can anybody tell me how my bipap machine delivers the two prescribed pressure. my current pressure is set to 20CMH20 does this mean that my machine uses a pressure of 10 for inhaling and another 10 for exhaling also I was also wondering when I was laying in bed I noticed that my Bipap machine would stay at 6.0 CMH2O will my machine get up to the prescribed pressure when I am sound asleep. thanks for answering.
Re: questions reguarding the two prssures for my bipap machine
the machine achieves different pressures by varying the speed of the blower to change the flow rate of the air.springman946 wrote:can anybody tell me how my bipap machine delivers the two prescribed pressure. my current pressure is set to 20CMH20 does this mean that my machine uses a pressure of 10 for inhaling and another 10 for exhaling also I was also wondering when I was laying in bed I noticed that my Bipap machine would stay at 6.0 CMH2O will my machine get up to the prescribed pressure when I am sound asleep. thanks for answering.
flow rate into a flow restriction (the vent on your mask) gives a pressure rise, greater flow, greater pressure, a flow sensor measures flow rate, and a pressure sensor measures the pressure that results, and a computer tells the blower how fast to spin.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Darth Lady
- Posts: 249
- Joined: Mon Feb 02, 2015 3:34 pm
- Location: Jersey City NJ
Re: questions reguarding the two prssures for my bipap machine
Are you using ramp? That would start out at a lower pressure and work its way up to prescribed pressure. A lot of people here (including me) have decided that ramp is more trouble than it is worth.springman946 wrote:can anybody tell me how my bipap machine delivers the two prescribed pressure. my current pressure is set to 20CMH20 does this mean that my machine uses a pressure of 10 for inhaling and another 10 for exhaling also I was also wondering when I was laying in bed I noticed that my Bipap machine would stay at 6.0 CMH2O will my machine get up to the prescribed pressure when I am sound asleep. thanks for answering.
It is also possible that 6 is a minimum pressure (see below), so it starts off with that, varying up from there as needed. Mine, for instance (I have this machine) is never below 10, and always starts out there.
With the machine you are using, pressures would be a minimum and a maximum, I believe both for inspiration and expiration. Anyway, it wouldn't be just "20 cm H2O." Do you have a copy of your prescription? Anyway, it's not 10 and 10; that's not how it works (or shouldn't be with this kind of equipment). Speaking of which, how is that workin' for you? If you don't have it, hopefully one of the helpful folks here will post that link for Sleepyhead, or you can get it from Pugsy's Pointers at the top, download it, and see how you are doing.
_________________
Mask: AirFit™ F10 Full Face Mask with Headgear |
Additional Comments: PR System One 960 BiPAP ASV Advanced |
Seeking the Dark Side 8 hours a night... complete with sound effects!
Re: questions reguarding the two prssures for my bipap machine
Bipap means that you have one pressure for inhalation and a slightly lower pressure for exhalation. A lower exhalation pressure makes it easier to exhale and a lower pressure is sufficient to keep the airway open.
If you have ramp on then turn it off. It is best to know how well your mask is sealing when you are at full pressure and still awake to make adjustments.
Your pressure is high. If you have trouble with your mask leaking, try pillows. Pillows are self sealing (the pressure makes the seal, not the straps) so I find they work much better.
If you have ramp on then turn it off. It is best to know how well your mask is sealing when you are at full pressure and still awake to make adjustments.
Your pressure is high. If you have trouble with your mask leaking, try pillows. Pillows are self sealing (the pressure makes the seal, not the straps) so I find they work much better.
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/ |
Use data to optimize your xPAP treatment:
how to see your data https://sleep.tnet.com/resources/sleepyhead
how to present your data https://sleep.tnet.com/resources/sleepyhead/shorganize
how to post your data https://sleep.tnet.com/reference/tips/imgur
how to see your data https://sleep.tnet.com/resources/sleepyhead
how to present your data https://sleep.tnet.com/resources/sleepyhead/shorganize
how to post your data https://sleep.tnet.com/reference/tips/imgur