AirFit F10 ventilation question - first night

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
mptolman
Posts: 3
Joined: Fri Jul 31, 2015 11:43 am

AirFit F10 ventilation question - first night

Post by mptolman » Fri Jul 31, 2015 11:57 am

In the corner of the elbow of the mask are what look like two large ventilation holes. When the air is blowing, a lot of air is coming out of those holes and it feels like I'm not getting as much as I should in the mask. Is this normal behavior? Or should something be blocking that vent unless the air is turned off? Forgive my ignorance if this is a silly question--last night was my first experience with CPAP.

User avatar
fdw
Posts: 260
Joined: Wed Aug 27, 2014 7:08 am
Location: D/FW TEXAS

Re: AirFit F10 ventilation question - first night

Post by fdw » Fri Jul 31, 2015 12:19 pm

mptolman wrote:In the corner of the elbow of the mask are what look like two large ventilation holes. When the air is blowing, a lot of air is coming out of those holes and it feels like I'm not getting as much as I should in the mask. Is this normal behavior? Or should something be blocking that vent unless the air is turned off? Forgive my ignorance if this is a silly question--last night was my first experience with CPAP.

I've used the F10 for a little more than 6 months now. IMO what your experiencing is normal with the amount of air coming out of the vent, do NOT block the vent or try and plug it. As far as you feeling like your not getting enough air, Im currently exploring the same idea. For example my last titration was with a Nasal mask with a pressure of 13.0, however currently Im set at 14.2 and thinking about another 0.4. bump in pressure. The main reason for my bump in pressure is because of the amount of time spent in hypopnea apena.
ResMed AirSense S10 Autoset (Ins. Pur. 3/16/2016)
AutoCPAP 10-15 EPR 1.0 (Latest Sleep Study 3-7-16)
Started CPAP March 1995 with a pressure of 11.0

User avatar
Pugsy
Posts: 64006
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AirFit F10 ventilation question - first night

Post by Pugsy » Fri Jul 31, 2015 12:42 pm

I thought the vent holes were the little diffused holes around where the elbow fits into the mask.
The 2 bigger holes appear to be on the elbow itself.
I think those larger holes are the anti Asphyxia valve...see pages 11 and the last page that shows the diagram of all the pieces.
http://www.resmed.com/us/dam/documents/ ... er_eng.pdf
No air should be coming out of them when the mask is under pressure.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
fdw
Posts: 260
Joined: Wed Aug 27, 2014 7:08 am
Location: D/FW TEXAS

Re: AirFit F10 ventilation question - first night

Post by fdw » Fri Jul 31, 2015 12:56 pm

Pugsy wrote:I thought the vent holes were the little diffused holes around where the elbow fits into the mask.
The 2 bigger holes appear to be on the elbow itself.
I think those larger holes are the anti Asphyxia valve...see pages 11 and the last page that shows the diagram of all the pieces.
http://www.resmed.com/us/dam/documents/ ... er_eng.pdf
No air should be coming out of them when the mask is under pressure.
Are saying that NO air should be blowing out of the Elbow Vent when under normal operating (treatment) pressure?????

Because it blow out of mine like crazy when under pressure.....I assumed that was normal.
ResMed AirSense S10 Autoset (Ins. Pur. 3/16/2016)
AutoCPAP 10-15 EPR 1.0 (Latest Sleep Study 3-7-16)
Started CPAP March 1995 with a pressure of 11.0

User avatar
Pugsy
Posts: 64006
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AirFit F10 ventilation question - first night

Post by Pugsy » Fri Jul 31, 2015 1:04 pm

fdw wrote:Are saying that NO air should be blowing out of the Elbow Vent when under normal operating (treatment) pressure?????

Because it blow out of mine like crazy when under pressure.....I assumed that was normal.
Are you sure that what you are feeling is coming from the anti asphyxia valve and not the vent holes around the elbow? It's in the same area.
My understanding of the valve is that when under pressure it should be closed. It's purpose is to open up and allow air exchange so that we don't re breathe too much of our own exhaled air should power loss occur and we don't wake up.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

mptolman
Posts: 3
Joined: Fri Jul 31, 2015 11:43 am

Re: AirFit F10 ventilation question - first night

Post by mptolman » Fri Jul 31, 2015 1:11 pm

Yes, there is a constant stream of air blowing out of the elbow itself when pressure is on. I'm wondering if this is normal. The machine isn't reporting any large leaks, so maybe it is normal? The hose is pointed directly at the holes in the elbow and there's nothing blocking the holes, so some air is going through the holes and the rest is traveling to the mask.

User avatar
fdw
Posts: 260
Joined: Wed Aug 27, 2014 7:08 am
Location: D/FW TEXAS

Re: AirFit F10 ventilation question - first night

Post by fdw » Fri Jul 31, 2015 1:28 pm

Pugsy wrote:
fdw wrote:Are saying that NO air should be blowing out of the Elbow Vent when under normal operating (treatment) pressure?????

