nearly new ASV settings

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
StephenR
Posts: 172
Joined: Fri Jul 25, 2014 4:43 pm

nearly new ASV settings

Post by StephenR » Wed May 13, 2015 3:38 pm

Picked up my nearly new ASV machine today. Just setting it up. On CPAP I have been on a pressure of 8cm, it takes care of my OA's.

I am going with : EPAP 6, Min PS 5, Max PS 10. Will try it for a hour and see what gives...

Stephen

User avatar
adipasqu
Posts: 128
Joined: Thu Apr 21, 2011 11:24 am
Location: San Francisco Bay Area, CA

Re: nearly new ASV settings

Post by adipasqu » Wed May 13, 2015 4:05 pm

If your CPAP pressure has been at 8 cm and it has been taking care of your OA's, you are likely going to want to set your EPAP at 8, as well. The 36007 does not have the auto-EPAP, so it will not increase in pressure if you start having OA's. PSmin of 5 might be too uncomfortable and unnecessary. I would start with PSmin = 2 and let the machine figure out what PS is best. Also, set PSmax as high as it will go so you don't limit the ASV function of the Adapt.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Now using a ResMed S9 VPAP Adapt (36037) with H5i...SleepyHead software
I am a doctor...just not that kind of doctor.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: nearly new ASV settings

Post by palerider » Wed May 13, 2015 5:43 pm

StephenR wrote:Picked up my nearly new ASV machine today. Just setting it up. On CPAP I have been on a pressure of 8cm, it takes care of my OA's.

I am going with : EPAP 6, Min PS 5, Max PS 10. Will try it for a hour and see what gives...

Stephen
I second the notion of leaving the max ps high, it defaults to 15. and it's the high possible PS that gives the adapt the ability to respond effectively to your centrals.

the min ps might be alot more than you're used to, but play with it and see how it feels, you might like it. I'd also agree with perhaps a higher epap, since the epap is what prevents the OAs.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

StephenR
Posts: 172
Joined: Fri Jul 25, 2014 4:43 pm

Re: nearly new ASV settings

Post by StephenR » Wed May 13, 2015 5:48 pm

palerider wrote:
StephenR wrote:Picked up my nearly new ASV machine today. Just setting it up. On CPAP I have been on a pressure of 8cm, it takes care of my OA's.

I am going with : EPAP 6, Min PS 5, Max PS 10. Will try it for a hour and see what gives...

Stephen
I second the notion of leaving the max ps high, it defaults to 15. and it's the high possible PS that gives the adapt the ability to respond effectively to your centrals.

the min ps might be alot more than you're used to, but play with it and see how it feels, you might like it. I'd also agree with perhaps a higher epap, since the epap is what prevents the OAs.
Oh, OK, I will set the EPAP to 8, I was just being a wus...
Have to say the hour I spent breathing with the machine was wonderful, I have a feeling this is the machine I "should" have been on in a perfect world..
Thanks guys...

Stephen

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: nearly new ASV settings

Post by palerider » Wed May 13, 2015 6:38 pm

StephenR wrote:Oh, OK, I will set the EPAP to 8, I was just being a wus...
the asv handles things very differently, so you MAY be able to lower the epap... but do raise the max ps.

what I'd do is leave it at 8, or no less than 7, and see what the numbers look like tomorrow.

if your apneas are low, then you're probably good.

have you watched the videos on youtube about the adapt?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 64027
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: nearly new ASV settings

Post by Pugsy » Wed May 13, 2015 6:53 pm

When I change to my Adapt I was able to use approx 2 cm lower EPAP than I was using on my other machines...so you might get by with lower EPAP and I would for sure keep that option open since we know that your centrals are directly related to pressure and thus we want to use as little pressure as possible to reduce the chance of triggering centrals ....if we can.
I do agree to open PS max to 15 though. The machine will likely need it to deal with the centrals and it won't go anywhere it doesn't think it needs to go anyway. If it doesn't need it then it won't go up there but it might be needed so don't tie its little hands with limited PS.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

StephenR
Posts: 172
Joined: Fri Jul 25, 2014 4:43 pm

Re: nearly new ASV settings

Post by StephenR » Wed May 13, 2015 8:03 pm

Thanks Pugsy and Palerider...
looking forward to my first night...lol

I clearly do not get out enough...

