PR System one 560 heated hose

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Hawthorne
Posts: 3973
Joined: Tue Oct 19, 2004 4:46 am
Location: London Ontario -Canada

PR System one 560 heated hose

Post by Hawthorne » Wed Nov 25, 2015 5:51 am

I have almost always used passive humidity (water in chamber - no heat). Since I got the 560 System One, I have been using some heat with my humidifier (1 or 2). I am thinking of getting the heated hose converter kit, since it is winter here. For those of you with the heated hose for this machine, what do you like about it over regular heated humidity? What benefit is it to you?

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments:  Backups- FX Nano masks. Backup machine- Airmini auto travel cpap

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: PR System one 560 heated hose

Post by JDS74 » Wed Nov 25, 2015 6:55 am

With the regular heated humidifier, there is rainout in the plain hose depending on the humidity level and room temperature.

With the heated hose, the rainout is almost eliminated under the same conditions. For most, it is completely eliminated.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
Pugsy
Posts: 63942
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: PR System one 560 heated hose

Post by Pugsy » Wed Nov 25, 2015 7:58 am

What I like most is I can get a constant humidity and have the hose air temp at separate controls without messing with the humidity delivered unless I want to.
A person can have as little (or none even) or max added humidity and can change the air temp in the hose.

What feels good in the summer for me in terms of air temp doesn't feel good in the winter because the bedroom is lot colder in the winter. But I don't want to vary my humidity levels because my nose is used to it a certain way and when I change it then I have problems.

Of course especially useful for rain out control but it would be useful for me just from hose air temp control even if I didn't prefer max humidity.
It would be useful for someone who preferred minimal added moisture whose bedroom was colder in the winter and they simply didn't care for the cold air. It all depends on how warm they keep their house.
I would roast in my mom's house at night because she keeps her house a lot warmer than I do my house.
I turn the heat way down at night. Saves fuel and I like to snuggle under the covers but my nose doesn't like that cold air going up my nose. I had lots of wake ups due to nose hurting from the cold even at maximum humidity setting with a regular hose and a hose cozy didn't really help and the weight was annoying pulling on my mask.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
Hawthorne
Posts: 3973
Joined: Tue Oct 19, 2004 4:46 am
Location: London Ontario -Canada

Re: PR System one 560 heated hose

Post by Hawthorne » Wed Nov 25, 2015 1:01 pm

Thanks for the responses, JDS74 and Pugsy!
Pugsy - does this mean I can still use just passive humidity and leave the heated hose attached?

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments:  Backups- FX Nano masks. Backup machine- Airmini auto travel cpap

User avatar
Pugsy
Posts: 63942
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: PR System one 560 heated hose

Post by Pugsy » Wed Nov 25, 2015 1:34 pm

I am pretty sure you can just use passive humidity if that is what you wish and still be able to adjust the hose air temp anyway you want...it starts with 4 degrees about room temp at the setting of 1 for temp and goes up from there.
See product description for more detail on temps
https://www.cpap.com/productpage/pr-sys ... ifier.html
When the heated hose is attached you have to go into the clinical setup menu to set the humidity and normally the settings are 1 to 3.
Give me just a little bit to dig out my 760 machine and see if it will let me set it to off instead of 1, 2 or 3 for humidity. I think it will but let me confirm to be sure.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
Pugsy
Posts: 63942
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: PR System one 560 heated hose

Post by Pugsy » Wed Nov 25, 2015 2:27 pm

I was wrong. With the heated hose attached I don't get a choice for "Off" for the humidity level.
All I get are the 3 choices with 1 being 70%
I can do it with the S9 but not with the 760 machine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
Hawthorne
Posts: 3973
Joined: Tue Oct 19, 2004 4:46 am
Location: London Ontario -Canada

Re: PR System one 560 heated hose

Post by Hawthorne » Wed Nov 25, 2015 2:34 pm

Thanks again, for checking that out Pugsy! I will change hoses if I go to passive again.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments:  Backups- FX Nano masks. Backup machine- Airmini auto travel cpap

User avatar
Uncle_Bob
Posts: 2777
Joined: Tue Feb 24, 2009 12:10 pm
Location: Arizona

Re: PR System one 560 heated hose

Post by Uncle_Bob » Wed Nov 25, 2015 3:19 pm

Pugsy wrote:I was wrong


library lady
Posts: 1116
Joined: Wed Feb 05, 2014 11:10 am
Location: Rochester, MN

Re: PR System one 560 heated hose

Post by library lady » Wed Nov 25, 2015 6:40 pm

I have the PR 560 with heated hose; when I use it with the heat off (on the dial, not in the menu) I don't use as much water as when I have the heater turned up, even at level 1. So, here in Minnesota winters, I often run through a full tank, whereas in the summer I go through very little water, filling the tank about a third of the way. Another advantage is that in winter with the house heat turned down at night, I feel the warmth in the hose. I have only the heated hose, use it year round.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments:  Sleepyhead
Now using AirFit F10 mask; Quattro Air is backup mask. RemZzzz mask liners with both.

User avatar
Hawthorne
Posts: 3973
Joined: Tue Oct 19, 2004 4:46 am
Location: London Ontario -Canada

Re: PR System one 560 heated hose

Post by Hawthorne » Wed Nov 25, 2015 6:57 pm

--so library lady--- you leave the heated hose on the machine year round. In the summer you set the humidity at 0 on the machine and use less water. Are you then using passive mode (no heat) in the summer and still have the heated hose on the machine? I'm trying to get this straight in my head!
Thanks

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments:  Backups- FX Nano masks. Backup machine- Airmini auto travel cpap

User avatar
bwexler
Posts: 1507
Joined: Mon Nov 08, 2010 11:52 pm
Location: San Marcos, Ca. USA

Re: PR System one 560 heated hose

Post by bwexler » Wed Nov 25, 2015 11:22 pm

I set the humidity and the hose heat when I got my PRS1 960 two years ago.
I haven't fussed with it since.
I have used a heated hose year round for 5 years now.

_________________
Mask: SleepWeaver 3D Soft Cloth Nasal CPAP Mask with Headgear
Additional Comments: AurCurve 10 ASV Also using Sleaplyhead 1.1, ResScan 6 and CMS50i

User avatar
M'ohms
Posts: 516
Joined: Fri Aug 09, 2013 2:44 pm
Location: Georgetown, Indiana

Re: PR System one 560 heated hose

Post by M'ohms » Thu Nov 26, 2015 7:16 pm

Year 'round for me too, and like Pugsy, I leave my humidity level on 3 and fiddle with the temp. only during the different seasons.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Supplemental oxygen at 2.5 lpm
AutoSV 960 with heated hose. Settings: EPAP Min-12, EPAP Max-17, PS Min-5.5, PS Max-13, Max Pressure-25, Rate-Auto, Rise Time 1. Use Sleepyhead and Encore Pro.

yaconsult
Posts: 1099
Joined: Fri Nov 14, 2014 10:43 pm
Location: "Silicon Valley", CA

Re: PR System one 560 heated hose

Post by yaconsult » Thu Nov 26, 2015 10:44 pm

On my AirSense 10, I have just left it set to auto as it seems to work well for me that way. I usually use between half and three-quarters of water in the tank each night.

BTW, the hose doesn't make the air you breathe feel warmer, just more moist.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: January 2015 Sleep Study Results: Apnea/Hypopnea Index (AHI): 80.2, Sleepyhead

User avatar
Hawthorne
Posts: 3973
Joined: Tue Oct 19, 2004 4:46 am
Location: London Ontario -Canada

Re: PR System one 560 heated hose

Post by Hawthorne » Fri Nov 27, 2015 7:29 am

Thanks all - for your responses! Very helpful! I am going to get the kit and convert my 560 to the heated hose!

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments:  Backups- FX Nano masks. Backup machine- Airmini auto travel cpap