Ammonia smell coming from machine
Ammonia smell coming from machine
The last two nights when my husband has turned on his cpap and put on his mask there has been a sudden, nose burning smell of ammonia getting blown out. He can't smell anything and there's nothing around the house that smells of it. His breath smells normal without the mask. In over ten years I've never smelled anything like it coming from him. It's so bad I got out of bed. The inside of my nose still burns. Any help?
- BlackSpinner
- Posts: 9745
- Joined: Sat Apr 25, 2009 5:44 pm
- Location: Edmonton Alberta
- Contact:
Re: Ammonia smell coming from machine
Have you rinsed the humidifier well? What is being cleaned and how?
What happens when you try it?
Ammonia can only come from chemicals. It doesn't come that way.
What happens if you have the humidifier turned off or not in use?
What happens when you try it?
Ammonia can only come from chemicals. It doesn't come that way.
What happens if you have the humidifier turned off or not in use?
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: Hybrid Full Face CPAP Mask with Nasal Pillows and Headgear |
Additional Comments: Quatro mask for colds & flus S8 elite for back up |
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal
-
- Posts: 101
- Joined: Mon Sep 28, 2015 2:04 pm
- Location: Halfway between Comfort and Welfare, Texas
Re: Ammonia smell coming from machine
New to me. Clean everything and make sure there is no issue with the motor on the CPAP machine. Burning smell could be a bearing or wire getting hot, and your husband cant smell it due to the filter on the machine.
That is very unusual.
Clean first and try it out. If still there, have your husband run the machine with mask off so he can smell it.
Changed the filter lately? Cleaned the hose?
Rich
That is very unusual.
Clean first and try it out. If still there, have your husband run the machine with mask off so he can smell it.
Changed the filter lately? Cleaned the hose?
Rich
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Re: Ammonia smell coming from machine
Stomach acid backing up....?....??? Jim
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
Re: Ammonia smell coming from machine
You are talking about his breath at night. I would have his blood sugar and kidneys checked.CPAPpartner wrote:The last two nights when my husband has turned on his cpap and put on his mask there has been a sudden, nose burning smell of ammonia getting blown out. He can't smell anything and there's nothing around the house that smells of it. His breath smells normal without the mask. In over ten years I've never smelled anything like it coming from him. It's so bad I got out of bed. The inside of my nose still burns. Any help?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Sleepyhead Software |
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760
Re: Ammonia smell coming from machine
Thinking the same thing. Same type of odor from neighbor's machine. I believe it was a wire getting too close to something which burned and caused the odor. He took it to his DME and they were able to fix it on the spot. Also, change the filter just in case.R1200ST wrote:Burning smell could be a bearing or wire getting hot, and your husband cant smell it due to the filter on the machine.
Resmed AirSense 10 Autoset for her w/humid air/heated Humidifier
Bleep/P10
Bleep/P10
- GettingBetter
- Posts: 371
- Joined: Thu Dec 03, 2015 5:08 pm
Re: Ammonia smell coming from machine
Sorry you are having these problems. It can be scary and just posting the question shows you are the kind of person that will act, not just keep asking questions.
An acrid smell from the machine is most likely hot metal. As posted by others here, taking the machine to your DME should resolve the problem.
An acrid, urine or ammonia smell coming from your husband's breath can be a symptom of serious disease and needs to be checked out immediately. A trip to the doctor is warranted here.
Hope you find the source and a remedy for your problem soon!
An acrid smell from the machine is most likely hot metal. As posted by others here, taking the machine to your DME should resolve the problem.
An acrid, urine or ammonia smell coming from your husband's breath can be a symptom of serious disease and needs to be checked out immediately. A trip to the doctor is warranted here.
Hope you find the source and a remedy for your problem soon!
Last edited by GettingBetter on Sun Feb 07, 2016 6:54 pm, edited 1 time in total.
Re: Ammonia smell coming from machine
A dry humidifier will cause a terrible smell.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is S9 Autoset...... |
- Stormynights
- Posts: 2273
- Joined: Wed Mar 10, 2010 7:01 pm
- Location: Oklahoma
Re: Ammonia smell coming from machine
Do you have pets?
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Pressure EPAP 5.8 IPAP 9.4-21.8 PS 3.6/16 S9 Vpap Adapt ASV |
- chunkyfrog
- Posts: 34397
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Ammonia smell coming from machine
Go to the doctor! Ammonia might be OK on a fireplug, but not in your breath or cpap.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
-
- Posts: 1562
- Joined: Tue May 06, 2008 5:57 pm
- Location: Murrysville, PA
Re: Ammonia smell coming from machine
The frog is right. Search the internet for exhaled breath condensate (EBC) and ammonia. An increase in ammonia is a marker of pneumonia and other lung conditions, including COPD and cystic fibrosis.
http://www.ncbi.nlm.nih.gov/pubmed/19670404
http://onlinelibrary.wiley.com/doi/10.1 ... 8/abstract
To narrow this down, run the machine without the patient to observe for any ammonia odors. Compare this to the mask exhaust while your husband uses the machine. Is it the machine or the man?
http://www.ncbi.nlm.nih.gov/pubmed/19670404
http://onlinelibrary.wiley.com/doi/10.1 ... 8/abstract
To narrow this down, run the machine without the patient to observe for any ammonia odors. Compare this to the mask exhaust while your husband uses the machine. Is it the machine or the man?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Sleepyhead software. Just changed from PRS1 BiPAP Auto DS760TS |