the 28th...at least per the machine.

It won't let you go backwards when there is data present during that reporting period.

Unless PR has a better way to fix it the only thing I know to do is erase all the data and then it will let you reset the calendar or wait until a night where you don't use the machine for some reason and then it will let you go backwards on the calendar.

I had that happen to the S9 once and I would have sworn I didn't touch the calendar to move it forward but forward it was.
I am still scratching my head over that one. I never mess with the calendar but it got messed with.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Sleephead graph settings suggestions

Post by palerider » Tue Feb 28, 2017 5:42 pm

Pugsy wrote:Unless PR has a better way to fix it the only thing I know to do is erase all the data and then it will let you reset the calendar
unfortunately, I do not.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.