I'm beginning to despair

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
roadrunnerNM
Posts: 7
Joined: Thu Jan 05, 2017 9:36 am

I'm beginning to despair

Post by roadrunnerNM » Thu Jan 05, 2017 12:09 pm

I've had my machine since March, and though it's helped, I still regularly have nights bad enough that I can't go to work the next day. After my titration study, pressure was set to 11. My doctor reviewed the data and said I was still having obstructions; so the pressure was raised. To make a long story short, we're now up to 16, and I'm still having problems. I have good nights, with AHI of about 5, occasionally, but right now, I've being getting AHI readings of 10-12 for the night for about 3 weeks now. I have absolutely no idea what makes the difference between a good and a bad night. I've tried everything I can think of, including banning the cats from the bedroom (a real sacrifice). My doctor has looked puzzled the last few times I've seen him. Does anybody have any suggestions?

_________________
Mask: Simplus Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamStation BiPAP autoSV

User avatar
OkyDoky
Posts: 2870
Joined: Mon Aug 25, 2014 5:18 pm

Re: I'm beginning to despair

Post by OkyDoky » Thu Jan 05, 2017 12:20 pm

roadrunnerNM wrote:I've had my machine since March, and though it's helped, I still regularly have nights bad enough that I can't go to work the next day. After my titration study, pressure was set to 11. My doctor reviewed the data and said I was still having obstructions; so the pressure was raised. To make a long story short, we're now up to 16, and I'm still having problems. I have good nights, with AHI of about 5, occasionally, but right now, I've being getting AHI readings of 10-12 for the night for about 3 weeks now. I have absolutely no idea what makes the difference between a good and a bad night. I've tried everything I can think of, including banning the cats from the bedroom (a real sacrifice). My doctor has looked puzzled the last few times I've seen him. Does anybody have any suggestions?
We need to start with what equipment you have. You can use this link to help you register it where it will show each time you post. wiki/index.php/Registering_Equipment_in_User_Profile
We need to find out if your machine is data capable and what type of apneas are making up your AHI. Are you having leaks? Do you wake up frequently?
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

User avatar
BlackSpinner
Posts: 9745
Joined: Sat Apr 25, 2009 5:44 pm
Location: Edmonton Alberta
Contact:

Re: I'm beginning to despair

Post by BlackSpinner » Thu Jan 05, 2017 12:23 pm

There is software called sleepyhead see tutorial in above section.

Install it, take screen shots, post them on photobucket and link them here.

Tell us EXACTLY what machine you use and its settings.

Get a copy of the clinical manual and learn to drive your machine. If you were a diabetic you would not be running to your doctor every time you need to adjust your insulin.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: Hybrid Full Face CPAP Mask with Nasal Pillows and Headgear
Additional Comments: Quatro mask for colds & flus S8 elite for back up
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal

User avatar
DreamStalker
Posts: 7509
Joined: Mon Aug 07, 2006 9:58 am
Location: Nowhere & Everywhere At Once

Re: I'm beginning to despair

Post by DreamStalker » Fri Jan 06, 2017 10:56 am

Doctors are pretty good at diagnosis but when it comes to CPAP therapy ... most are about as stupid as they come.

Most failures are due to inability to understand pressure leakages. Increasing pressure without fixing leak problems just leads to more leaks and ineffective therapy.

As others have noted, you need to provide more details re: blower machine, settings, mask type, etc.
President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: I'm beginning to despair

Post by Wulfman... » Fri Jan 06, 2017 2:05 pm

roadrunnerNM wrote:I've had my machine since March, and though it's helped, I still regularly have nights bad enough that I can't go to work the next day. After my titration study, pressure was set to 11. My doctor reviewed the data and said I was still having obstructions; so the pressure was raised. To make a long story short, we're now up to 16, and I'm still having problems. I have good nights, with AHI of about 5, occasionally, but right now, I've being getting AHI readings of 10-12 for the night for about 3 weeks now. I have absolutely no idea what makes the difference between a good and a bad night. I've tried everything I can think of, including banning the cats from the bedroom (a real sacrifice). My doctor has looked puzzled the last few times I've seen him. Does anybody have any suggestions?
WHAT machine?
Please give specific make and model. If it's capable of collecting full data (not just hours used), you need to use software to see what's going on.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Dennis de NorthWet
Posts: 19
Joined: Wed May 09, 2012 1:01 pm
Location: 4th Corner - Pacific NorthWet

Re: I'm beginning to despair

Post by Dennis de NorthWet » Sun Jan 08, 2017 6:29 pm

roadrunnerNM wrote:
"......... I have absolutely no idea what makes the difference between a good and a bad night. I've tried everything I can think of, including banning the cats from the bedroom (a real sacrifice). My doctor has looked puzzled the last few times I've seen him. Does anybody have any suggestions?"

Not directly C-PAP related, but if you want to know what's going on while you're asleep, set up a video system that will show you. Who knows, perhaps the cats are sitting on your chest, or some such. If you are thrashing around, or sleeping in a position that would reduce your breathing ability or air supply, you'll be able to sort that out.
Regards,
D

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Eson™ Nasal CPAP Mask with Headgear
Additional Comments: Started with a S9. Now have AirSense 10. S10 is noisier than S9 even though it's supposed to be the quietest thing on the market.

roadrunnerNM
Posts: 7
Joined: Thu Jan 05, 2017 9:36 am

Re: I'm beginning to despair

Post by roadrunnerNM » Sun Feb 05, 2017 5:06 pm

OK, I got my act together sufficiently to download Sleepyhead and take a look. I have a Respironics Auto-CPAP, and I use a full-face mask.