Because it blow out of mine like crazy when under pressure.....I assumed that was normal.
Are you sure that what you are feeling is coming from the anti asphyxia valve and not the vent holes around the elbow? It's in the same area.
My understanding of the valve is that when under pressure it should be closed. It's purpose is to open up and allow air exchange so that we don't re breathe too much of our own exhaled air should power loss occur and we don't wake up.
100% positive.......maybe that's why my pressure is much higher than titration levels.
ResMed AirSense S10 Autoset (Ins. Pur. 3/16/2016)
AutoCPAP 10-15 EPR 1.0 (Latest Sleep Study 3-7-16)
Started CPAP March 1995 with a pressure of 11.0

User avatar
Pugsy
Posts: 64006
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AirFit F10 ventilation question - first night

Post by Pugsy » Fri Jul 31, 2015 2:21 pm

The vent holes are right next to the elbow. I watched the video for cleaning and that's where they suggest using a toothbrush to clean out the tiny holes if they are blocked. The vent holes are tiny but they do encircle the elbow.

I am just going by the diagram I linked to above from the ResMed website...those 2 holes are for the AA valve and are not the vent holes.

Now I always thought that the anti asphyxia valve was or should be closed when pressure was within the circuit (machine blowing air).

Higher pressures would cause more air out the vent holes though...so maybe that is what is being felt at the elbow.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
Midnight Strangler
Posts: 248
Joined: Tue Oct 07, 2014 5:39 am

Re: AirFit F10 ventilation question - first night

Post by Midnight Strangler » Fri Jul 31, 2015 2:41 pm

The exhaust port is a circle of tiny openings in the front of the mask. There should be air blowing out through the openings at all times the system is operating.

Image

_______________________________________________________________________________________
_______________________________________________________________________________________

The anti-asphyxia valves are on the upper sides of the elbow. They should be closed with no air coming out or in when the system is working. When the system is not working, they should open to allow the air of breathing to move in and out.

If either of these valves is open while the CPAP process is running, your machine should register a large leak and your therapy may not be effective.

Image

mptolman
Posts: 3
Joined: Fri Jul 31, 2015 11:43 am

Re: AirFit F10 ventilation question - first night

Post by mptolman » Fri Jul 31, 2015 2:49 pm

Yeah, I'm getting a stream of air coming out of the anti-asphyxia valve within the elbow itself while the system is running. There is no barrier between the valve openings and the air coming from the hose, so a lot of the air is blowing straight up and out the valve. It sounds like this shouldn't be happening, so I'll visit my provider and see about getting a replacement.

User avatar
ChicagoGranny
Posts: 14454
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: AirFit F10 ventilation question - first night

Post by ChicagoGranny » Fri Jul 31, 2015 3:07 pm

mptolman wrote:There is no barrier between the valve openings and the air coming from the hose
I've never had that mask to examine. How is the AA valve engineered to open and close?

The AA valve on my mask has a flap that opens when the circuit is not operating. When the circuit is operating, the pressure inside the elbow forces the flap closed. When I turn my machine on, the flap can be heard making a tiny "slam-close" noise.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AirFit F10 ventilation question - first night

Post by palerider » Fri Jul 31, 2015 3:08 pm

mptolman and fdw, I'm pretty sure you're both wrong about the anti-asphyxiation holes, there's a little flap that should seal those when the blower is on, and then air comes out from the holes around where the elbow fits into the mask proper.

get a mirror and examine those holes *closely* while the machine is running, or, pop the elbow off the mask and put your thumb over the main opening, then turn on the cpap, the flap should close and no air should come out of it.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Last edited by palerider on Fri Jul 31, 2015 5:30 pm, edited 1 time in total.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
ChicagoGranny
Posts: 14454
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: AirFit F10 ventilation question - first night

Post by ChicagoGranny » Fri Jul 31, 2015 5:27 pm

palerider wrote:I'm pretty sure you're both wrong about the anti-asphyxiation holes
What do you mean? Are you saying Strangler has the descriptions reversed? I don't think so.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AirFit F10 ventilation question - first night

Post by palerider » Fri Jul 31, 2015 5:29 pm

ChicagoGranny wrote:
palerider wrote:I'm pretty sure you're both wrong about the anti-asphyxiation holes
What do you mean? Are you saying Strangler has the descriptions reversed? I don't think so.
oh, no, I'll go back and clarify, I was referring to mptolman and fdw, strangler is 100% correct.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AirFit F10 ventilation question - first night

Post by palerider » Fri Jul 31, 2015 5:32 pm

ChicagoGranny wrote:
mptolman wrote:There is no barrier between the valve openings and the air coming from the hose
I've never had that mask to examine. How is the AA valve engineered to open and close?

The AA valve on my mask has a flap that opens when the circuit is not operating. When the circuit is operating, the pressure inside the elbow forces the flap closed. When I turn my machine on, the flap can be heard making a tiny "slam-close" noise.
yeah, unless someone has ripped off the flapper with over exuberant cleaning, it'll close and block off airflow from the AA holes.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.