Understanding Flow Rate (ASV)

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Uncle Flapp
Posts: 247
Joined: Thu Nov 14, 2013 2:24 pm
Location: Arizona

Understanding Flow Rate (ASV)

Post by Uncle Flapp » Mon Jan 26, 2015 6:20 pm

Things with my sleep have been relatively uneventful (no pun intended) for months but I recently observed a pattern that has me scratching my head. I am hoping y'all can help me understand the data.

The snippet here shows me awake and drifting off:
Image

This is my normal sleep pattern and what I see throughout most of the night. Wave form is pretty regular:
Image

A couple of times per night, some weird stuff happens. This is where I could use some help. I'm not sure what causes this.
I see a series of short breaths followed by large breaths. The series repeats for about 15 or 20 minutes before returning to normal:
Image

Back to normal again:
Image

And then more weird activity including some CAs with PB before returning to normal.
Image

I've seem this pattern for the last several nights. Not sure why. Any ideas? My CA and OA index for the night remains <1.0 but I would like to better understand those anomalies.

Thanks for your input!

- Flappy

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dreamstation ASV. UPPP in 2007; Untreated AHI 84

Sleeprider
Posts: 1562
Joined: Tue May 06, 2008 5:57 pm
Location: Murrysville, PA

Re: Understanding Flow Rate (ASV)

Post by Sleeprider » Mon Jan 26, 2015 7:11 pm

I get the same thing occasionally. It is periodic breathing or Cheyne Stokes Respirations. You don't want to see a lot of that. It can be a component of central apnea and indicates a waxing and waning of respiratory effort.

Here is my version from last night...not good, but probably SWJ.

Image

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleepyhead software. Just changed from PRS1 BiPAP Auto DS760TS
Last edited by Sleeprider on Mon Jan 26, 2015 8:43 pm, edited 1 time in total.

User avatar
Uncle Flapp
Posts: 247
Joined: Thu Nov 14, 2013 2:24 pm
Location: Arizona

Re: Understanding Flow Rate (ASV)

Post by Uncle Flapp » Mon Jan 26, 2015 8:06 pm

Thanks for sharing, Sleeprider. Makes sense now that you call it out.
My second episode lasted 8 minutes but only the last 3 were flagged by the machine. The first was not flagged at all. Is that normal? I may have to review some old data to see if I can find other examples.

In thinking aloud, I wonder what causes my periodic breathing. I had an EKG a few months ago and everything appeared normal. Perhaps another trip to the cardiologist is in order.

- Flappy

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dreamstation ASV. UPPP in 2007; Untreated AHI 84

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Understanding Flow Rate (ASV)

Post by palerider » Mon Jan 26, 2015 8:37 pm

Uncle Flapp wrote:Perhaps another trip to the cardiologist is in order.
I don't have an opinion on what you've posted, if it's just a few minutes here or there... I dunno...

but this is one view of congestive heart failure:

Image

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Sleeprider
Posts: 1562
Joined: Tue May 06, 2008 5:57 pm
Location: Murrysville, PA

Re: Understanding Flow Rate (ASV)

Post by Sleeprider » Mon Jan 26, 2015 8:42 pm

Really the intermittent stuff like you and I posted is as likely to be 'sleep/wake junk as anything real. If it's not a routine part of your sleep then nothing to worry about. Last night mine was definitely swj and if you look at the time, I was probably half awake. I didn't do myself any favors by drinking a bit more than I should the night before.

Pale, that is a graph I hope never to see first-hand. My dad died of CHF. Had oxygen but no ventilation.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleepyhead software. Just changed from PRS1 BiPAP Auto DS760TS

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Understanding Flow Rate (ASV)

Post by palerider » Mon Jan 26, 2015 8:56 pm

Sleeprider wrote:Pale, that is a graph I hope never to see first-hand. My dad died of CHF. Had oxygen but no ventilation.
that's from a friend of mine, I saw that and my immediate response was "holy shit"... and I started frantically learning as much as I could, decided he needed an asv, stumbled across one on craigslist, bought it, said "here, use this, AND GO TO YOUR DOCTOR AND DEMAND A REFERRAL TO A CARDIO"

he ain't ded yet

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Last edited by palerider on Sun Apr 26, 2015 11:21 am, edited 1 time in total.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Uncle Flapp
Posts: 247
Joined: Thu Nov 14, 2013 2:24 pm
Location: Arizona

Re: Understanding Flow Rate (ASV)

Post by Uncle Flapp » Mon Jan 26, 2015 9:34 pm

Sleeprider wrote:Really the intermittent stuff like you and I posted is as likely to be 'sleep/wake junk as anything real. If it's not a routine part of your sleep then nothing to worry about. Last night mine was definitely swj and if you look at the time, I was probably half awake. I didn't do myself any favors by drinking a bit more than I should the night before.
Could be swj, I dunno... I got more questions than answers. I see similar spells almost nightly but they are usually HYP clusters and don't always include CAs. I previously assumed it was just the PRS1 being overly sensitive.

After seeing Pale Rider's screenshot, I am not going to worry too much and just keep an eye on things. That graph is nasty. PR, best of luck to your friend.

- Flappy

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dreamstation ASV. UPPP in 2007; Untreated AHI 84

User avatar
Uncle Flapp
Posts: 247
Joined: Thu Nov 14, 2013 2:24 pm
Location: Arizona

Re: Understanding Flow Rate (ASV)

Post by Uncle Flapp » Tue Jan 27, 2015 5:48 am

A call from work woke me early so I decided to take a quick look at last night's data. There were two more instances of this weird behavior. The second has no flags at all:
Image
Image

Is that periodic breathing? More swj? A dream where the dog and I are chasing rabbits? My OCD is having a difficult time letting this one go.

