Can Itamar WatchPAT distinguish types of events?
- Darth Lady
- Posts: 249
- Joined: Mon Feb 02, 2015 3:34 pm
- Location: Jersey City NJ
Can Itamar WatchPAT distinguish types of events?
I asked a more involved question that included this as an element, and didn't get a lot of response. So I'm breaking it down.
My machine says my AHI is under 1 every night. Sometimes even zero! A follow-up home study with a WatchPAT, on the other hand, said my AHI was 12 and my RDI was 22!
Now I'm going in for a re-titration study where, of course, they are going to crank the pressure to fix the "events" that the WatchPat study said I was having.
So here's the question: Can WatchPAT distinguish between obstructive/CA events, and "events" generated by, say, plain old O2 desaturation (which it found a lot of) from some other cause, like, say, insufficient respiratory effort while sleeping?
I've searched older threads here, but haven't found a definitive answer.
My machine says my AHI is under 1 every night. Sometimes even zero! A follow-up home study with a WatchPAT, on the other hand, said my AHI was 12 and my RDI was 22!
Now I'm going in for a re-titration study where, of course, they are going to crank the pressure to fix the "events" that the WatchPat study said I was having.
So here's the question: Can WatchPAT distinguish between obstructive/CA events, and "events" generated by, say, plain old O2 desaturation (which it found a lot of) from some other cause, like, say, insufficient respiratory effort while sleeping?
I've searched older threads here, but haven't found a definitive answer.
_________________
Mask: AirFit™ F10 Full Face Mask with Headgear |
Additional Comments: PR System One 960 BiPAP ASV Advanced |
Seeking the Dark Side 8 hours a night... complete with sound effects!
Re: Can Itamar WatchPAT distinguish types of events?
Were you using your CPAP when you took the WatchPAT test?Darth Lady wrote:I asked a more involved question that included this as an element, and didn't get a lot of response. So I'm breaking it down.
My machine says my AHI is under 1 every night. Sometimes even zero! A follow-up home study with a WatchPAT, on the other hand, said my AHI was 12 and my RDI was 22!
I thought the WatchPAT was quackery when I first heard about it, but after a little reading, I think there's some real science there. I'm not entirely convinced that there's enough real world experience yet on how well it works in real life.Darth Lady wrote:So here's the question: Can WatchPAT distinguish between obstructive/CA events, and "events" generated by, say, plain old O2 desaturation (which it found a lot of) from some other cause, like, say, insufficient respiratory effort while sleeping?
Does it claim to distinguish central from obstructive?
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
- Darth Lady
- Posts: 249
- Joined: Mon Feb 02, 2015 3:34 pm
- Location: Jersey City NJ
Re: Can Itamar WatchPAT distinguish types of events?
Yes, I was using my xPAP when I took the WatchPAT test. We ran it because I'm still not feeling rested at night, or even much better than before I started therapy, even though my numbers are great.
I don't think Itamar claims that WatchPAT can distinguish types of apnea. But when they have run it head to head against polysomnograpy, it has been a pretty accurate reflection of what the full-scale study gives for AHI/RDI numbers.
My issue is that I suspect I'm just not breathing deeply enough to get enough oxygen, and that WatchPAT was reading the desaturations and resulting arousals as events (fair enough), and that the sleep lab is reading those events as obstructive and/or central (perhaps not warranted by the evidence).
I don't think Itamar claims that WatchPAT can distinguish types of apnea. But when they have run it head to head against polysomnograpy, it has been a pretty accurate reflection of what the full-scale study gives for AHI/RDI numbers.
My issue is that I suspect I'm just not breathing deeply enough to get enough oxygen, and that WatchPAT was reading the desaturations and resulting arousals as events (fair enough), and that the sleep lab is reading those events as obstructive and/or central (perhaps not warranted by the evidence).
_________________
Mask: AirFit™ F10 Full Face Mask with Headgear |
Additional Comments: PR System One 960 BiPAP ASV Advanced |
Seeking the Dark Side 8 hours a night... complete with sound effects!
Re: Can Itamar WatchPAT distinguish types of events?
Do you have your data from the ASV the night of the WatchPAT test? Be sure and save a copy of the SD card from that date.
If the ASV says no apneas on that night, and the WatchPAT says yes, the WatchPAT is wrong. It might be interesting to look at your flow rate waveforms and other data at the times the WatchPAT shows events.
However, you might have some more subtle problem that the WatchPAT is seeing. Not "apnea," but some other problem such as hypoventillation.
