Distilled Water Only while traveling?
-
- Posts: 17
- Joined: Tue Apr 14, 2015 2:39 pm
Distilled Water Only while traveling?
So I have a conundrum ...
I've had my Airsense S10 Autoset for only 2-3 days now and I leave on a business trip on Sunday for one week. I'll be speaking at a tech conference in Philly and will be at the conference hotel without a rental car.
I was told to use Distilled Water only, otherwise sediment can collect and rust the bottom of my humidifier. However I don't know if I will have access to a drug store within walking distance, especially when I arrive in the late evening on a Sunday.
Is tap water a real no no? Or can I used bottled water which should be available at the hotel?
I've had my Airsense S10 Autoset for only 2-3 days now and I leave on a business trip on Sunday for one week. I'll be speaking at a tech conference in Philly and will be at the conference hotel without a rental car.
I was told to use Distilled Water only, otherwise sediment can collect and rust the bottom of my humidifier. However I don't know if I will have access to a drug store within walking distance, especially when I arrive in the late evening on a Sunday.
Is tap water a real no no? Or can I used bottled water which should be available at the hotel?
Re: Distilled Water Only while traveling?
Lots of people use tap water most of the time, then just clean the tanks more carefully and more often. Tap water won't matter for a 1-2 night stay unless the water's full of e.g. tar, and you can still clean if you feel the need when you get home.
_________________
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear |
Humidifier: IntelliPAP Integrated Heated Humidifier |
-
- Posts: 17
- Joined: Tue Apr 14, 2015 2:39 pm
Re: Distilled Water Only while traveling?
Wow, that makes me fill a ton better. Thank you.
Re: Distilled Water Only while traveling?
While I was in Orlando last week I used tap water. I just made sure that I emptied the tank every morning to lessen the chance of mineral deposits building up on the inside.
_________________
Mask: Eson™ Nasal CPAP Mask with Headgear |
Humidifier: 60 Series Heated Humidifier for Non-Heated Tubing |
- ChicagoGranny
- Posts: 14512
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Distilled Water Only while traveling?
In that case, just feel your humidifier tank with any old water.Madzamboni wrote:makes me fill a ton better
First read: viewtopic/t172378/Sticky--Newbies-PLEAS ... STING.html
Then get a free account: https://home.sleephq.com/
Post links in this thread for members to see and make recommendations.
Then get a free account: https://home.sleephq.com/
Post links in this thread for members to see and make recommendations.
Re: Distilled Water Only while traveling?
though I'd strongly suggest you use much less than a ton of it.ChicagoGranny wrote:In that case, just feel your humidifier tank with any old water.Madzamboni wrote:makes me fill a ton better
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
- Posts: 2744
- Joined: Tue Oct 12, 2010 6:42 pm
Re: Distilled Water Only while traveling?
Another option is to call the hotel. They may be able to supply you with distilled water. The cost may be inflated a little but still manageable. In other words, don't expect it for free.
_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier. |
SpO2 96+% and holding...
Re: Distilled Water Only while traveling?
Bottled water or tap water will be fine for a few days. I've used bottled water for a few weeks in a row while traveling without any problems.
I once resorted to murky tap water in a Dublin hostel that made for a very unpleasant night. But I just cleaned the tank the next day and it was fine.
I once resorted to murky tap water in a Dublin hostel that made for a very unpleasant night. But I just cleaned the tank the next day and it was fine.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Re: Distilled Water Only while traveling?
I have even used tap water here at home (on a well and we have a high mineral content) when I was out of distilled and unable to get to the store to get more.
I just dump the water in the AM and let the tank air dry and then fill it again at night.
If there is any mineral build up it's easily removed with just a little water/vinegar solution sloshed around in the chamber.
I just dump the water in the AM and let the tank air dry and then fill it again at night.
If there is any mineral build up it's easily removed with just a little water/vinegar solution sloshed around in the chamber.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Distilled Water Only while traveling?
One important note - do not use any other kind of bottled water besides distilled because they all have some other stuff in them that you wouldn't want even if it's great to drink... i.e. Evian, sparkling, etc. etc.
_________________
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear |
Humidifier: IntelliPAP Integrated Heated Humidifier |
Re: Distilled Water Only while traveling?
It will kill you in your sleep!Julie wrote:One important note - do not use any other kind of bottled water besides distilled because they all have some other stuff in them that you wouldn't want even if it's great to drink... i.e. Evian, sparkling, etc. etc.
- ChicagoGranny
- Posts: 14512
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Distilled Water Only while traveling?
Quit it!Chevie wrote:It will kill you in your sleep!
First read: viewtopic/t172378/Sticky--Newbies-PLEAS ... STING.html
Then get a free account: https://home.sleephq.com/
Post links in this thread for members to see and make recommendations.
Then get a free account: https://home.sleephq.com/
Post links in this thread for members to see and make recommendations.
Re: Distilled Water Only while traveling?
The bottled plain drinking water isn't as good as distilled water, but is no worse than tap water. I agree to not use sparkling water, etc.Julie wrote:One important note - do not use any other kind of bottled water besides distilled because they all have some other stuff in them that you wouldn't want even if it's great to drink... i.e. Evian, sparkling, etc. etc.
Whoever told you that is wrong, especially for short term use. No tap water that's safe to drink is going to eat up your water tank in a few days. Remember that you can buy a replacement tank, and should replace it every 6 months or so anyway.Madzamboni wrote:I was told to use Distilled Water only, otherwise sediment can collect and rust the bottom of my humidifier. However I don't know if I will have access to a drug store within walking distance, especially when I arrive in the late evening on a Sunday.
Is tap water a real no no? Or can I used bottled water which should be available at the hotel?
ResMed does specify "distilled water only" for the "standard" water tank. It probably says that on the top of the water tank somewhere. There's also a "cleanable" or "dishwashable" water tank that is OK for tap water, but you should dump it and rinse every morning. Personally, I would use tap water anyway for a few days in the standard tank and wash it really well afterwards.
Do not attempt to clean the water tank with a Zamboni or use the meltwater in the CPAP machine.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Re: Distilled Water Only while traveling?
I used Sparklets (drinking water from the cooler) when I run out of distilled water. I only has a few minerals in it to add "taste" and it is only for a night or two.Julie wrote:One important note - do not use any other kind of bottled water besides distilled because they all have some other stuff in them that you wouldn't want even if it's great to drink... i.e. Evian, sparkling, etc. etc.
Carbinated anything is not a good idea.
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin
Re: Distilled Water Only while traveling?
If you are taking a cab to the hotel, you could ask the driver to stop at a pharmacy if there isn't one near your hotel. Since I've been on CPAP, I'm surprised by how many pharmacies there are. It seems there is a CVS or Walgreens around every other corner. Go to Google Maps and enter in your hotel's address. After that, ask google maps for "pharmacies near XXX address."Madzamboni wrote: I've had my Airsense S10 Autoset for only 2-3 days now and I leave on a business trip on Sunday for one week. I'll be speaking at a tech conference in Philly and will be at the conference hotel without a rental car.
Of course, as others have said, a couple of days of bottled or tap water should be fine. The only real downside is the increased minerals that will give you a little extra cleaning to do.
Good Luck!
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Zeo Bedside, CMS-50E Pulse Oximeter |