BiPAP not picking up inspiration sometimes
-
- Posts: 15
- Joined: Wed Jan 03, 2024 3:04 pm
BiPAP not picking up inspiration sometimes
Hi! I'm migrating from CPAP to BiPAP and have been struggling with the device not picking up my inspiration and not triggering the IPAP often. Could this be a breathing problem or a device issue? I initially noticed a clogged nostril but even after decongestion it kept happening.
Last edited by unputdownnable on Wed Jan 03, 2024 10:00 pm, edited 1 time in total.
Re: BiPAP not picking up inspiration sometimes
Knowing brand and model of machine might be helpful to any responders.
_________________
Machine: Airsense 10 Card to Cloud |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
The people who confuse "entomology" and "etymology" really bug me beyond words.
---
A love song to a CPAP? Oh please!:
https://youtu.be/_e32lugxno0?si=W4W9EnrZZTD5Ow6p
---
A love song to a CPAP? Oh please!:
https://youtu.be/_e32lugxno0?si=W4W9EnrZZTD5Ow6p
- ChicagoGranny
- Posts: 14543
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: BiPAP not picking up inspiration sometimes
Many possibilities. Please state machine and mask models, mode setting, pressure settings including ramp, ...unputdownnable wrote: ↑Wed Jan 03, 2024 3:11 pmHi! I'm migrating from CPAP to BiPAP and has struggling with the device not picking up my inspiration and not triggering the IPAP often. Could this be a breathing problem or a device issue? I initially noticed a clogged nostril but even after decongestion it kept happening.
If you don't know how to get into the clinical menu to check the settings, find a youtube video for your model machine explaining it.
-
- Posts: 15
- Joined: Wed Jan 03, 2024 3:04 pm
Re: BiPAP not picking up inspiration sometimes
Here is my setup:
Device: BPAP AUTO GII BMC T-20A Ser. (https://en.bmc-medical.com/products/npa ... t-25a.html)
Mode: S
Mask: Phillips Nasal DreamWear
EPAP: 6 cmH2O
IPAP: 10.5 cmH2O
Ramp: 10min
I Sense: 8
E Sense: 8
Hose length: ~1.75m (no bacterial filter)
P.S.: BTW, the hose is passing through a hanger on the wall so that it does not get too lengthy on the bed.
Device: BPAP AUTO GII BMC T-20A Ser. (https://en.bmc-medical.com/products/npa ... t-25a.html)
Mode: S
Mask: Phillips Nasal DreamWear
EPAP: 6 cmH2O
IPAP: 10.5 cmH2O
Ramp: 10min
I Sense: 8
E Sense: 8
Hose length: ~1.75m (no bacterial filter)
P.S.: BTW, the hose is passing through a hanger on the wall so that it does not get too lengthy on the bed.
Last edited by unputdownnable on Wed Jan 03, 2024 8:22 pm, edited 6 times in total.
- ChicagoGranny
- Posts: 14543
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: BiPAP not picking up inspiration sometimes
Is it the AirCurve 10V? The model name should be on the front of the machine. https://www.cpap.com/productpage/resmed ... el-machine
Can you post a typical day in imgur.com and share the link in this thread? Instructions in my signature.
Can you post a typical day in imgur.com and share the link in this thread? Instructions in my signature.
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Re: BiPAP not picking up inspiration sometimes
As a general rule, ResMed bilevels have a setting in the clinician menu for adjusting trigger sensitivity to help it recognize when you breathe in. There should also be a sensitivity adjustment for cycle, when you breathe out.
I believe most users do well with "medium" for both, although a few users are more comfortable with changing those for increased or decreased sensitivity."Under normal conditions, the device triggers (initiates IPAP) and cycles (terminates IPAP and changes to EPAP) as it senses the change in patient flow. . . . The device has adjustable trigger/cycle sensitivity to optimize the sensing level according to patient conditions."--https://www.resmed.com/en-us/healthcare ... echnology/
_________________
Machine: Airsense 10 Card to Cloud |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
The people who confuse "entomology" and "etymology" really bug me beyond words.
