Weird power consumption in "Off" mode
Weird power consumption in "Off" mode
I just bought a AirSense 10 (used). I put my watt meter on it to see how it compared to my old one. The old one had a Purge mode after it was turned off. So I unplugged it every morning. Why waste electricity to run it when I am not using it. Since I wasn't using humidity I didn't need the purge.
I notice strange power readings on the new AirSense 10.
When I plug it in there is a quick spike and then it settles to about 5 watts for 3 minutes or so till the display turns off.
Then it goes up a lot to about 25-35 watts. It then cycles back and forth between about 25 watts to 35 watts. It keeps doing this.
Then I turn it on and it uses 25 watts wide open and 10 watts at breathing pressure. This is what I expect.
Turn it off and it goes to 5 watts while the display is on for about 3 minutes.
When the display goes off it goes back to its 25-35 watt cycle back and forth. This makes no sense. I am not using it so it should be about zero.
So for this one I will have to make a switch to turn it off during the day. I shouldn't have to do this.
I notice strange power readings on the new AirSense 10.
When I plug it in there is a quick spike and then it settles to about 5 watts for 3 minutes or so till the display turns off.
Then it goes up a lot to about 25-35 watts. It then cycles back and forth between about 25 watts to 35 watts. It keeps doing this.
Then I turn it on and it uses 25 watts wide open and 10 watts at breathing pressure. This is what I expect.
Turn it off and it goes to 5 watts while the display is on for about 3 minutes.
When the display goes off it goes back to its 25-35 watt cycle back and forth. This makes no sense. I am not using it so it should be about zero.
So for this one I will have to make a switch to turn it off during the day. I shouldn't have to do this.
Re: Weird power consumption in "Off" mode
Many devices draw current when turned off, it called standby, after all they aren't paying your electric bill. Then there's wall warts, that power the world, with my external drives and chargers I'll bet my house have 10 to 15 running. Then count t.v.'s I plugged mine in last week it sparked like I plugged a 1/2 horsepower motor, energy star my butt.
Either pay the bills or buy a dozen power strips, and box of bandaid to pad your worn out fingers from pushing the switches. Jim
Either pay the bills or buy a dozen power strips, and box of bandaid to pad your worn out fingers from pushing the switches. Jim
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: Weird power consumption in "Off" mode
If you don't already have it plugged into one, I'd suggest a surge protector outlet bar with on/off switch.Leea wrote:I just bought a AirSense 10 (used). I put my watt meter on it to see how it compared to my old one. The old one had a Purge mode after it was turned off. So I unplugged it every morning. Why waste electricity to run it when I am not using it. Since I wasn't using humidity I didn't need the purge.
I notice strange power readings on the new AirSense 10.
When I plug it in there is a quick spike and then it settles to about 5 watts for 3 minutes or so till the display turns off.
Then it goes up a lot to about 25-35 watts. It then cycles back and forth between about 25 watts to 35 watts. It keeps doing this.
Then I turn it on and it uses 25 watts wide open and 10 watts at breathing pressure. This is what I expect.
Turn it off and it goes to 5 watts while the display is on for about 3 minutes.
When the display goes off it goes back to its 25-35 watt cycle back and forth. This makes no sense. I am not using it so it should be about zero.
So for this one I will have to make a switch to turn it off during the day. I shouldn't have to do this.
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: Weird power consumption in "Off" mode
inrush current that charges the caps in a switch mode power supply rally has nothing at all to do with steady state power consumption.Goofproof wrote:Then count t.v.'s I plugged mine in last week it sparked like I plugged a 1/2 horsepower motor, energy star my butt.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- TangledHose
- Posts: 862
- Joined: Wed Mar 13, 2013 11:13 am
- Location: Centennial, CO
Re: Weird power consumption in "Off" mode
Push and hold the power button on top for three seconds, it'll put the machine in a low power sleep mode.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Simplus Full Face CPAP Mask with Headgear |
Additional Comments: Also own F&P Icon Auto for backup and travel; and a Transcend when packing space is limited |
Link to Sleepyhead:
http://sleepfiles.com/SH2/
http://sleepfiles.com/SH2/
Re: Weird power consumption in "Off" mode
Is out timepalerider wrote:inrush current that charges the caps in a switch mode power supply rally has nothing at all to do with steady state power consumption.Goofproof wrote:Then count t.v.'s I plugged mine in last week it sparked like I plugged a 1/2 horsepower motor, energy star my butt.