Image

Not sure what to make of it, but I guess you can see why I'm tired.

_________________
Mask: Simplus Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamStation BiPAP autoSV

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: I'm beginning to despair

Post by palerider » Sun Feb 05, 2017 6:01 pm

the summary is pretty useless.

https://sleep.tnet.com/resources/sleepyhead/shorganize

do it like that, and show us the detail chart.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
bonjour
Posts: 322
Joined: Thu May 14, 2009 5:39 pm

Re: I'm beginning to despair

Post by bonjour » Sun Feb 05, 2017 6:24 pm

getting close,

Still need the charts, I I do see a 90% number of 16 cmH2O so we will need to make changes,

Let us see the charts so we can get a better idea of what is going on.

see the organization portion as that is important

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: VAuto 11-25cmH2O PS 3 --- OSCAR software, Many masks - Amara View FFM to P10 Pillows Several Nasal and FFM

User avatar
Pugsy
Posts: 64027
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: I'm beginning to despair

Post by Pugsy » Sun Feb 05, 2017 6:37 pm

Please be aware what a 90% number represents....in this case it's a pressure number and all it means is for 90% of the time you were at OR BELOW that number. 10% of the time you were above that number. How far above depends on what your maximum setting might have been and how long the machine was at the max or close to it. It does NOT mean you were there at that pressure for 90% of the night.

Really need to see the detailed graphs but while it's obvious the 13 cm minimum isn't getting the job done very well I don't know that you will need to go all the way to 16 as a minimum. It will probably be close though.

Are you sleeping on your back very much? Back sleeping will sometimes affect the pressure needed for some people.
It might explain some of the good vs bad nights...if you happen to be one of those people whose OSA is worse when on their back and need more pressure. It's fairly common. Maybe some nights you were on your back more than other nights.
REM sleep stage same thing...and fairly common but we can't do anything about REM...we can sometimes do something about back sleeping.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: I'm beginning to despair

Post by raisedfist » Sun Feb 05, 2017 6:44 pm

Don't despair, you have identified that your current therapy is not working, which is great because many people on this forum can help you.

_________________
Mask
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

roadrunnerNM
Posts: 7
Joined: Thu Jan 05, 2017 9:36 am

Re: I'm beginning to despair

Post by roadrunnerNM » Sun Feb 05, 2017 8:52 pm

More like this?

Image
Pugsy wrote: Are you sleeping on your back very much? Back sleeping will sometimes affect the pressure needed for some people.
It might explain some of the good vs bad nights...if you happen to be one of those people whose OSA is worse when on their back and need more pressure. It's fairly common. Maybe some nights you were on your back more than other nights.
REM sleep stage same thing...and fairly common but we can't do anything about REM...we can sometimes do something about back sleeping.
I wondered the same thing; so I've slept with a pillow jammed behind my back for about a week or so. I'm pretty sure I've been on my side since I get pressure points on my hips (which I would consider a totally reasonable trade-off for a good night's sleep). But it hasn't helped, as far as I can tell.

_________________
Mask: Simplus Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamStation BiPAP autoSV

User avatar
Pugsy
Posts: 64027
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: I'm beginning to despair

Post by Pugsy » Sun Feb 05, 2017 8:57 pm

Open your maximum back to 20 cm.
Are you having any problems using the 13 cm starting point?....problems breathing in and out at that pressure?
You are fast approaching the need for a bilevel machine in terms of pressure needs.

Do you think you could handle changing the minimum to 15? If that's too much try 14 for a few nights then try going up to 15.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: I'm beginning to despair

Post by palerider » Sun Feb 05, 2017 9:49 pm

roadrunnerNM wrote: I get pressure points on my hips
try adding some thickness of pillow just above your hips when sleeping on your side, reduces the pressure on the hip bones.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

D.H.
Posts: 3469
Joined: Sun Sep 04, 2016 7:07 pm

Re: I'm beginning to despair

Post by D.H. » Mon Feb 06, 2017 4:47 pm

Pugsy wrote:Open your maximum back to 20 cm.
Are you having any problems using the 13 cm starting point?....problems breathing in and out at that pressure?
You are fast approaching the need for a bilevel machine in terms of pressure needs.

Do you think you could handle changing the minimum to 15? If that's too much try 14 for a few nights then try going up to 15.
Yes, I agree to put the max as high as it goes, which is 20. You are actually hitting the celing you set of 16, so you want to let it go higher. You don't seem to have any significant "clear airway" apneas, so you should be OK with this. If you're having issues falling asleep, you can try turning the minimum down and/or using the ramp. Personally, I turned my ramp off some time in 1999 and never used it again; couldn't stand it! However, that doesn't mean that nobody else should use the ramp.

Also, if you're topping still topping out at 20, you'll need to consider a special machine that goes higher (usually a Bi-PAP will go as high as 25, a few models can go higher).

Furthermore, if that doesn't work, you can try "pressure relief" *(another feature that I couldn't stand).

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max