Increased noise on inhalation.

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Jackiep
Posts: 49
Joined: Thu Jun 04, 2015 4:10 pm
Location: Alberta, Canada

Increased noise on inhalation.

Post by Jackiep » Thu Jul 02, 2015 9:02 am

I have been testing this machine for almost two months and am now ready to purchase. Lately I have become more aware of increased noise in the machine on inhalation...at least I think the noise is coming from the machine ...it is not a leak type noise. Last night it actually kept me awake...of course the more aware I am of it the more it bothers me. Is it normal to have this increased noise on inhalation ? Thanks for any input.

User avatar
Julie
Posts: 19910
Joined: Tue Feb 28, 2006 12:58 pm

Re: Increased noise on inhalation.

Post by Julie » Thu Jul 02, 2015 9:10 am

Where is the noise coming from - front, back, hose, mask, humidifier, etc. etc?

User avatar
Jackiep
Posts: 49
Joined: Thu Jun 04, 2015 4:10 pm
Location: Alberta, Canada

Re: Increased noise on inhalation.

Post by Jackiep » Thu Jul 02, 2015 9:26 am

It is not coming from the hose or mask....seems to be in the machine ... can't determine exactly where. Of course the machine makes a constant purr when in use (not loud)....the purr just increases substantially on inhalation.

User avatar
Julie
Posts: 19910
Joined: Tue Feb 28, 2006 12:58 pm

Re: Increased noise on inhalation.

Post by Julie » Thu Jul 02, 2015 9:29 am

Have you talked to the DME or brought the machine in?

User avatar
Jackiep
Posts: 49
Joined: Thu Jun 04, 2015 4:10 pm
Location: Alberta, Canada

Re: Increased noise on inhalation.

Post by Jackiep » Thu Jul 02, 2015 9:34 am

I will call DME today.......you guys are my first "go to " people....but you're right.. I should contact them.Thanks

WildMtnHoney
Posts: 140
Joined: Sat Feb 28, 2015 10:38 am

Re: Increased noise on inhalation.

Post by WildMtnHoney » Thu Jul 02, 2015 9:45 am

This may not be it, but - my totally different machine, with differently named settings, does something similar. The setting that is my expiratory relief means that the fan drive noise goes down a little bit because the fan slows down a little bit (lower pressure) during each exhalation. I believe that your machine's Aflex works similarly. So, to me, it's not that it is "more" noise on inhalation, it's that there is "less" noise on exhalation - does that make sense?

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleephead (whatever version was current as of 3/15)

User avatar
Jackiep
Posts: 49
Joined: Thu Jun 04, 2015 4:10 pm
Location: Alberta, Canada

Re: Increased noise on inhalation.

Post by Jackiep » Thu Jul 02, 2015 9:51 am

That certainly does make sense...I have a call in to the technician now ...we will see if he concurs. Thanks

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Increased noise on inhalation.

Post by palerider » Thu Jul 02, 2015 10:37 am

the fans on the machines do change speed breath by breath, they're trying to maintain a desired pressure, either a slightly changing one with epr or *flex, or a constant one with those off. when you inhale and exhale, the fan has to change speed to maintain the pressure against the changing airflow, so even with a fixed pressure, the fan will speed up when you inhale, and slow down as you exhale, thus maintaining pressure.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Jackiep
Posts: 49
Joined: Thu Jun 04, 2015 4:10 pm
Location: Alberta, Canada

Re: Increased noise on inhalation.

Post by Jackiep » Thu Jul 02, 2015 7:13 pm

Thanks for the great explanation Palerider ! Sounds like this is the way it is and I am just going to have to get used to it... I'll try !

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Increased noise on inhalation.

Post by palerider » Thu Jul 02, 2015 7:22 pm

Jackiep wrote:Thanks for the great explanation Palerider ! Sounds like this is the way it is and I am just going to have to get used to it... I'll try !
most folks get used to the noise after a bit... then it becomes an old friend, and when it changes in some way, you gotta get used to that...

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.