My new Airsense 10 arrived yesterday. Thought I set the date and time correctly before I went to sleep, but I see now that it is an hour fast.
When I tried to change time from the displayed 12:30 to the correct 11:30, the cheeky machine has the gall to tell me that "the time cannot be set in the past"!
Anyone ever seen something like this?
Doug
New PAP won't let me set correct time!
Re: New PAP won't let me set correct time!
This is what the manual says.
If you set a new time that occurs in the past then an error message is
displayed. Before this change can be made, erase the compliance data
available under the Configuration menu.
If you have any compliance data on the SD card you might want to save that data. Then you have to erase the data from the machine and SD card. You do this by going into the clinical settings menu - Configuration.
If you set a new time that occurs in the past then an error message is
displayed. Before this change can be made, erase the compliance data
available under the Configuration menu.
If you have any compliance data on the SD card you might want to save that data. Then you have to erase the data from the machine and SD card. You do this by going into the clinical settings menu - Configuration.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Sleepyhead Software |
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760
Re: New PAP won't let me set correct time!
A) Thanks very much!
B) What a stupid paradigm.... Erase all my data to change the clock? Happy - really, really happy - I discovered this on Day Two and not on Day Eighty-seven!!
B) What a stupid paradigm.... Erase all my data to change the clock? Happy - really, really happy - I discovered this on Day Two and not on Day Eighty-seven!!
Re: New PAP won't let me set correct time!
Yes it is. I believe it has something to do with preventing people from trying to use the clock to falsify their compliance but I'm not sure. Respironics has a clock set at the factory and cannot be set by the user.dvejr wrote:A) Thanks very much!
B) What a stupid paradigm.... Erase all my data to change the clock? Happy - really, really happy - I discovered this on Day Two and not on Day Eighty-seven!!
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Sleepyhead Software |
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760
Re: New PAP won't let me set correct time!
wait until the machine rolls over to a new day (noon by it's clock) and then you can set the time back.dvejr wrote:A) Thanks very much!
B) What a stupid paradigm.... Erase all my data to change the clock? Happy - really, really happy - I discovered this on Day Two and not on Day Eighty-seven!!
you just can't set the time back when there is data for that day.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: New PAP won't let me set correct time!
Thanks PR, I was getting information from the manual and other users. My own machine has been off 10 minutes and I got the same message but it really didn't matter to me so I left it that way. I just went back and changed mine like you said. So now I have 1st hand experience.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Sleepyhead Software |
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760
Re: New PAP won't let me set correct time!
I'm pretty sure I got that from posts here on the forum, not out of one of the manuals
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: New PAP won't let me set correct time!
What the machine won't allow you to do is set the clock back in time before the last minute for which you have data. For instance if the machine thinks it was 9 AM January 1 when you last used the machine, you can't set the clock back before 9 AM Jan 1.palerider wrote:wait until the machine rolls over to a new day (noon by it's clock) and then you can set the time back.dvejr wrote:A) Thanks very much!
B) What a stupid paradigm.... Erase all my data to change the clock? Happy - really, really happy - I discovered this on Day Two and not on Day Eighty-seven!!
you just can't set the time back when there is data for that day.
If your clock is 1 hour ahead, turn the machine off and don't use it for a little over an hour. After that, you can set the time back one hour. Even easier, just turn the machine off when you get up, and then try to set the clock back before going to bed, you'll probably never see the problem.
No need to erase your data unless you've got a big time jump to make.
BTW, you can always set the clock forward without resetting data.
Source: I've dealt with this several times with my S9 AutoSet for DST changes and have experimented with the possible changes. I presume the A10 models do the same.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.