Centrals Predominate - 2 1/2 Month Check In

Post by Zomby Woof » Tue Jun 30, 2015 9:43 am

My sleep therapy results are much better thanks to this forum's advice.
Now at 75 days into treatment I'm checking in here as my progress seems to have stalled.

My settings have remained constant for a month now - Min EPAP 10cwp, Max IPAP 18cwp, Min PS 3cwp.
While OAs have slowly trended downward, CAs have stayed steady or actually increased.

Your comments about which direction I might take next would be greatly appreciated!


Below are representative SH screenshots from this week:

Image

Image

Image

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Oracle HC452 Oral CPAP Mask
Additional Comments: EPAP 12.6 - 25, PS 4, 0min Ramp, Humidity Setting 4
You know I'll never sleep no more, It seem to me that it just ain't wise
Didja ever wake up in the mornin', With a ZOMBY WOOF behind your eyes?
Just about as evil as you could be.
Tellin' you all the Zomby troof, Here I'm is, the ZOMBY WOOF
- Zappa

User avatar
Julie
Posts: 19908
Joined: Tue Feb 28, 2006 12:58 pm

Re: Centrals Predominate - 2 1/2 Month Check In

Post by Julie » Tue Jun 30, 2015 9:48 am

I personally would try to do something about those leaks - they may not be officially too high as far as the machine goes, but they are so constant, and in the interests of curiosity if nothing else, I'd experiment with taping for at least a couple of nights, or a FF mask if you have one that fits relatively well, just to see if they help. I'm always being lectured on how leaks are not a problem because of what the machine says, but if mine were as constant as yours (unless the read outs really reflect centrals and not air leaks) I'd want to see about improving them.

User avatar
Zomby Woof
Posts: 122
Joined: Tue Apr 07, 2015 2:28 pm
Location: Dallas TX

Re: Centrals Predominate - 2 1/2 Month Check In

Post by Zomby Woof » Tue Jun 30, 2015 10:05 am

Julie wrote:I personally would try to do something about those leaks - ,
I'd experiment with taping for at least a couple of nights,
or a FF mask if you have one that fits relatively well, . I'd want to see about improving them.
Thank you for your comments Julie.
I have done quite a bit of experimenting with taping; I use bandage tape
ear-to-ear every night and have added a cervical collar at times.

FF masks I've tried and ruled out due to their harnesses causing headaches, and,
their leak rates which were much higher than you see in these screenshots.

My thinking is that the leaks are caused by pillow contact or facial distortion when side-sleeping.
Any suggestions for a replacement pillow?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Oracle HC452 Oral CPAP Mask
Additional Comments: EPAP 12.6 - 25, PS 4, 0min Ramp, Humidity Setting 4
You know I'll never sleep no more, It seem to me that it just ain't wise
Didja ever wake up in the mornin', With a ZOMBY WOOF behind your eyes?
Just about as evil as you could be.
Tellin' you all the Zomby troof, Here I'm is, the ZOMBY WOOF
- Zappa

User avatar
Krelvin
Posts: 1977
Joined: Tue Jun 06, 2006 5:23 pm
Location: Metro Phx Area - Dry Heat!

Re: Centrals Predominate - 2 1/2 Month Check In

Post by Krelvin » Tue Jun 30, 2015 10:42 am

deforeman wrote:My sleep therapy results are much better thanks to this forum's advice.
Now at 75 days into treatment I'm checking in here as my progress seems to have stalled.
Looks to me that most of your CA's are centered around Large Leaks and when you are going to sleep or waking up.

CA's while you are not sleeping are not really something you should need to care about.

Controlling the large leaks will settle you down and most non-wake CA's most likely will settle down.
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin

User avatar
Zomby Woof
Posts: 122
Joined: Tue Apr 07, 2015 2:28 pm
Location: Dallas TX

Re: Centrals Predominate - 2 1/2 Month Check In

Post by Zomby Woof » Tue Jun 30, 2015 11:12 am

Krelvin wrote: Looks to me that most of your CA's are centered around Large Leaks and when you are going to sleep or waking up.
CA's while you are not sleeping are not really something you should need to care about.
Controlling the large leaks will settle you down and most non-wake CA's most likely will settle down.
Thanks Krelvin!

BTW, I'm shopping for a cpap pillow. The cpap.com site has 9 varieties available.

Do you use a specialty pillow? Have any suggestions?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Oracle HC452 Oral CPAP Mask
Additional Comments: EPAP 12.6 - 25, PS 4, 0min Ramp, Humidity Setting 4
You know I'll never sleep no more, It seem to me that it just ain't wise
Didja ever wake up in the mornin', With a ZOMBY WOOF behind your eyes?
Just about as evil as you could be.
Tellin' you all the Zomby troof, Here I'm is, the ZOMBY WOOF
- Zappa