Stephen

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: nearly new ASV settings

Post by palerider » Wed May 13, 2015 8:24 pm

StephenR wrote:Thanks Pugsy and Palerider...
looking forward to my first night...lol

I clearly do not get out enough...

Stephen
well, we've both got experience first hand with the adapts, pugsy's been using one, and I take care of a friend who uses one....they're very nice machines... pretty simple to set, and very automatic afterwards

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
adipasqu
Posts: 128
Joined: Thu Apr 21, 2011 11:24 am
Location: San Francisco Bay Area, CA

Re: nearly new ASV settings

Post by adipasqu » Wed May 13, 2015 9:09 pm

palerider wrote:[

well, we've both got experience first hand with the adapts, pugsy's been using one, and I take care of a friend who uses one....they're very nice machines... pretty simple to set, and very automatic afterwards
Automatic is correct. Personally I have found the best results by only really setting EPAPmin to prevent OA's, use a minimal PSmin (2 in my case) and let the machine figure out the rest. It's amazing how well and how quickly the algorithms on these high end machines work. Its when you try to limit them that they don't work well (high AHI) and can actually make therapy uncomfortable (aerophagia).

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Now using a ResMed S9 VPAP Adapt (36037) with H5i...SleepyHead software
I am a doctor...just not that kind of doctor.

StephenR
Posts: 172
Joined: Fri Jul 25, 2014 4:43 pm

Re: nearly new ASV settings

Post by StephenR » Thu May 14, 2015 10:44 am

all I can say is I am shocked and amazed...in a good way...

first night...

did reduce EPAP tp 6.6, 8 was just too much...lol

[img][IMG]http://i36.photobucket.com/albums/e22/s ... 44wdu0.jpg[/img][/img]

Stephen
Last edited by StephenR on Thu May 14, 2015 3:54 pm, edited 1 time in total.

User avatar
Pugsy
Posts: 64027
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: nearly new ASV settings

Post by Pugsy » Thu May 14, 2015 10:55 am

The report doesn't surprise me at all....have you ever known a woman to pass up the opportunity to say "I told you so" Well...here it is....."I told you so and a long time ago it was"

There's no need to use a higher EPAP at this point especially since you aren't comfortable with more and there certainly isn't any urgent need to use more.

I know it's probably too soon to notice much difference in overall sleep quality or how you feel...but you look real good on paper at least.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

StephenR
Posts: 172
Joined: Fri Jul 25, 2014 4:43 pm

Re: nearly new ASV settings

Post by StephenR » Thu May 14, 2015 11:00 am

Pugsy wrote:The report doesn't surprise me at all....have you ever known a woman to pass up the opportunity to say "I told you so" Well...here it is....."I told you so and a long time ago it was"

There's no need to use a higher EPAP at this point especially since you aren't comfortable with more and there certainly isn't any urgent need to use more.

I know it's probably too soon to notice much difference in overall sleep quality or how you feel...but you look real good on paper at least.
lol...yes you did tell me...and don't I look good...

Still its early days...one day at a time...

I am forcing myself not to read the thread on the ResMed failed test...

I just could not go back to CPAP at this point...

Stephen

User avatar
Pugsy
Posts: 64027
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: nearly new ASV settings

Post by Pugsy » Thu May 14, 2015 11:14 am

That failed ResMed thread thing....those people were already real sick before the experiment and with any treatment there are always risks...even taking that once a day baby aspirin to ward off heart attacks.
With any treatment we simply have to weight the risks vs the rewards.
I look at xpap therapy as a preventative thing...preventing that bad congestive heart disease in the first place is a much better reward than trying to fix it after it develops. If there is the slightest chance that optimal therapy is preventing something bad from developing down the road...then I am all for it. It's preventing further damage/stress to the body and thus there's a good chance it is preventing that really bad stuff from happening down the road.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: nearly new ASV settings

Post by palerider » Thu May 14, 2015 11:15 am

StephenR wrote:lol...yes you did tell me...and don't I look good...

Still its early days...one day at a time...

I am forcing myself not to read the thread on the ResMed failed test...

I just could not go back to CPAP at this point...

Stephen
can I get in with a 'me too' on what pugsy said?

as to the resmed failed thing, that's apparently a particular subset of a particular condition, a type of congestive heart failure and csr... more details to follow, I'm sure.

that's a great looking chart.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: nearly new ASV settings

Post by HoseCrusher » Thu May 14, 2015 12:08 pm

I love it when a plan comes together...

and it works (as expected).

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...