Staring to have acne around where my mask sits

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
pettyfan45
Posts: 160
Joined: Fri Jul 10, 2015 9:34 am
Contact:

Staring to have acne around where my mask sits

Post by pettyfan45 » Fri Sep 23, 2016 1:04 am

So I have been on CPAP for over a month now and it gets rather hot here and I wake up with sweat around where my mask sits and I think it is causing to have acne break outs where my mask sits. Any suggestions on how to deal with that?

_________________
MachineMask
Additional Comments: Pressure at 7-20, Climateline Air heated tube

D.H.
Posts: 3469
Joined: Sun Sep 04, 2016 7:07 pm

Re: Staring to have acne around where my mask sits

Post by D.H. » Fri Sep 23, 2016 5:16 am

First of all, wash the mask - at least the part that comes in direct contact with the affected areas - with soap and water every time you finish using the machine. Wash you face as well at that time.

Also, when you change the mask or the part in question, soak the new mask (or face piece) in 50% vinegar solution for at least an hour.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

nobody
Posts: 1038
Joined: Mon Dec 15, 2008 8:50 am

Re: Staring to have acne around where my mask sits

Post by nobody » Fri Sep 23, 2016 5:27 am

I have the same problem. I wash my mask and face before using but it doesn't matter. My solution was to alternate masks (2 different styles), that way it's not sitting on the same spot night after night. I wouldn't recommend washing or soaking your mask in vinegar or anything with a strong smell because it can linger around for a long time and irritate nose and throat.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Additional Comments: Respironics Simplicity nasal mask small

User avatar
ChicagoGranny
Posts: 14454
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Staring to have acne around where my mask sits

Post by ChicagoGranny » Fri Sep 23, 2016 5:51 am

pettyfan45 wrote:So I have been on CPAP for over a month now and it gets rather hot here and I wake up with sweat around where my mask sits and I think it is causing to have acne break outs where my mask sits. Any suggestions on how to deal with that?
Wash your face with a non-comedogenic soap (https://www.google.com/search?q=Non-Com ... 8&oe=utf-8) just before bed each night. See here for additional tips- http://www.menscience.com/Comedogenic-v ... p_177.html
it gets rather hot here and I wake up with sweat
Turn the A/C thermostat down at bedtime.

User avatar
chunkyfrog
Posts: 34385
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: Staring to have acne around where my mask sits

Post by chunkyfrog » Fri Sep 23, 2016 7:23 am

pettyfan45 wrote:So I have been on CPAP for over a month now and it gets rather hot here and I wake up with sweat around where my mask sits and I think it is causing to have acne break outs where my mask sits. Any suggestions on how to deal with that?
Check out PadaCheek.com. Karen is a forum member and long time papper.
A mask liner will help keep your mask cool in summer; and make it warm and cuddly in winter.
It will also extend the life of the cushion on your mask.
Her liners are washable and reusable for quite a long time.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

nanwilson
Posts: 3464
Joined: Tue Apr 13, 2010 10:35 am
Location: Southern Alberta

Re: Staring to have acne around where my mask sits

Post by nanwilson » Fri Sep 23, 2016 9:03 am

A quick trial to see if a cover would work.........
Take a well washed white t-shirt, lay it flat and place your mask on it. Now trace your mask onto the fabric .... adding about 3/4 of an inch all round. Cut this out and then cut a mouth and nose hole so that you can breath through it. Voila!!! you have a soft fabric between the mask and your face to sop up any sweat or face oils. Karen's padacheek liners are made of much more durable fabric and are made to fit each mask type differently, as Chunkyfrog said, they can be washed and dried almost indefinitely. I have had one for every ffm I have owned and recommend them to anyone who complains about their mask. Hint.. hint.. they also cut the sound of any mask farts and leaks you have. its a win, win!
Cheers
Nan
Started cpap in 2010.. still at it with great results.

User avatar
chunkyfrog
Posts: 34385
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: Staring to have acne around where my mask sits

Post by chunkyfrog » Fri Sep 23, 2016 9:50 am

You can even cut a liner out of a soft Viva paper towel.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Staring to have acne around where my mask sits

Post by palerider » Fri Sep 23, 2016 10:49 am

noise wrote:Also, when you change the mask or the part in question, soak the new mask (or face piece) in 50% vinegar solution for at least an hour.
that makes no sense at all.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Last edited by palerider on Tue Aug 07, 2018 1:08 pm, edited 1 time in total.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
knothead
Posts: 538
Joined: Fri Jan 24, 2014 8:58 am
Location: Corpus Christi, Tx

Re: Staring to have acne around where my mask sits

Post by knothead » Fri Sep 23, 2016 11:02 am

D.H. wrote:First of all, wash the mask - at least the part that comes in direct contact with the affected areas - with soap and water every time you finish using the machine. Wash you face as well at that time.