- Flappy

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dreamstation ASV. UPPP in 2007; Untreated AHI 84

User avatar
Pesser
Posts: 547
Joined: Sat Jan 11, 2014 4:41 pm
Location: Nova Scotia Canada

Re: Understanding Flow Rate (ASV)

Post by Pesser » Tue Jan 27, 2015 12:42 pm

Uncle Flapp wrote:Things with my sleep have been relatively uneventful (no pun intended) for months but I recently observed a pattern that has me scratching my head. I am hoping y'all can help me understand the data.

The snippet here shows me awake and drifting off:
Image

This is my normal sleep pattern and what I see throughout most of the night. Wave form is pretty regular:
Image

A couple of times per night, some weird stuff happens. This is where I could use some help. I'm not sure what causes this.
I see a series of short breaths followed by large breaths. The series repeats for about 15 or 20 minutes before returning to normal:
Image

Back to normal again:
Image

And then more weird activity including some CAs with PB before returning to normal.
Image

I've seem this pattern for the last several nights. Not sure why. Any ideas? My CA and OA index for the night remains <1.0 but I would like to better understand those anomalies.

Thanks for your input!

- Flappy
I certainly am not qualified to give an opinion on your graphs but I have a question for you. I have seen myself, in my graphs, breathing in and not out. I can see that in yours. Is it possible that you are exhausting air through your mask while you FFM is a little to the side of your face? I use a nasal pillow and I do exhale from my mouth a little. This only happens when I am on my back.

_________________
MachineMaskHumidifier
Additional Comments: Fear of devil with halo

User avatar
Uncle Flapp
Posts: 247
Joined: Thu Nov 14, 2013 2:24 pm
Location: Arizona

Re: Understanding Flow Rate (ASV)

Post by Uncle Flapp » Tue Jan 27, 2015 6:43 pm

Pesser wrote:I certainly am not qualified to give an opinion on your graphs but I have a question for you. I have seen myself, in my graphs, breathing in and not out. I can see that in yours. Is it possible that you are exhausting air through your mask while you FFM is a little to the side of your face? I use a nasal pillow and I do exhale from my mouth a little. This only happens when I am on my back.
Thanks for the reply, Pesser. I'm not an expert either and welcome all opinions. Anything is possible and I am a FFM side sleeper.

For this event:
Image
Sleepy Head reports a gradual leak rate rise from 0L/min to 4L/min and back down again. Total leak ranged from 34L/min to 38L/min and back. Coincidence or not, it does seem to correlate with the event. On the other hand, I did have an equally large leak rate in other areas and the graphs appear normal. Most of the time Leak Rate is 0 and Total Leak well within normal parameters. In the other two graph examples from my earlier post, Leak Rate remained 0 and Total Leak just under 40L/min.

It's a good theory but I am still not sure what to think.

- Flappy

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dreamstation ASV. UPPP in 2007; Untreated AHI 84

-SWS
Posts: 5301
Joined: Tue Jan 11, 2005 7:06 pm

Re: Understanding Flow Rate (ASV)

Post by -SWS » Tue Jan 27, 2015 9:06 pm

Uncle Flapp wrote: I've seem this pattern for the last several nights. Not sure why. Any ideas? My CA and OA index for the night remains <1.0 but I would like to better understand those anomalies.
Sleeprider wrote: Here is my version from last night...not good, but probably SWJ.
Sleeprider is probably right about sleep-wake junk in the flow waveform.
http://www.ncbi.nlm.nih.gov/pubmed/16024516

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: Understanding Flow Rate (ASV)

Post by Jay Aitchsee » Wed Jan 28, 2015 6:50 am

And here's another I found interesting referenced at the address provided by -SWS
http://www.ncbi.nlm.nih.gov/pmc/articles/PMC3378217/

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
Uncle Flapp
Posts: 247
Joined: Thu Nov 14, 2013 2:24 pm
Location: Arizona

Re: Understanding Flow Rate (ASV)

Post by Uncle Flapp » Wed Jan 28, 2015 8:29 am

Thanks for all the responses.

I see a more telling picture when I enable the Timed Breath flag and think this is likely just part of my complex apnea. PTBs range from 50% - 85% on any given night so I usually keep the flag off to avoid the extra clutter.

Image
I suppose it could be sleep-wake junk or, more likely, the reason I am on ASV in the first place.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dreamstation ASV. UPPP in 2007; Untreated AHI 84

Sleeprider
Posts: 1562
Joined: Tue May 06, 2008 5:57 pm
Location: Murrysville, PA

Re: Understanding Flow Rate (ASV)

Post by Sleeprider » Wed Jan 28, 2015 10:15 am

Your machine is capable of dealing with that issue. Do you have timed breaths enabled?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleepyhead software. Just changed from PRS1 BiPAP Auto DS760TS

User avatar
grayghost4
Posts: 1554
Joined: Wed Aug 20, 2014 6:52 pm
Location: Norther Illinois
Contact:

Re: Understanding Flow Rate (ASV)

Post by grayghost4 » Wed Jan 28, 2015 10:32 am

I get patterns like that every night .. first one is usually about 90 min into the night and then they repeat through out the night ..... "REM" sleep.