I got my o2 setup Friday afternoon. After reading that you should not use petroleum products or grease with the oxygen, I asked the man who set it up about Lansinoh. I use it on my lips every night. It is the only product that doesn't dry out over night and I want to be able to continue using it. He didn't know the answer. Does anyone know if it is safe to use with oxygen?
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Additional Comments: Supplemental oxygen at 2.5 lpm |
AutoSV 960 with heated hose. Settings: EPAP Min-12, EPAP Max-17, PS Min-5.5, PS Max-13, Max Pressure-25, Rate-Auto, Rise Time 1. Use Sleepyhead and Encore Pro.
Re: Lansinoh and O2?
Lanolin isn't flammable.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear |
Additional Comments: Oscar and Rescan |
Mask: bleep and F30i
Machine: Resmed S10 Autoset
Machine: Resmed S10 Autoset
- SleepDisturbed
- Posts: 276
- Joined: Thu Apr 25, 2013 1:52 pm
- Location: East Texas
Re: Lansinoh and O2?
I can say with 100% certainty that it is not a Petroleum product.
I don't know what they mean by grease, as that is a very vague and broad term. I suspect they are referring to petroleum based grease, but i dont know. However, since Lansinoh has been used for a very long time on nipples to soothe irritation from breast feeding, unless your oxygen equipment is more delicate than a newborn baby, I would not really expect any issues.
But don't send me a repair bill if I am wrong, as this is just a SWAG!
I don't know what they mean by grease, as that is a very vague and broad term. I suspect they are referring to petroleum based grease, but i dont know. However, since Lansinoh has been used for a very long time on nipples to soothe irritation from breast feeding, unless your oxygen equipment is more delicate than a newborn baby, I would not really expect any issues.
But don't send me a repair bill if I am wrong, as this is just a SWAG!
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: 9.6 - 15 Pressure on APAP. EPR 2 / Sleepyhead software |
What I lack in verbosity, I make up in brevity.
Re: Lansinoh and O2?
Are you sure? Even if it isn't flammable like other animal fats, many things are flammable in pure O2 even if they aren't otherwise flammable?rdfry wrote:Lanolin isn't flammable.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Re: Lansinoh and O2?
That is my concern. I haven't found any information about it specifically, and don't want to risk it if it is a potential fire hazard. OTOH, I don't want to give it up unnecessarily.
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Additional Comments: Supplemental oxygen at 2.5 lpm |
AutoSV 960 with heated hose. Settings: EPAP Min-12, EPAP Max-17, PS Min-5.5, PS Max-13, Max Pressure-25, Rate-Auto, Rise Time 1. Use Sleepyhead and Encore Pro.
Re: Lansinoh and O2?
oh, it probably is.rdfry wrote:Lanolin isn't flammable.
you'd be surprised what burns in an oxygen rich environment.
however, 2-3lpm suppox in a cpap feed likely to be an issue.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Lansinoh and O2?
Fixed that for you pr...palerider wrote:oh, it probably is.rdfry wrote:Lanolin isn't flammable.
you'd be surprised what burns in an oxygen rich environment.
however, 2-3lpm suppox in a cpap feed isn't likely to be an issue.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Backup/travel unit is an identical S9 AutoSet for Her w/Eson nasal mask |
Re: Lansinoh and O2?
I honestly wouldn't worry about it... added o2 in a cpap feed results in much less o2 than wearing a canula or mask....M'ohms wrote:That is my concern. I haven't found any information about it specifically, and don't want to risk it if it is a potential fire hazard. OTOH, I don't want to give it up unnecessarily.
it's extra, and things would burn better... but you don't have to be as careful as with pure o2 (which makes steel wool burn like gunpowder )
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- TangledHose
- Posts: 862
- Joined: Wed Mar 13, 2013 11:13 am
- Location: Centennial, CO
Re: Lansinoh and O2?
I use a substantial application of Vaseline Petroleum Jelly on lips every single night before putting on my FFM and turning on APAP with 2 lpm supplemental O2......been doing it for years.......can't see how Lansinoh would be any worse as far as O2 is concerned. I agree with PR, just not a high enough concentration of O2 to make much difference....... just don't smoke your Crack Pipe in the process......
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Simplus Full Face CPAP Mask with Headgear |
Additional Comments: Also own F&P Icon Auto for backup and travel; and a Transcend when packing space is limited |
Link to Sleepyhead:
http://sleepfiles.com/SH2/
http://sleepfiles.com/SH2/
Re: Lansinoh and O2?
Thanks for the answers. Tangled Hose, I will think of you every time I slather up, crack pipe and all!
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Additional Comments: Supplemental oxygen at 2.5 lpm |
AutoSV 960 with heated hose. Settings: EPAP Min-12, EPAP Max-17, PS Min-5.5, PS Max-13, Max Pressure-25, Rate-Auto, Rise Time 1. Use Sleepyhead and Encore Pro.
Re: Lansinoh and O2?
This MSDS lists it as flammable.
http://www.pharmcoaaper.com/pages/MSDS/ ... anolin.pdf
Sounds like it's not terribly flammable, but once again, O2 changes the equation.
http://www.pharmcoaaper.com/pages/MSDS/ ... anolin.pdf
Sounds like it's not terribly flammable, but once again, O2 changes the equation.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Re: Lansinoh and O2?
Well then I stand corrected. I doubt that the little bit used is ever going to be a problem.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear |
Additional Comments: Oscar and Rescan |
Mask: bleep and F30i
Machine: Resmed S10 Autoset
Machine: Resmed S10 Autoset
- SleepDisturbed
- Posts: 276
- Joined: Thu Apr 25, 2013 1:52 pm
- Location: East Texas
Re: Lansinoh and O2?
After doing some additional reading on this I believe this is a simple misunderstanding. It seems to me that the prohibition on petroleum products and grease is intended to apply to using it on the O2 equipment itself, such as in fittings.
I don't think it has to do with on the patient.
But, I may be wrong.
I don't think it has to do with on the patient.
But, I may be wrong.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: 9.6 - 15 Pressure on APAP. EPR 2 / Sleepyhead software |
What I lack in verbosity, I make up in brevity.
- TangledHose
- Posts: 862
- Joined: Wed Mar 13, 2013 11:13 am
- Location: Centennial, CO
Re: Lansinoh and O2?
M'ohms wrote:Thanks for the answers. Tangled Hose, I will think of you every time I slather up, crack pipe and all!
<giggle>.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Simplus Full Face CPAP Mask with Headgear |
Additional Comments: Also own F&P Icon Auto for backup and travel; and a Transcend when packing space is limited |
Link to Sleepyhead:
http://sleepfiles.com/SH2/
http://sleepfiles.com/SH2/
Re: Lansinoh and O2?
I read that, while your're not supposed to use it on the machine itself, you are not to use it around the oxygen (patient) either.
From WebMD: http://www.webmd.com/lung/copd/safety-t ... ing-oxygen
"Do not use oil, grease or petroleum-based products on the equipment. Do not use it near you while you use oxygen. These materials are highly flammable and will burn readily with the presence of oxygen. Avoid petroleum-based lotions or creams, like Vaseline, on your face or upper chest."
From WebMD: http://www.webmd.com/lung/copd/safety-t ... ing-oxygen
"Do not use oil, grease or petroleum-based products on the equipment. Do not use it near you while you use oxygen. These materials are highly flammable and will burn readily with the presence of oxygen. Avoid petroleum-based lotions or creams, like Vaseline, on your face or upper chest."