Problem setting Remstar Series M

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
old_man
Posts: 1
Joined: Sat Apr 18, 2015 12:16 pm

Problem setting Remstar Series M

Post by old_man » Sat Apr 18, 2015 12:22 pm

I picked up a used Remstar Series M model 1022334 for use on the road. I can't seem to get it programmed.

I have read all the info and watched the videos but they all are for the unit with the three buttons on the top. Mine has only two control buttons visible when the top is closed. When I perform the power up with the arrows held down, it does not go into the programming mode.

Does anybody have any info on the magic incantation required for this unit.

User avatar
Pugsy
Posts: 64014
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Problem setting Remstar Series M

Post by Pugsy » Sat Apr 18, 2015 12:30 pm

Have you been here and to the proper manual for whichever M series machine you have?
There's some quick guides and full manuals.
http://www.apneaboard.com/adjust-cpap-p ... tup-manual

Holding down the 2 arrow buttons at the same time while plugging the little plug into the back of the machine should work. I did it often with my M series machine and I think that little step should work with all the M series machines.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Problem setting Remstar Series M

Post by Wulfman... » Sat Apr 18, 2015 1:45 pm

old_man wrote:I picked up a used Remstar Series M model 1022334 for use on the road. I can't seem to get it programmed.

I have read all the info and watched the videos but they all are for the unit with the three buttons on the top. Mine has only two control buttons visible when the top is closed. When I perform the power up with the arrows held down, it does not go into the programming mode.

Does anybody have any info on the magic incantation required for this unit.
According to the service manual for the M Series machines, that "model number" you referenced is the humidifier model number.
You need to open the lid and press the two (top) opposing arrow buttons down (on either side of the LCD screen) while you plug it in and wait for two beeps.

Den


Accessing provider mode unlocks additional settings that cannot be modified while in user mode. To access provider mode:
1. Plug the AC power cord into the power supply and an electrical outlet, as described in Chapter 3.
2. Hold down both the left and right <- -> buttons while simultaneously plugging the power supply into the device.
Once the device beeps twice, it is in provider mode and you can release the buttons.
You can view and modify prescription settings using the Provider Setup screens and view data about device usage on the Provider Data screens.
To navigate the display screens:
• Use the button to navigate to the next screen and the button to navigate to the previous screen.
• Use the + and — buttons to adjust the settings on the Setup screens.


PROVIDER MODE
The following settings can be modified when in Provider Mode:

• Therapy Mode

• (Max) IPAP (BiPAP Auto M Series only - Max if in Auto Bi-LEvel Mode)

• (Min) EPAP (BiPAP Auto M Series only - Min if in Auto Bi-LEvel Mode)

• Max Pressure Support (BiPAP Plus M Series and BiPAP Auto M Series - If in Auto Bi-LEvel Mode)

• Pressure setting (REMstar Pro M Series, REMstar Plus M Series, and REMstar Auto M Series)

• Flex setting (REMstar Pro M Series, REMstar Plus M Series, REMstar Auto M Series, BiPAP Plus M Series, and BiPAP Auto M Series)

• Altitude adjustment setting (REMstar M Series and REMstar Plus M Series)

• Fine pressure adjustment setting

• Ramp time setting

• Ramp starting pressure

• Mask disconnect alert (enable/disable [REMstar Pro M Series, REMstar Plus M Series, and REMstar Auto M Series])

• Auto-off setting (enable/disable [REMstar Pro M Series, REMstar Plus M Series, and REMstar Auto M Series])

• Split Night Time (REMstar Auto M Series and BiPAP Auto M Series - If in Auto mode)

• Patient Reminder (REMstar Pro M Series, REMstar Plus M Series, and REMstar Auto M Series) Additionally, the following data can be viewed when in Provider Mode:

• Therapy usage hours

• Number of sessions greater than four hours

• Blower hours

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Problem setting Remstar Series M

Post by Wulfman... » Sat Apr 18, 2015 1:54 pm

And, if that isn't good enough or if you need a "visual aid", check out this link:

http://www.cpap-reviews.com/2008/cpap-s ... p-machine/


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05