Pressure settings on A10 autoset

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
chefmate
Posts: 30
Joined: Sun Mar 08, 2015 5:33 am

Pressure settings on A10 autoset

Post by chefmate » Thu Mar 26, 2015 5:59 pm

My doctor prescribed me a pressure of 7-20 H20 cm but when I looked at the clinical settings on autoset A10 my minimum pressure was set at 4 and maximum was set at 20. Why is the minimum pressure not set at 7 according to the prescription?
Machine: AirSense™ 10 AutoSet CPAP Machine with HumidAir™ Heated Humidifier
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
ClimateLineAir™ Heated Tube for AirSense™ 10 and AirCurve™ 10 Machines

User avatar
SleepDisturbed
Posts: 276
Joined: Thu Apr 25, 2013 1:52 pm
Location: East Texas

Re: Pressure settings on A10 autoset

Post by SleepDisturbed » Thu Mar 26, 2015 6:46 pm

Seems as if your DME handed you an unconfigured machine, and did not set it up according to your prescription. I believe hat 4-20 is the factory default. Not unusual behavior for a DME.

You can either set it up yourself or take it back to them.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: 9.6 - 15 Pressure on APAP. EPR 2 / Sleepyhead software
What I lack in verbosity, I make up in brevity.

bucksnort
Posts: 6
Joined: Thu Mar 26, 2015 9:48 am

Re: Pressure settings on A10 autoset

Post by bucksnort » Thu Mar 26, 2015 10:26 pm

As explained to me when i picked up my machine today. It starts at 4 and ramps up to at least 7 and then adjusts as needed. No need to worry

User avatar
TangledHose
Posts: 862
Joined: Wed Mar 13, 2013 11:13 am
Location: Centennial, CO

Re: Pressure settings on A10 autoset

Post by TangledHose » Thu Mar 26, 2015 10:44 pm

The ramp start pressure can be adjusted in the clinical menu, as well as therapy pressure.

Many folks can not tolerate 4cm pressure as it is so low most people get that "starved for air" feeling.

To enter the clinical menu push and hold both the round selector button and the home button simultaneously for about three seconds or so, you'll see the screen change indicating access to that menu.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: Also own F&P Icon Auto for backup and travel; and a Transcend when packing space is limited
Link to Sleepyhead:
http://sleepfiles.com/SH2/

User avatar
chunkyfrog
Posts: 34397
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: Pressure settings on A10 autoset

Post by chunkyfrog » Fri Mar 27, 2015 9:43 am

I take my reported pressure off the screen, and bracket it a couple of cm. either way.
I write down my pressure (well, everything) every morning, and can change pressure to accommodate any change.
I monitor my air like I do my blood sugar, but I don't have to get stuck since I fired Crapria.