Problem with dates on either SH or my F&P Icon Auto
Problem with dates on either SH or my F&P Icon Auto
Hi all,
When I import data from my F&P Icon Auto into SH, sometimes it goes in the wrong day and combines with existing dates. I literally have two lots of data from different nights (ie, in the events, it lists the event times ie 1:02 ... 1:32... 2:33... etc then 1:22... 2:02... 3:33... - these are made up times btw).
The time is set correctly. The timezone is set correctly and always has been (finally found out where it is in the menu).
I presumed this was a problem with SH but I've just noticed something odd on my machine.
Last night I changed from APAP mode to CPAP mode to try the prescribed pressure I was given the other day (7).
But, after midday when the machine gives me last night's numbers, I looked and it said the 95% pressure was 17. But it's 100% set to CPAP 7, and I could tell it was that low because 1) I felt crap when I woke up, and 2) my mask and hose was full of water which it never is.
So now I'm wondering if SH isn't to blame for the data problem I'm seeing, and it's the machine.
Does anyone have any ideas please?
Many thanks
John
When I import data from my F&P Icon Auto into SH, sometimes it goes in the wrong day and combines with existing dates. I literally have two lots of data from different nights (ie, in the events, it lists the event times ie 1:02 ... 1:32... 2:33... etc then 1:22... 2:02... 3:33... - these are made up times btw).
The time is set correctly. The timezone is set correctly and always has been (finally found out where it is in the menu).
I presumed this was a problem with SH but I've just noticed something odd on my machine.
Last night I changed from APAP mode to CPAP mode to try the prescribed pressure I was given the other day (7).
But, after midday when the machine gives me last night's numbers, I looked and it said the 95% pressure was 17. But it's 100% set to CPAP 7, and I could tell it was that low because 1) I felt crap when I woke up, and 2) my mask and hose was full of water which it never is.
So now I'm wondering if SH isn't to blame for the data problem I'm seeing, and it's the machine.
Does anyone have any ideas please?
Many thanks
John
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: New User. Prescription of 7. Still trying various APAP settings. |
- GettingBetter
- Posts: 371
- Joined: Thu Dec 03, 2015 5:08 pm
Re: Problem with dates on either SH or my F&P Icon Auto
I have the same machine and the same issue. I have not been able to resolve, but also haven't put in too much effort (priorities and my ahi hasn't been horrible).
Like you, the time is set correctly on the icon and my computer. I am assuming you also are uploading the Icon data from the memory stick from the Icon? Just so we are on the same plane here. My date issues also occur when an automatic change from APAP to CPAP also occurs. I can verify this with the charts since for each of those change dates I have only one session showing. It shows I don't meet compliance on one day and have an abundance of hours on the day before the change.
I am not making the change back and forth between APAP and CPAP. I'm pretty sure the machine is adjusting as programmed. The only thing I adjust on my machine is the ramp feature when I want to.
GB
Like you, the time is set correctly on the icon and my computer. I am assuming you also are uploading the Icon data from the memory stick from the Icon? Just so we are on the same plane here. My date issues also occur when an automatic change from APAP to CPAP also occurs. I can verify this with the charts since for each of those change dates I have only one session showing. It shows I don't meet compliance on one day and have an abundance of hours on the day before the change.
I am not making the change back and forth between APAP and CPAP. I'm pretty sure the machine is adjusting as programmed. The only thing I adjust on my machine is the ramp feature when I want to.
GB
johnrrr wrote:Hi all,
When I import data from my F&P Icon Auto into SH, sometimes it goes in the wrong day and combines with existing dates. I literally have two lots of data from different nights (ie, in the events, it lists the event times ie 1:02 ... 1:32... 2:33... etc then 1:22... 2:02... 3:33... - these are made up times btw).
The time is set correctly. The timezone is set correctly and always has been (finally found out where it is in the menu).
I presumed this was a problem with SH but I've just noticed something odd on my machine.
Last night I changed from APAP mode to CPAP mode to try the prescribed pressure I was given the other day (7).
But, after midday when the machine gives me last night's numbers, I looked and it said the 95% pressure was 17. But it's 100% set to CPAP 7, and I could tell it was that low because 1) I felt crap when I woke up, and 2) my mask and hose was full of water which it never is.
