Pr system one clinical mode

Post by Hurricane » Tue Mar 31, 2015 4:52 pm

Just got my replacement Pr System One Remstar 60 series BiPap auto sv advanced machine being first one was defective, noticed AHI feature is not enabled on new machine, how to enable AHI screen.
Ed

_________________
Mask: Swift™ LT Nasal Pillow CPAP Mask with Headgear
Additional Comments: Some times you are the bug, sometimes the windshield.

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Pr system one clinical mode

Post by robysue » Tue Mar 31, 2015 5:05 pm

Turn the knob to select Setup from the LCD screen. Then press and hold the knob AND the ramp button down until the machine beeps at you. (It takes several seconds so be patient). You're then in the Clinician's mode. Select Setup and press the knob to enter the clinician's set up memory. Turn the knob to scroll down through the settings until you reach the one that says something like Show AHI and make sure it's set to On. To set a particular setting, you click the setting while it is highlighted. Turn the knob until the desired value for the setting is highlighted. Click the knob to set the new value.

To get out of the Clinical mode, scroll down to exit and press the knob.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

Hurricane
Posts: 578
Joined: Thu Oct 26, 2006 1:46 pm
Location: New Jersey

Re: Pr system one clinical mode

Post by Hurricane » Tue Mar 31, 2015 5:29 pm

Thanks Robysue,
All fixed,, thank you again.