Do you have real data from the WatchPAT, such as graphs of SpO2 or whatever it records?
It might be worthwhile for you to get one of the CMS recording oximeters that works with SleepyHead and get some more data.
If the ASV says no apneas on that night, and the WatchPAT says yes, the WatchPAT is wrong. It might be interesting to look at your flow rate waveforms and other data at the times the WatchPAT shows events.
However, you might have some more subtle problem that the WatchPAT is seeing. Not "apnea," but some other problem such as hypoventillation.
Do you have real data from the WatchPAT, such as graphs of SpO2 or whatever it records?
It might be worthwhile for you to get one of the CMS recording oximeters that works with SleepyHead and get some more data.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Re: Can Itamar WatchPAT distinguish types of events?
Darthlady, you can get your print out of the detailed WatchPat report with all the graphs. I think you're Kaiser, so you can request it, and may be able to get it electronically.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: DreamWear Nasal CPAP Mask with Headgear |
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Re: Can Itamar WatchPAT distinguish types of events?
Yes. And you can also walk into the Member Services office of your local Kaiser and they should be able to print it out for you.Janknitz wrote:Darthlady, you can get your print out of the detailed WatchPat report with all the graphs. I think you're Kaiser, so you can request it, and may be able to get it electronically.
_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle |
Mask: Aloha Nasal Pillow CPAP Mask with Headgear |
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura |
- Darth Lady
- Posts: 249
- Joined: Mon Feb 02, 2015 3:34 pm
- Location: Jersey City NJ
Re: Can Itamar WatchPAT distinguish types of events?
I've got all the data; I download it into SleepyHead regularly. I don't have real data from the WatchPAT, just the hypercondensed printout they stick in the report. I actually do suspect that I'm hypoventilating rather than having obstructive events, and I suspect (don't know, which is why I started this thread) that WatchPAT uses desats along with circulatory changes/arousals to flag events. I also suspect (but don't know) that it can't tell hypoventilation from an obstructive event.archangle wrote:Do you have your data from the ASV the night of the WatchPAT test? Be sure and save a copy of the SD card from that date.
If the ASV says no apneas on that night, and the WatchPAT says yes, the WatchPAT is wrong. It might be interesting to look at your flow rate waveforms and other data at the times the WatchPAT shows events.
However, you might have some more subtle problem that the WatchPAT is seeing. Not "apnea," but some other problem such as hypoventillation.
Do you have real data from the WatchPAT, such as graphs of SpO2 or whatever it records?
It might be worthwhile for you to get one of the CMS recording oximeters that works with SleepyHead and get some more data.
_________________
Mask: AirFit™ F10 Full Face Mask with Headgear |
Additional Comments: PR System One 960 BiPAP ASV Advanced |
Seeking the Dark Side 8 hours a night... complete with sound effects!
- Darth Lady
- Posts: 249
- Joined: Mon Feb 02, 2015 3:34 pm
- Location: Jersey City NJ
Re: Can Itamar WatchPAT distinguish types of events?
I'm not Kaiser; I'm BCBS, but I can ask my sleep doc for the whole data set/extended report. I'm seeing her tomorrow, to discuss whether more pressure is really the answer to everythingkaiasgram wrote:Yes. And you can also walk into the Member Services office of your local Kaiser and they should be able to print it out for you.Janknitz wrote:Darthlady, you can get your print out of the detailed WatchPat report with all the graphs. I think you're Kaiser, so you can request it, and may be able to get it electronically.
_________________
Mask: AirFit™ F10 Full Face Mask with Headgear |
Additional Comments: PR System One 960 BiPAP ASV Advanced |
Seeking the Dark Side 8 hours a night... complete with sound effects!
Re: Can Itamar WatchPAT distinguish types of events?
If you have real central apnea (cycling) the PAT signal may have the same waxing/waning characteristics:Darth Lady wrote:Can WatchPAT distinguish between obstructive/CA events, and "events" generated by, say, plain old O2 desaturation (which it found a lot of) from some other cause, like, say, insufficient respiratory effort while sleeping?
Re: Can Itamar WatchPAT distinguish types of events?
That's the ASV Fallacy (Cover It With A Band Aid - If You Don't See It, It Must Not Be There).Darth Lady wrote:My machine says my AHI is under 1 every night. Sometimes even zero! A follow-up home study with a WatchPAT, on the other hand, said my AHI was 12 and my RDI was 22!
- Darth Lady
- Posts: 249
- Joined: Mon Feb 02, 2015 3:34 pm
- Location: Jersey City NJ
Re: Can Itamar WatchPAT distinguish types of events?
I do have a periodicity to my breathing, sometimes quite pronounced, that the machine tries to correct, that doesn't "grow up" into anything it considers flag-able. As for the rest of it, I guess I'll have to put up with yet another sleep study . I just hope I can get my doc to look for underlying causes, and at other solutions besides "crank the pressure."
_________________
Mask: AirFit™ F10 Full Face Mask with Headgear |
Additional Comments: PR System One 960 BiPAP ASV Advanced |
Seeking the Dark Side 8 hours a night... complete with sound effects!
- Darth Lady
- Posts: 249
- Joined: Mon Feb 02, 2015 3:34 pm
- Location: Jersey City NJ
Re: Can Itamar WatchPAT distinguish types of events?
Oh, and I'm curious to know what it is about the ASV algorithms that would make it so inaccurate, if that's the issue.Morbius wrote:That's the ASV Fallacy (Cover It With A Band Aid - If You Don't See It, It Must Not Be There).Darth Lady wrote:My machine says my AHI is under 1 every night. Sometimes even zero! A follow-up home study with a WatchPAT, on the other hand, said my AHI was 12 and my RDI was 22!
_________________
Mask: AirFit™ F10 Full Face Mask with Headgear |
Additional Comments: PR System One 960 BiPAP ASV Advanced |
Seeking the Dark Side 8 hours a night... complete with sound effects!
Re: Can Itamar WatchPAT distinguish types of events?
I did not say it was inaccurate.Darth Lady wrote:Oh, and I'm curious to know what it is about the ASV algorithms that would make it so inaccurate...
I said there is a Fallacy.
This was disclosed during the famous State of Confusion v. ASV (2005) trial where, during brisk cross examination, it was revealed:
Prosecution: Did you send in nice-looking breaths?
ASV: Yes.
Prosecution: Did they actually do anything?
ASV: No.
ASV technology was designed to break the chain of central events, thus solving the problem. It will rapidly attack central apneas and back off during normal breathing until the chain is broken and only normal breathing ensues.
However, if your machine is attacking all night the problem is never getting solved. The ASV will put up some nice waveforms and read it as acceptable breathing, but the underlying event persists.
I am postulating that's why the discrepancy exists between the ASV readings and Watch PAT(which can look under the Band-Aid).
Re: Can Itamar WatchPAT distinguish types of events?
Morbius,Morbius wrote: ASV technology was designed to break the chain of central events, thus solving the problem. It will rapidly attack central apneas and back off during normal breathing until the chain is broken and only normal breathing ensues.
However, if your machine is attacking all night the problem is never getting solved. The ASV will put up some nice waveforms and read it as acceptable breathing, but the underlying event persists.
I am postulating that's why the discrepancy exists between the ASV readings and Watch PAT(which can look under the Band-Aid).
Are you saying the ASV is reporting a nice low AHI because it's confusing the "nice regular waveform" that's being generated by the ASV's aggressive attacking of the central apneas with real normal breathing? And the Watch PAT is not confused by the ASV-generated "nice regular waveform" because it's looking at different data? And the Watch PAT can see that in spite of the ASV-triggered inhalations which smooth the waveform out, the desats are still happening?
And a more general question for you: When is "ASV confusion" most likely to happen?
_________________
Machine: DreamStation BiPAP® Auto Machine |
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5 |
- Darth Lady
- Posts: 249
- Joined: Mon Feb 02, 2015 3:34 pm
- Location: Jersey City NJ
Re: Can Itamar WatchPAT distinguish types of events?
I'm also looking forward to the answer to that question.
Saw the sleep doc yesterday. She seems convinced that I'm having obstructive events although she also says that WatchPAT can't tell what kind of event you're having. I told her that my concern is that this may be part of other neurological problems I am having; that the muscles I use for breathing are being affected and when I am asleep I can't compensate. I will just have to see how Sleep Study No. 4 goes.
Saw the sleep doc yesterday. She seems convinced that I'm having obstructive events although she also says that WatchPAT can't tell what kind of event you're having. I told her that my concern is that this may be part of other neurological problems I am having; that the muscles I use for breathing are being affected and when I am asleep I can't compensate. I will just have to see how Sleep Study No. 4 goes.
_________________
Mask: AirFit™ F10 Full Face Mask with Headgear |
Additional Comments: PR System One 960 BiPAP ASV Advanced |
Seeking the Dark Side 8 hours a night... complete with sound effects!