CFLEX AND CFLEX+

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Carry On
Posts: 5
Joined: Sun Nov 22, 2015 4:37 pm

CFLEX AND CFLEX+

Post by Carry On » Mon Nov 30, 2015 10:53 pm

IS ONE BETTER THAN THEORY OTHER?

SET CFLEX PER THE DR. POST SLEEP STUDY, SHOWING MILD TO MODERATE SLEEP APNEA, TO 12 PRESSURE AND 1 FLEX. NOTICED BIG DROP IN APNI BUT SOMETIMES HAD TO EXHALE SLIGHTLY FROM MY MOUTH WHEN THE PRESSURE BUILT UP THERE. I USE THE DREAM PADS ON THE NOSE ONLY. NOT A MOUTH BREATHER.

DOES THIS MATTER OR SEEM NORMAL? I THINK THE DREAM MACHINE SEES THIS A SMALL MASK LEAK OF ABOUT 1 OR 2 %.

THANKS.......

_________________
Machine: DreamStation Auto CPAP Machine
Mask: DreamWear Nasal CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: CPAP and CFLEX+ @12

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: CFLEX AND CFLEX+

Post by Wulfman... » Tue Dec 01, 2015 12:26 am

They're both exhale relief. You can pick whichever one feels better to you or turn it off.
I used C-Flex at a setting of 2 for about 8 years and then decided to try turning it off entirely and wondered why I hadn't done it long ago.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Pugsy
Posts: 64020
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: CFLEX AND CFLEX+

Post by Pugsy » Tue Dec 01, 2015 8:44 am

CFlex is just the slightly older version of CFlex +...the Plus part was added a little later for more gentler transition during the exhale relief.

I think all the newer machines come with CFlex + or AFlex depending on mode choice.
My first M Series machine many years ago only had CFlex available even in APAP mode but later M series machine that I got had AFlex and CFlex +.

Doctors usually just talk about CFlex or Flex in general and don't even bother with the + part of the description.

I terms of therapy performance the plus or lack of a plus isn't going to make any difference. It's a comfort feature and the plus part just has to do with a gentler transition.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

Carry On
Posts: 5
Joined: Sun Nov 22, 2015 4:37 pm

Re: CFLEX AND CFLEX+

Post by Carry On » Wed Dec 16, 2015 10:55 am

Thanks

I can't tell the difference between cflex and cflex plus. Except with either one sometimes air collects in my cheek. I wonder if that is because of some asynchronousity between my breathing pattern and the machine's. Is that an unavoidable aspect of CFLEX? Would using just CPAP be better. The doctor did not know.

It seems to me that the machine can be too sophisticated that the patient does not know if it is operating properly.

I have been trying to see a difference in O2 with a WM 50 or at least confirm that the machine is working. O2 is above 92 all night.

Is the API a good measure?

What would the Sleepyhead software tell me if I could get it to work on the SIM from this machine.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: DreamWear Nasal CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: CPAP and CFLEX+ @12