Maximum hose length

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
wlenz
Posts: 127
Joined: Wed May 28, 2008 8:08 pm
Location: Stockholm, NJ
Contact:

Maximum hose length

Post by wlenz » Mon Aug 03, 2015 3:36 pm

I have been using an 8 ft hose on my auto BiPAP machine. I toss and turn a lot and the extra length reduces pulling the machine off my nightstand. I get 6 ft hoses for free and would like to use them. An in-line coupler would cascade two 6 ft hoses. Since the sensors for auto ipap/epap are in the machine, is there a maximum length that will allow the machine to sense what is happening at the mask? Does some manufacturer sell a 2 ft extension?
There is only one way to achieve happiness on this terrestrial ball, And that is to have either a clear conscience, or none at all.

Ogden Nash

User avatar
Pugsy
Posts: 64029
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Maximum hose length

Post by Pugsy » Mon Aug 03, 2015 3:43 pm

There are people using 2 of the 6 ft hoses without any apparent issues with 12 ft.
Manufacturer's do make a 10 ft hose that is readily available so I would assume that 10 foot isn't a problem.
I can't say for sure about 12 ft though...might make yourself a manometer to check the pressure you are getting to see if it corresponds with what you are supposed to get.

They do make 2 foot hose extensions though.
https://www.cpap.com/cpap-hose/cpap-hose-short.php

Before anyone asks...heated hoses are pretty much 6 ft only...if you want to use a heated hose then you would have to piggy back a non heated extension to the end of the heated hose. You would lose some of the heated hose benefit but it still might be enough to get the job done.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
Sheriff Buford
Posts: 4082
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: Maximum hose length

Post by Sheriff Buford » Mon Aug 03, 2015 4:26 pm

I've been using the 8' hose for years. I need the 2' so I'll be getting the 10'.

Sheriff

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Maximum hose length

Post by Wulfman... » Mon Aug 03, 2015 4:44 pm

wlenz wrote:I have been using an 8 ft hose on my auto BiPAP machine. I toss and turn a lot and the extra length reduces pulling the machine off my nightstand. I get 6 ft hoses for free and would like to use them. An in-line coupler would cascade two 6 ft hoses. Since the sensors for auto ipap/epap are in the machine, is there a maximum length that will allow the machine to sense what is happening at the mask? Does some manufacturer sell a 2 ft extension?
Over the years, there have been discussions about this subject. I remember somebody saying that they even connected two 10' hoses with (virtually) no pressure loss with fixed pressure. However, the cautionary points were about when using ranges of pressures and the possible delay in the machine sensors being able to function properly. Straight pressure presents virtually no problems, but with pressure ranges there may be some, depending how long the distance is.

Found a couple of those old threads:
I believe the later one also references the earlier ones.
The forum member "dsm" did a variety of tests, too.

viewtopic.php?f=1&t=21043&

viewtopic.php?f=1&t=37508&


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
TangledHose
Posts: 862
Joined: Wed Mar 13, 2013 11:13 am
Location: Centennial, CO

Re: Maximum hose length

Post by TangledHose » Mon Aug 03, 2015 5:26 pm

I am not going to mention my maximum hose length here.......it's just too personal


_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: Also own F&P Icon Auto for backup and travel; and a Transcend when packing space is limited
Link to Sleepyhead:
http://sleepfiles.com/SH2/

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Maximum hose length

Post by palerider » Mon Aug 03, 2015 5:57 pm

wlenz wrote:I have been using an 8 ft hose on my auto BiPAP machine. I toss and turn a lot and the extra length reduces pulling the machine off my nightstand. I get 6 ft hoses for free and would like to use them. An in-line coupler would cascade two 6 ft hoses. Since the sensors for auto ipap/epap are in the machine, is there a maximum length that will allow the machine to sense what is happening at the mask? Does some manufacturer sell a 2 ft extension?
the resmed bilevel machines have a setting for a 10 foot hose. (which I use on mine) this allows better pressure delivery, since the resistance of the hose will have some affect on the pressures at the mask....

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: Maximum hose length

Post by HoseCrusher » Mon Aug 03, 2015 6:57 pm

Just as a point of reference...