So now I'm wondering if SH isn't to blame for the data problem I'm seeing, and it's the machine.
Does anyone have any ideas please?
Many thanks
John
Re: Problem with dates on either SH or my F&P Icon Auto
Hi GB,GettingBetter wrote:I have the same machine and the same issue. I have not been able to resolve, but also haven't put in too much effort (priorities and my ahi hasn't been horrible).
Like you, the time is set correctly on the icon and my computer. I am assuming you also are uploading the Icon data from the memory stick from the Icon? Just so we are on the same plane here. My date issues also occur when an automatic change from APAP to CPAP also occurs. I can verify this with the charts since for each of those change dates I have only one session showing. It shows I don't meet compliance on one day and have an abundance of hours on the day before the change.
I am not making the change back and forth between APAP and CPAP. I'm pretty sure the machine is adjusting as programmed. The only thing I adjust on my machine is the ramp feature when I want to.
Yes, I am uploading the data from the USB stick.
I don't normally change from APAP to CPAP, although that's what I did last night. But the problem also happens on other days. I can't see a pattern in why it combines days.
But now I've noticed that the machine is saying my 95% pressure was 17 last night, when it was set on CPAP 7, that suggests strongly the machine is to blame, not the card or SH.
I really wanted to know my AHI last night as 7 is the prescribed pressure I got the other day and I'm sure it's too low. It's really frustrating not to be sure what it was
Do you know if there are any software updates for the machine? I think i read you can put them on the USB stick, insert it, and it auto-updates?
Thanks
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: New User. Prescription of 7. Still trying various APAP settings. |
- TangledHose
- Posts: 862
- Joined: Wed Mar 13, 2013 11:13 am
- Location: Centennial, CO
Re: Problem with dates on either SH or my F&P Icon Auto
On the Icon machine itself when viewing sleep data on the machines screen, you won't see the last nights data until after 12 noon the next day. So if you look at the machines data before noon you'll be looking at the data not from the last night, but the night before that.
Not sure if this has any bearing on importing to Sleepyhead, but you might try waiting till after the noon hour before importing the data and see if that makes any difference in SH reports.
Not sure if this has any bearing on importing to Sleepyhead, but you might try waiting till after the noon hour before importing the data and see if that makes any difference in SH reports.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Simplus Full Face CPAP Mask with Headgear |
Additional Comments: Also own F&P Icon Auto for backup and travel; and a Transcend when packing space is limited |
Link to Sleepyhead:
http://sleepfiles.com/SH2/
http://sleepfiles.com/SH2/
Re: Problem with dates on either SH or my F&P Icon Auto
Yes, I know I can't see it until midday but I hadn't thought of the idea that I should wait until midday to import the data. That's a good idea, thanks.
But separately from that, last night's data is still wrong on the machine.
Today I did import the data before midday into SH and it had the problem.
But I did wait until midday to look at the machine data. I've just looked again (nearly 4pm) and it's still wrong.
The machine data is wrong as it shows an APAP 95% pressure of 17, instead of the CPAP pressure of 7...
Unless taking the card out before 12 and putting it back in caused that problem?
But separately from that, last night's data is still wrong on the machine.
Today I did import the data before midday into SH and it had the problem.
But I did wait until midday to look at the machine data. I've just looked again (nearly 4pm) and it's still wrong.
The machine data is wrong as it shows an APAP 95% pressure of 17, instead of the CPAP pressure of 7...
Unless taking the card out before 12 and putting it back in caused that problem?
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: New User. Prescription of 7. Still trying various APAP settings. |
- GettingBetter
- Posts: 371
- Joined: Thu Dec 03, 2015 5:08 pm
Re: Problem with dates on either SH or my F&P Icon Auto
Okay, on the statistics report, there is a record of when my Icon switches from auto to fixed (I don't do this). This is "Changes to Perscription Settings" at the bottom of the report. On those days the machine will throw any current data sessions under the APAP to the day before so the APAP data stays together and the CPAP data sessions are together, regardless of the midday cut off. The last time it did this was Feb 10, which had non-compliant hours per Sleepyhead, even though I was compliant for that day. The upload sent to the DME shows me as compliant.johnrrr wrote:Hi GB,GettingBetter wrote:I have the same machine and the same issue. I have not been able to resolve, but also haven't put in too much effort (priorities and my ahi hasn't been horrible).
