Do S9 auto's wear out?
Do S9 auto's wear out?
Hi,
My pressure I thought was very high over the last 3 years. My ResMed S9 autoset APAP needed to be bumped from 9 min 20 max to 14 min 20 max. My nightly average was 19.1 or 20. My AHI was all over the map from 7AHI to 20 AHI. I do use a full face mask called Quattro Pro. Sleepyhead shows I have no leaks.
I recently bought a Z1 travel APAP set at 11-20. My pressure stays at 11 it never goes up above 12. But my AHI went down to 1 or 2. I didn’t believe the night log tool. I verified my findings with a Oxy meter CMS 50E. With the Res Med s9 I am up and down from 95 down to 88 SPO2. With the Z1 its above 95 SPO2.
I wonder if my S9 is broken. I called ResMed they said call the vendor and wouldn’t talk about if it was broken or not. Its 4 years old. It’ll cost me $50.00 to return it.
I think it is out of warrantee.
Will the internet vendor do anything for me? Do these things just wear out? I just forked out $900 for the Z1. I do like it but it’s a little noisy. I did like the S9 being quiet. Maybe the S9 is just old technology?
Thanks in advance,
Phil
My pressure I thought was very high over the last 3 years. My ResMed S9 autoset APAP needed to be bumped from 9 min 20 max to 14 min 20 max. My nightly average was 19.1 or 20. My AHI was all over the map from 7AHI to 20 AHI. I do use a full face mask called Quattro Pro. Sleepyhead shows I have no leaks.
I recently bought a Z1 travel APAP set at 11-20. My pressure stays at 11 it never goes up above 12. But my AHI went down to 1 or 2. I didn’t believe the night log tool. I verified my findings with a Oxy meter CMS 50E. With the Res Med s9 I am up and down from 95 down to 88 SPO2. With the Z1 its above 95 SPO2.
I wonder if my S9 is broken. I called ResMed they said call the vendor and wouldn’t talk about if it was broken or not. Its 4 years old. It’ll cost me $50.00 to return it.
I think it is out of warrantee.
Will the internet vendor do anything for me? Do these things just wear out? I just forked out $900 for the Z1. I do like it but it’s a little noisy. I did like the S9 being quiet. Maybe the S9 is just old technology?
Thanks in advance,
Phil
- Wulfman...
- Posts: 6692
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: Do S9 auto's wear out?
You can check it yourself (for free) by using the method in the following link.mnPhil wrote:Hi,
My pressure I thought was very high over the last 3 years. My ResMed S9 autoset APAP needed to be bumped from 9 min 20 max to 14 min 20 max. My nightly average was 19.1 or 20. My AHI was all over the map from 7AHI to 20 AHI. I do use a full face mask called Quattro Pro. Sleepyhead shows I have no leaks.
I recently bought a Z1 travel APAP set at 11-20. My pressure stays at 11 it never goes up above 12. But my AHI went down to 1 or 2. I didn’t believe the night log tool. I verified my findings with a Oxy meter CMS 50E. With the Res Med s9 I am up and down from 95 down to 88 SPO2. With the Z1 its above 95 SPO2.
I wonder if my S9 is broken. I called ResMed they said call the vendor and wouldn’t talk about if it was broken or not. Its 4 years old. It’ll cost me $50.00 to return it.
I think it is out of warrantee.
Will the internet vendor do anything for me? Do these things just wear out? I just forked out $900 for the Z1. I do like it but it’s a little noisy. I did like the S9 being quiet. Maybe the S9 is just old technology?
Thanks in advance,
Phil
Or, you can either buy a gauge manometer from CPAP.COM for about $78.
viewtopic/t113072/viewtopic.php?f=1&t=4 ... ter#p37259
I used that (free) method for a number of years to check some of mine and then broke down and bought a gauge manometer in 2011 to double check them.
I would suggest setting the machine to a straight pressure when checking it.......and maybe a number of different settings to verify several pressures. If it's set for a range, it may or may not self-adjust to a different pressure.
Most machines have a plus or minus of .5 to 1.0 centimeter of leeway on pressure accuracy.
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: Do S9 auto's wear out?
Hi,
Excellent idea. I love the pressure check with water.
I am going to check it this weekend.
Thanks again
God bless
Phil
Excellent idea. I love the pressure check with water.
I am going to check it this weekend.