---
A love song to a CPAP? Oh please!:
https://youtu.be/_e32lugxno0?si=W4W9EnrZZTD5Ow6p
---
A love song to a CPAP? Oh please!:
https://youtu.be/_e32lugxno0?si=W4W9EnrZZTD5Ow6p
-
- Posts: 15
- Joined: Wed Jan 03, 2024 3:04 pm
Re: BiPAP not picking up inspiration sometimes
My device has sensitivities ranging from 1 to 8. I have played with both "I Sensitivity" and "E Sensitivity" for many days. Both are at maximum lately (=8). But still, every once in a while when I finish breathing out and start to breathe in, it does not switch the pressure and it gets hard to breathe in.
Re: BiPAP not picking up inspiration sometimes
How long is the hose?unputdownnable wrote: ↑Wed Jan 03, 2024 3:11 pmHi! I'm migrating from CPAP to BiPAP and has struggling with the device not picking up my inspiration and not triggering the IPAP often. Could this be a breathing problem or a device issue? I initially noticed a clogged nostril but even after decongestion it kept happening.
Do you have anything in the hose line like a bacterial filter?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
Happiness is being on Dog Slobber pr & zonkers foe list
CrankyGranny is Whale Road + many other ids
They are here to help.
CrankyGranny is Whale Road + many other ids
They are here to help.
zonkers + palerider aka GrumpyHere wrote: ↑What exactly do you think you're adding to this thread?
-
- Posts: 15
- Joined: Wed Jan 03, 2024 3:04 pm
Re: BiPAP not picking up inspiration sometimes
It's actually "BPAP AUTO GII BMC T-20A Ser.". I will have a look at the instructions.ChicagoGranny wrote: ↑Wed Jan 03, 2024 6:36 pmIs it the AirCurve 10V? The model name should be on the front of the machine. https://www.cpap.com/productpage/resmed ... el-machine
Can you post a typical day in imgur.com and share the link in this thread? Instructions in my signature.
Re: BiPAP not picking up inspiration sometimes
I'm not aware of any such model. what's it say on the front of the machine?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
- Posts: 15
- Joined: Wed Jan 03, 2024 3:04 pm
Re: BiPAP not picking up inspiration sometimes
Hose is about 1.7-1.8m. No bacterial filter.SleepGeek wrote: ↑Wed Jan 03, 2024 7:44 pmHow long is the hose?unputdownnable wrote: ↑Wed Jan 03, 2024 3:11 pmHi! I'm migrating from CPAP to BiPAP and has struggling with the device not picking up my inspiration and not triggering the IPAP often. Could this be a breathing problem or a device issue? I initially noticed a clogged nostril but even after decongestion it kept happening.
Do you have anything in the hose line like a bacterial filter?
-
- Posts: 15
- Joined: Wed Jan 03, 2024 3:04 pm
Re: BiPAP not picking up inspiration sometimes
Re: BiPAP not picking up inspiration sometimes
RESmart GII BPAP System (T-20A / T-25A)
https://en.bmc-medical.com/contact/
International Headquarters
BMC Medical Co., Ltd.
Add: Room 110 Tower A Fengyu Building, No.115 Fucheng Road, Haidian, 100036 Beijing, People's Republic of China
Phone: +86 18622933511
Fax: 86-10-5166 3880 ext. 810
Email: intl@bmc-medical.com
For Technical service: service1@bmc-medical.com
https://en.bmc-medical.com/contact/
International Headquarters
BMC Medical Co., Ltd.
Add: Room 110 Tower A Fengyu Building, No.115 Fucheng Road, Haidian, 100036 Beijing, People's Republic of China
Phone: +86 18622933511
Fax: 86-10-5166 3880 ext. 810
Email: intl@bmc-medical.com
For Technical service: service1@bmc-medical.com
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
- ChicagoGranny
- Posts: 14543
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: BiPAP not picking up inspiration sometimes
BMC's ( ) machines are not supported by OSCAR.
-
- Posts: 15
- Joined: Wed Jan 03, 2024 3:04 pm
Re: BiPAP not picking up inspiration sometimes
Just wanted to share that I've noticed that this was happening during ramp only. I've made 2 changes that seem to not trigger the issue anymore for the last 2 nights:unputdownnable wrote: ↑Wed Jan 03, 2024 3:11 pmHi! I'm migrating from CPAP to BiPAP and have been struggling with the device not picking up my inspiration and not triggering the IPAP often. Could this be a breathing problem or a device issue? I initially noticed a clogged nostril but even after decongestion it kept happening.
1) Increased the ramp time from 10min to 15min.
2) Reduced the "E Sensibility" from 8 (max) to 7.
If the issue comes back I will update in the thread.