I ran this question by a process engineer. Assuming 50 feet of 22 mm diameter smooth bore hose and a pressure of 20 cm H2O at a constant flow of 50 liters per minute the "rough" calculation indicates that you would have a little less than 0.05 cm H2O difference at the output.

That means that if you set the pressure on your machine to 20 cm H2O and use a 50 foot hose (22 mm diameter) you would only end up with about 19.95 cm H2O at the mask.

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Maximum hose length

Post by Goofproof » Mon Aug 03, 2015 7:22 pm

The manufacturers set a 10 ft maximum hose length for a reason, adding more than that degrades the machines ability to detect the need for pressure changes, and the collecting of reliable data. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Maximum hose length

Post by Wulfman... » Mon Aug 03, 2015 7:26 pm

HoseCrusher wrote:Just as a point of reference...

I ran this question by a process engineer. Assuming 50 feet of 22 mm diameter smooth bore hose and a pressure of 20 cm H2O at a constant flow of 50 liters per minute the "rough" calculation indicates that you would have a little less than 0.05 cm H2O difference at the output.

That means that if you set the pressure on your machine to 20 cm H2O and use a 50 foot hose (22 mm diameter) you would only end up with about 19.95 cm H2O at the mask.
And, if you look in the machine manuals, the pressure specifications typically show an accuracy of +/- 0.5 to +/- 1.0 depending on the pressure settings.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Maximum hose length

Post by palerider » Mon Aug 03, 2015 7:26 pm

HoseCrusher wrote:Just as a point of reference...

I ran this question by a process engineer. Assuming 50 feet of 22 mm diameter smooth bore hose and a pressure of 20 cm H2O at a constant flow of 50 liters per minute the "rough" calculation indicates that you would have a little less than 0.05 cm H2O difference at the output.

That means that if you set the pressure on your machine to 20 cm H2O and use a 50 foot hose (22 mm diameter) you would only end up with about 19.95 cm H2O at the mask.
the tricky bits, however, come when you've got varying flow, as you do when you're breathing, that causes more or less pressure drop over the length of the hose.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: Maximum hose length

Post by JDS74 » Mon Aug 03, 2015 7:31 pm

wlenz wrote:I have been using an 8 ft hose on my auto BiPAP machine. I toss and turn a lot and the extra length reduces pulling the machine off my nightstand. I get 6 ft hoses for free and would like to use them. An in-line coupler would cascade two 6 ft hoses. Since the sensors for auto ipap/epap are in the machine, is there a maximum length that will allow the machine to sense what is happening at the mask? Does some manufacturer sell a 2 ft extension?
Have you tried a hose hanger to isolate your movement from the machine/
I use a hose hanger and also tie my hose to my four poster bed and cannot pull the machine at all.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
wlenz
Posts: 127
Joined: Wed May 28, 2008 8:08 pm
Location: Stockholm, NJ
Contact:

Re: Maximum hose length

Post by wlenz » Mon Aug 03, 2015 8:21 pm

Thanks for all the great replies. I will only use 8' hose, it is anchored to my bed with a velcro strap, and I will monitor the data frequently.
Again, thanks
There is only one way to achieve happiness on this terrestrial ball, And that is to have either a clear conscience, or none at all.

Ogden Nash

User avatar
Zomby Woof
Posts: 122
Joined: Tue Apr 07, 2015 2:28 pm
Location: Dallas TX

Re: Maximum hose length

Post by Zomby Woof » Tue Aug 04, 2015 9:05 am

TangledHose wrote:I am not going to mention my maximum hose length here.......it's just too personal
... feeling inadequate 'cause I've just got an "average" hose

Could I be suffering from "hose envy"?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Oracle HC452 Oral CPAP Mask
Additional Comments: EPAP 12.6 - 25, PS 4, 0min Ramp, Humidity Setting 4
You know I'll never sleep no more, It seem to me that it just ain't wise
Didja ever wake up in the mornin', With a ZOMBY WOOF behind your eyes?
Just about as evil as you could be.
Tellin' you all the Zomby troof, Here I'm is, the ZOMBY WOOF
- Zappa