Thanks again
God bless
Phil
Re: Do S9 auto's wear out?
They only wear out if they are man made and have parts, and are turned on. Jim
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
Re: Do S9 auto's wear out?
Or, you can buy a digital manometer from ebay for under 30$.Wulfman... wrote:Or, you can either buy a gauge manometer from CPAP.COM for about $78.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Sir NoddinOff
- Posts: 4189
- Joined: Mon May 14, 2012 5:30 pm
- Location: California
Re: Do S9 auto's wear out?
Or take it to a local DME and sweet talk them into testing it for free (it just takes a few minutes). Nowadays tho, it's mostly big national chains with no local outlets. <sigh>
_________________
| Mask: AirFit™ F10 Full Face Mask with Headgear |
| Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2. |
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.
- Wulfman...
- Posts: 6692
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: Do S9 auto's wear out?
Many/most hospitals have Respiratory Therapists who could check it, too. If a person has any "connections" (contacts) to their local hospitals, that might be an option.Sir NoddinOff wrote:Or take it to a local DME and sweet talk them into testing it for free (it just takes a few minutes). Nowadays tho, it's mostly big national chains with no local outlets. <sigh>
But, the bucket of water with a ruler method is still pretty accurate for occasional checking.
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: Do S9 auto's wear out?
That's assuming they know what a manometer is. Clearly, some get what you mean about testing the pressure and some don't. If they don't know what a manometer is, chances are, you need to talk to a tech or go somewhere else.Sir NoddinOff wrote:Or take it to a local DME and sweet talk them into testing it for free (it just takes a few minutes). Nowadays tho, it's mostly big national chains with no local outlets. <sigh>
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Sleepyhead |
Diabetes 2, RLS & bradycardia
Airsense For Her; Settings: range 8-12, Airfit P10 (M)
Airsense For Her; Settings: range 8-12, Airfit P10 (M)
Re: Do S9 auto's wear out?
Or go to the library and have them let you see a dictionary, but it's easier to use google. JimSewTired wrote:That's assuming they know what a manometer is. Clearly, some get what you mean about testing the pressure and some don't. If they don't know what a manometer is, chances are, you need to talk to a tech or go somewhere else.Sir NoddinOff wrote:Or take it to a local DME and sweet talk them into testing it for free (it just takes a few minutes). Nowadays tho, it's mostly big national chains with no local outlets. <sigh>
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
Re: Do S9 auto's wear out?
Hi,
I wanted to report back. I always what to know what happens after I give advice. I turned the S9 to cpap mode. I set the pressure at 11 and 17.8. I put a 3 rubber bands around the hose and the ruler and did the water trick.
The S9 at cpap 11 was 4 and 1/3 inches.
The s9 at cpap 17.8 was 7 inches.
So I think the pressure is working.
I set the s9 back to apap 12 min 20 max and will check how it goes tonight. I did turn EPR to of I used to run it at 3.
So I don’t think I need to throw my old s9 machine.
I’m still not sure why my new little Z1 gives me better AHI numbers.
Thanks again for saving me some $$.
Phil
I wanted to report back. I always what to know what happens after I give advice. I turned the S9 to cpap mode. I set the pressure at 11 and 17.8. I put a 3 rubber bands around the hose and the ruler and did the water trick.
The S9 at cpap 11 was 4 and 1/3 inches.
The s9 at cpap 17.8 was 7 inches.
So I think the pressure is working.
I set the s9 back to apap 12 min 20 max and will check how it goes tonight. I did turn EPR to of I used to run it at 3.
So I don’t think I need to throw my old s9 machine.
I’m still not sure why my new little Z1 gives me better AHI numbers.
Thanks again for saving me some $$.
Phil
Re: Do S9 auto's wear out?
Perhaps differences in the reporting criteria for the various apnea events?mnPhil wrote:I’m still not sure why my new little Z1 gives me better AHI numbers.
Meaning maybe the Z1 has a stricter criteria compared to the ResMed event definitions.
Like maybe the % of reduction needed to meet criteria is not as much with the ResMed hence more events getting flagged by the ResMed because the range to earn a flag is larger.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Do S9 auto's wear out?
Another possibility is the algorithm that the Z1 works better for you than the Resmed algorithm. Horses for courses.mnPhil wrote:Hi,
I wanted to report back. I always what to know what happens after I give advice. I turned the S9 to cpap mode. I set the pressure at 11 and 17.8. I put a 3 rubber bands around the hose and the ruler and did the water trick.
The S9 at cpap 11 was 4 and 1/3 inches.
The s9 at cpap 17.8 was 7 inches.
So I think the pressure is working.
I set the s9 back to apap 12 min 20 max and will check how it goes tonight. I did turn EPR to of I used to run it at 3.
So I don’t think I need to throw my old s9 machine.
I’m still not sure why my new little Z1 gives me better AHI numbers.
Thanks again for saving me some $$.
Phil
Re: Do S9 auto's wear out?
Hi,
With the z1 I start at 11 and go up to 12.5 pressure.
Last night I tried my s9. App pressure set at 12 min 20 max no edr. My ahi was 12.5 Horrible.
I have a cms50e. My o2 was dropping down to the low 90s.
The pressure shot up to 20. No leaks.
Do any apaps try to do burst air? Instead of a steady pressure increase. I like the z1 but it a little noisy and does not have water.
Thanks
Phil
With the z1 I start at 11 and go up to 12.5 pressure.
Last night I tried my s9. App pressure set at 12 min 20 max no edr. My ahi was 12.5 Horrible.
I have a cms50e. My o2 was dropping down to the low 90s.
The pressure shot up to 20. No leaks.
Do any apaps try to do burst air? Instead of a steady pressure increase. I like the z1 but it a little noisy and does not have water.
Thanks
Phil
Re: Do S9 auto's wear out?
I'm in the camp that thinks your S9 may have always had issues, especially since your AHI has been high from the beginning and you are running pressures so much higher. I highly question an AHI that changes from 1 to 2 to over 12 and up to 20 due to a machine change. I'm sure that Govt agencies that require OSA testing and monitoring as a criteria for employment would be all over it if it were just algorithms. I can you that I would have a tough time flying if the pilot had OSA and was running at 20AHI but was able to cheat the system by simply changing machines.
I do have an almost new older S8 I was using as a backup on vacation that gave me horrible numbers and I never felt very good in the morning after using it.
The big question is how do you feel?
You may want to test your S9 at 11 or 12 for a few hours to see if what happens. It may start to decrease as the machine warms up and parts that are close to being out of tolerance fail to maintain the pressure. After all, it is just a pump.
I recently got an S9 Autoset and, for me it is an improvement, over running straight CPAP. But I think straight CPAP machine will last longer than an APAP. My first CPAP machine was an S6 and that was going strong after 10 years. My S9 Elite is over 5 years old and has 14,000 hours. Even with lower pressures and a tighter range, I don't anticipate my Autoset will last as long.
John
I do have an almost new older S8 I was using as a backup on vacation that gave me horrible numbers and I never felt very good in the morning after using it.
The big question is how do you feel?
You may want to test your S9 at 11 or 12 for a few hours to see if what happens. It may start to decrease as the machine warms up and parts that are close to being out of tolerance fail to maintain the pressure. After all, it is just a pump.
I recently got an S9 Autoset and, for me it is an improvement, over running straight CPAP. But I think straight CPAP machine will last longer than an APAP. My first CPAP machine was an S6 and that was going strong after 10 years. My S9 Elite is over 5 years old and has 14,000 hours. Even with lower pressures and a tighter range, I don't anticipate my Autoset will last as long.
John
_________________
| Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
| Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
AHI: 2.5
Central: 1.7
Obstructive: 0.3
Hypopnea: 0.5
Pressure: 6.0-8.0cm on back with cervical collar.
Compliance: 15 Years
Central: 1.7
Obstructive: 0.3
Hypopnea: 0.5
Pressure: 6.0-8.0cm on back with cervical collar.
Compliance: 15 Years
Re: Do S9 auto's wear out?
No...none of the auto adjusting machines will do a burst of air unless you are using one of the ASV types of machines.mnPhil wrote:Do any apaps try to do burst air? Instead of a steady pressure increase.
What is that AHI made up of?mnPhil wrote:My ahi was 12.5 Horrible.
What is the Central/Clear Airway Index?
What is the Obstructive Apnea Index?
What is the Hyponea Index?
Finally...with the Z machine...same questions...what is the AHI made up of? Does it flag Centrals or does it just flag OAs and Hyponeas?