I'm well aware of how electronics work, just saying the way products are built today has little regard for power efficiency. Is our time worth so much we can't be bothered to walk over and turn on our toys..... Off no longer means off, in English or Spanish. Jim
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
- chunkyfrog
- Posts: 34378
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Weird power consumption in "Off" mode
Could capacitor drain and recharge make it more economical to leave some things plugged in?
I understand that some lights die sooner if you shut them off and turn them on repeatedly.
The cost of replacing the device may outweigh the energy expenditure.
{{{I will grasp at any excuse to have lights on, as dark rooms are depressing.}}}
Anyhow, my APAP resets the display every time I unplug it, so, not doing it every day.
I understand that some lights die sooner if you shut them off and turn them on repeatedly.
The cost of replacing the device may outweigh the energy expenditure.
{{{I will grasp at any excuse to have lights on, as dark rooms are depressing.}}}
Anyhow, my APAP resets the display every time I unplug it, so, not doing it every day.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Weird power consumption in "Off" mode
[quote="chunkyfrog"{{{I will grasp at any excuse to have lights on, as dark rooms are depressing.}}} [/quote]
LEDs don't care, and are one of the most efficient lighting sources currently.
of course, that said, they don't care at all whether you turn them on and off... their power supplies *might* but the leds themselves don't
LEDs don't care, and are one of the most efficient lighting sources currently.
of course, that said, they don't care at all whether you turn them on and off... their power supplies *might* but the leds themselves don't
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Weird power consumption in "Off" mode
From the user guide:TangledHose wrote:Push and hold the power button on top for three seconds, it'll put the machine in a low power sleep mode.
Power save mode
Your AirSense 10 device records your therapy data. In order to allow it to transmit the data to your
care provider, you should not unplug the device. However, you can put it into power save mode to
save electricity.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: PAPCap, 3M Microfoam Surgical Tape, PoliGrip Strip, APAP 12.0 - 14.0 cm |
- chunkyfrog
- Posts: 34378
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Weird power consumption in "Off" mode
Every light in the house is either LED or CFL--it adds up. Been doing that for over thirty years.palerider wrote: LEDs don't care, and are one of the most efficient lighting sources currently. of course, that said,
they don't care at all whether you turn them on and off... their power supplies *might* but the leds themselves don't
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Weird power consumption in "Off" mode
The annoying thing that my apap does when I unplug it and plug it back in for some reason is to change the default period for the "sleep report" to 1 month rather than 1 day. If you don't notice that, you can have some misleading information presented to you *smile*. Is that what you meant by "resets the display"? Just curious.chunkyfrog wrote: Anyhow, my APAP resets the display every time I unplug it, so, not doing it every day.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear |
- chunkyfrog
- Posts: 34378
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Weird power consumption in "Off" mode
Yes, I just hate when it does that.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Weird power consumption in "Off" mode
And the Chinese and Koreans, thank you for their jobs. While Americans are collecting the last of their food stamps.chunkyfrog wrote:Every light in the house is either LED or CFL--it adds up. Been doing that for over thirty years.palerider wrote: LEDs don't care, and are one of the most efficient lighting sources currently. of course, that said,
they don't care at all whether you turn them on and off... their power supplies *might* but the leds themselves don't
CFL's are the worst, short lived, maximum environment polution, led's area much better solution. Jim Next year I'm switching to fireflies for lighting (true cold fusion)...
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire