Just starting but diagnosed setting not enough O2

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
ByronB
Posts: 3
Joined: Tue Mar 31, 2015 4:21 pm

Just starting but diagnosed setting not enough O2

Post by ByronB » Wed Apr 01, 2015 8:23 pm

Just diagnosed with rather severe apnea, titration determined setting of 7 needed and I'm trying a donated CPAP machine.

Problem: feel suffocated and leg is seriously falling asleep. I'm not getting enough air.

Having the same problem with:
- pressure setting of 7 and even 7.4.
- EPR on or off
- Ramp feature on or off (and when I use it, after the ramp time is up, the pressure seems to go down--that's a head-scratcher)

I know this defies common sense, but it's my experience over and over. Very maddening.

I see a tech tomorrow, but I really would relish your ideas on how to solve this. Thanks ahead of time.
Last edited by ByronB on Wed Apr 01, 2015 8:40 pm, edited 1 time in total.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Just starting but diagnosed setting not enough O2

Post by palerider » Wed Apr 01, 2015 8:33 pm

ByronB wrote:Just diagnosed with rather severe apnea, titration determined setting of 7 needed and I'm trying a donated CPAP machine.

Problem: feel suffocated and leg is seriously falling asleep. I'm not getting enough air.

Having the same problem with:
- pressure setting of 7 and even 7.4.
- EPR on or off
- Ramp feature on or off (and when I use it, after the ramp time is up, the pressure seems to go down--that's a head-scratcher)

I see a tech tomorrow, but I really would relish your ideas on how to solve this. Thanks ahead of time.
8 or 9.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Just starting but diagnosed setting not enough O2

Post by Goofproof » Wed Apr 01, 2015 8:45 pm

Without a steering wheel installed, (A machine that collects full data, a program that reads it, there is not way to even guess what the problem is)

For treatment to work, the pressure in your airway has to be high enough to hold the airway open. Excessive leaks can stop treatment from working. You can try higher pressure if leaks aren't the problem, if leaks are the problem will get worse. Jim

If using the ramp feature, turn it off, it can starve you of air.
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
Pugsy
Posts: 64030
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Just starting but diagnosed setting not enough O2

Post by Pugsy » Wed Apr 01, 2015 8:57 pm

What mask are you using?
If it is a nasal pillow mask can you try the next size larger in the nasal pillows?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

ByronB
Posts: 3
Joined: Tue Mar 31, 2015 4:21 pm

Re: Just starting but diagnosed setting not enough O2

Post by ByronB » Wed Apr 01, 2015 10:26 pm

Thought my profile was supposed to show my machine:
Resmed S9 Elite with ERP. Using a Nasal mask.

Mask leak seems to not be an issue according to display.

Ramp was off last time I tried, as I said (either way, problem occurs).

Looks like I need to up the pressure substantially. Thanks for your comments so far.

User avatar
Pugsy
Posts: 64030
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Just starting but diagnosed setting not enough O2

Post by Pugsy » Wed Apr 01, 2015 10:37 pm

Equipment profile isn't showing anything yet.
Did you click on "submit" after you made your choices? If you didn't click on "submit" the changes won't stick and nothing will show up.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.