Also, when you change the mask or the part in question, soak the new mask (or face piece) in 50% vinegar solution for at least an hour.



I wouldn't do this. When I first started using my cpap I tried cleaning it with vinegar....not a good thing. Left red burn marks on sides of my nose. Use a non scented Dawn soap every night to clean the cushion, it will help, along with some of the other comments.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: BP down & in control, Resmed S9 Autoset for backup & travel.

D.H.
Posts: 3469
Joined: Sun Sep 04, 2016 7:07 pm

Re: Staring to have acne around where my mask sits

Post by D.H. » Fri Sep 23, 2016 5:21 pm

palerider wrote:
D.H. wrote:Also, when you change the mask or the part in question, soak the new mask (or face piece) in 50% vinegar solution for at least an hour.
stupid.
This time, I'll reply to this rude person.

I always soak new face pieces in vinegar solution. I am using the ComfortGel Blue Nasal CPAP Mask. It's a good mask for me, as I generally get a 100% fit. However, I did find that when I changed the face piece, I was getting a bad allergic reaction. The vinegar solution solved that problem. If the vinegar odor is objectionable, just wash very thoroughly afterwards.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Staring to have acne around where my mask sits

Post by palerider » Fri Sep 23, 2016 5:30 pm

D.H. wrote:
palerider wrote:
D.H. wrote:Also, when you change the mask or the part in question, soak the new mask (or face piece) in 50% vinegar solution for at least an hour.
stupid.
I always soak new face pieces in vinegar solution. .
you can do whatever you want, nobody cares, it's when you spread your stupid ideas without any thought that it becomes a problem for others.

I'll quit being "rude" just as soon as you quit posting stupid shit.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

D.H.
Posts: 3469
Joined: Sun Sep 04, 2016 7:07 pm

Re: Staring to have acne around where my mask sits

Post by D.H. » Fri Sep 23, 2016 5:44 pm

knothead wrote:
D.H. wrote:First of all, wash the mask - at least the part that comes in direct contact with the affected areas - with soap and water every time you finish using the machine. Wash you face as well at that time.

Also, when you change the mask or the part in question, soak the new mask (or face piece) in 50% vinegar solution for at least an hour.



I wouldn't do this. When I first started using my cpap I tried cleaning it with vinegar....not a good thing. Left red burn marks on sides of my nose. Use a non scented Dawn soap every night to clean the cushion, it will help, along with some of the other comments.
Funny, I was allergic to the latex (or whatever the material is). You're obviously have a contact allergy to vinegar (even if you eat it in your food). If you have a contact vinegar allergy, obviously this won't work for you! Also, I said only to do this to the face piece, not to any other piece.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

tlohse
Posts: 208
Joined: Fri Jun 27, 2014 5:32 pm
Location: West Bend, Wisconsin
Contact:

Re: Staring to have acne around where my mask sits

Post by tlohse » Fri Sep 23, 2016 6:11 pm

I soak my mask in a half vinager and half water solution for cleaning. always add some water to the vinager. Acne comes from not washing your mask or your face often enough and not changing your towel you wash your face with.
Thomas Lohse
Machine: ResMed AirSense 10 Auto Set With Heated hose and Humidifier
Mask: ResMed AirFit P10 Size M Nasal Pillow with headgear
Mask: ResMed AirFit N30i/P30i
Original Mask: ResMed Mirage FX Nasal
CPAP USER SINCE JUNE 2013

Cici
Posts: 5
Joined: Sun Apr 08, 2018 9:43 am

Re: Staring to have acne around where my mask sits

Post by Cici » Sun Apr 08, 2018 12:58 pm

chunkyfrog wrote:
Fri Sep 23, 2016 7:23 am
pettyfan45 wrote:So I have been on CPAP for over a month now and it gets rather hot here and I wake up with sweat around where my mask sits and I think it is causing to have acne break outs where my mask sits. Any suggestions on how to deal with that?
Check out PadaCheek.com. Karen is a forum member and long time papper.
A mask liner will help keep your mask cool in summer; and make it warm and cuddly in winter.
It will also extend the life of the cushion on your mask.
Her liners are washable and reusable for quite a long time.
I was having the same issue and saw this suggestion....a couple of weeks ago, I bought two liners from Padacheek....I have had the best night's sleep since I started using them. They not only eliminated the sweat/oil issues, they also took care of the slight leakage I was having at the top of my mask which kept waking me up at night. WIN-WIN! :P

_________________
MachineMask
Additional Comments: S10 AutoSet 4-8
CiCi

User avatar
chunkyfrog
Posts: 34385
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: Staring to have acne around where my mask sits

Post by chunkyfrog » Sun Apr 08, 2018 7:02 pm

It surprises me that providers almost universally know nothing of mask liners.
But maybe it should not . . .

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her