Like you, the time is set correctly on the icon and my computer. I am assuming you also are uploading the Icon data from the memory stick from the Icon? Just so we are on the same plane here. My date issues also occur when an automatic change from APAP to CPAP also occurs. I can verify this with the charts since for each of those change dates I have only one session showing. It shows I don't meet compliance on one day and have an abundance of hours on the day before the change.
I am not making the change back and forth between APAP and CPAP. I'm pretty sure the machine is adjusting as programmed. The only thing I adjust on my machine is the ramp feature when I want to.
Yes, I am uploading the data from the USB stick.
I don't normally change from APAP to CPAP, although that's what I did last night. But the problem also happens on other days. I can't see a pattern in why it combines days.
But now I've noticed that the machine is saying my 95% pressure was 17 last night, when it was set on CPAP 7, that suggests strongly the machine is to blame, not the card or SH.
I really wanted to know my AHI last night as 7 is the prescribed pressure I got the other day and I'm sure it's too low. It's really frustrating not to be sure what it was
Do you know if there are any software updates for the machine? I think i read you can put them on the USB stick, insert it, and it auto-updates?
Thanks
So what is going on is the Icon is doing what it has been set to do. It is automatically seeing if you are ready for a fixed setting or still need auto settings. Instead of you working the numbers, evaluating pressures it is doing it for you. You yourself can always set it to CPAP settings and set the pressure to a static number and see if that works, but this seems easier to me.
Am I worried about this? No. Is this a problem to report on SleepyHead? I don't think so. It is clear to me what is happening and putting CPAP and APAP sessions together just so dates match is silly.
Re: Problem with dates on either SH or my F&P Icon Auto
My overview screen says i switched to fixed CPAP of 7 on the 11th... but I did it yesterday on the 12th (in fact, I actually did it gone midnight, so it was the 13th but for the 'night' of the 12th).
I imported the data on the morning of the 13th (before noon - maybe around 8am?).
But the data went in to SH as the 11th, along with data that was already there on the 11th previously.
That's what I don't get. I can now see a potential problem of importing before noon (Even though I knew the machine didn't show last night's data until midday, I thought I'd read somewhere I could import before - but I'm guessing not).
But even if that was the problem, shouldn't it have imported the data into the 12th (Friday night)? But it imported it into the 11th (Thursday night)
Or am I totally confused?
I imported the data on the morning of the 13th (before noon - maybe around 8am?).
But the data went in to SH as the 11th, along with data that was already there on the 11th previously.
That's what I don't get. I can now see a potential problem of importing before noon (Even though I knew the machine didn't show last night's data until midday, I thought I'd read somewhere I could import before - but I'm guessing not).
But even if that was the problem, shouldn't it have imported the data into the 12th (Friday night)? But it imported it into the 11th (Thursday night)
Or am I totally confused?
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: New User. Prescription of 7. Still trying various APAP settings. |
- BlackSpinner
- Posts: 9745
- Joined: Sat Apr 25, 2009 5:44 pm
- Location: Edmonton Alberta
- Contact:
Re: Problem with dates on either SH or my F&P Icon Auto
It completely depends on the internal clock and software. Most equipment goes by Greenwich time because then it doesn't matter where the machine it. Your cell phone for instances corrects the display for your location. The same goes for your computer. That means your date doesn't change until the London time does if the software doesn't handle things well. Dates and time are one big giant headache, always have been when writing software. .johnrrr wrote:My overview screen says i switched to fixed CPAP of 7 on the 11th... but I did it yesterday on the 12th (in fact, I actually did it gone midnight, so it was the 13th but for the 'night' of the 12th).
I imported the data on the morning of the 13th (before noon - maybe around 8am?).
But the data went in to SH as the 11th, along with data that was already there on the 11th previously.
That's what I don't get. I can now see a potential problem of importing before noon (Even though I knew the machine didn't show last night's data until midday, I thought I'd read somewhere I could import before - but I'm guessing not).
But even if that was the problem, shouldn't it have imported the data into the 12th (Friday night)? But it imported it into the 11th (Thursday night)
Or am I totally confused?
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: Hybrid Full Face CPAP Mask with Nasal Pillows and Headgear |
Additional Comments: Quatro mask for colds & flus S8 elite for back up |
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal
Re: Problem with dates on either SH or my F&P Icon Auto
Agreed, I'm a programmer by background but in this case the machine is set to GMT+1, as is my computer - which is correct for where I am (Spain).BlackSpinner wrote:It completely depends on the internal clock and software. Most equipment goes by Greenwich time because then it doesn't matter where the machine it. Your cell phone for instances corrects the display for your location. The same goes for your computer. That means your date doesn't change until the London time does if the software doesn't handle things well. Dates and time are one big giant headache, always have been when writing software. .johnrrr wrote:My overview screen says i switched to fixed CPAP of 7 on the 11th... but I did it yesterday on the 12th (in fact, I actually did it gone midnight, so it was the 13th but for the 'night' of the 12th).
I imported the data on the morning of the 13th (before noon - maybe around 8am?).
But the data went in to SH as the 11th, along with data that was already there on the 11th previously.
That's what I don't get. I can now see a potential problem of importing before noon (Even though I knew the machine didn't show last night's data until midday, I thought I'd read somewhere I could import before - but I'm guessing not).
But even if that was the problem, shouldn't it have imported the data into the 12th (Friday night)? But it imported it into the 11th (Thursday night)
Or am I totally confused?
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: New User. Prescription of 7. Still trying various APAP settings. |
Re: Problem with dates on either SH or my F&P Icon Auto
Today (gone midday), my machine is still not showing the correct data (it shows 95% pressure at 17 still and the same AHI as two days ago).
But importing the card in SH does show the pressure was at a fixed 8 (I upped it last night) and an AHI of 3.33 - but it's imported it in to Friday 12th (it's Sunday 14th now).
If anyone has any real solution to this, or even ideas, that'd be great. Thanks.
But importing the card in SH does show the pressure was at a fixed 8 (I upped it last night) and an AHI of 3.33 - but it's imported it in to Friday 12th (it's Sunday 14th now).
If anyone has any real solution to this, or even ideas, that'd be great. Thanks.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: New User. Prescription of 7. Still trying various APAP settings. |
Re: Problem with dates on either SH or my F&P Icon Auto
have you tried unplugging it and letting the little computer inside reset?johnrrr wrote:Today (gone midday), my machine is still not showing the correct data (it shows 95% pressure at 17 still and the same AHI as two days ago).
But importing the card in SH does show the pressure was at a fixed 8 (I upped it last night) and an AHI of 3.33 - but it's imported it in to Friday 12th (it's Sunday 14th now).
If anyone has any real solution to this, or even ideas, that'd be great. Thanks.
I've heard of some resmed s9 machines that started reporting 0 ahi and needing a reset... maybe yours has a 'stuck pressure' bug?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Problem with dates on either SH or my F&P Icon Auto
Thanks. I have unplugged it a few times but nothing...palerider wrote:have you tried unplugging it and letting the little computer inside reset?johnrrr wrote:Today (gone midday), my machine is still not showing the correct data (it shows 95% pressure at 17 still and the same AHI as two days ago).
But importing the card in SH does show the pressure was at a fixed 8 (I upped it last night) and an AHI of 3.33 - but it's imported it in to Friday 12th (it's Sunday 14th now).
If anyone has any real solution to this, or even ideas, that'd be great. Thanks.
I've heard of some resmed s9 machines that started reporting 0 ahi and needing a reset... maybe yours has a 'stuck pressure' bug?
I decided I was going to wipe the machine (I have data on SH) and start again. The instructions in the manual I have didn't work but a search found an alternative method (I had to set it to Menu 2, instead of Menu 1)...
Just before I hit the final 'reset' button I thought I'd check the figures again and would you believe it... it is now showing last night's data correctly!
No idea why, but ...
So, see what tomorrow brings!
I put the newly wiped usb stick into the machine and copied the data off, and put it in to a new SH profile though and it's still coming up on the wrong dates, so that problem is on-going (well, maybe it will be fixed going forward, but it's not retrospectively fixed).
Thanks