Do I need to raise my Maximum Pressure (Clusters)

Post by fdw » Tue Jan 13, 2015 3:01 pm

I've been doing OK for a while but recently I've noticed cluster OA's in some of the OA's they've lasted as long as 47 seconds (according to SH). Anyway I now have a Pacemaker/Defibrillator implanted back in mid October 2014 which should not have anything to do with the clusters Im beginning to see, along with the long duration of the OA's (just thought I'd let you know about the implant). That's why I ask about should I increase my pressure.

I attached 2 different graphs one with a large cluster and the other with none but just last nights results.

Thanks for any insight.
https://lh4.googleusercontent.com/-roDw ... 44430.jpeg

https://lh6.googleusercontent.com/-0yeg ... 44353.jpeg
Last edited by fdw on Tue Jan 13, 2015 3:12 pm, edited 3 times in total.
ResMed AirSense S10 Autoset (Ins. Pur. 3/16/2016)
AutoCPAP 10-15 EPR 1.0 (Latest Sleep Study 3-7-16)
Started CPAP March 1995 with a pressure of 11.0

User avatar
Julie
Posts: 19910
Joined: Tue Feb 28, 2006 12:58 pm

Re: Do I need to raise my Maximum Pressure (Clusters)

Post by Julie » Tue Jan 13, 2015 3:07 pm

Your links don't work.

User avatar
fdw
Posts: 260
Joined: Wed Aug 27, 2014 7:08 am
Location: D/FW TEXAS

Re: Do I need to raise my Maximum Pressure (Clusters)

Post by fdw » Tue Jan 13, 2015 3:13 pm

They should work now...Thank you.
ResMed AirSense S10 Autoset (Ins. Pur. 3/16/2016)
AutoCPAP 10-15 EPR 1.0 (Latest Sleep Study 3-7-16)
Started CPAP March 1995 with a pressure of 11.0

User avatar
OkyDoky
Posts: 2870
Joined: Mon Aug 25, 2014 5:18 pm

Re: Do I need to raise my Maximum Pressure (Clusters)

Post by OkyDoky » Tue Jan 13, 2015 3:20 pm

Your numbers aren't bad but you are bumping the top most of the time so it wouldn't hurt to increase top and see what happens. In APAP the machine won't higher if not needed.
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Do I need to raise my Maximum Pressure (Clusters)

Post by Wulfman... » Tue Jan 13, 2015 3:43 pm

Just throwing out some thoughts......

It appears to me that your OA events are occurring when your pressure is at its peak.
Now......which is the chicken and which is the egg? Pressures don't rise to treat apneas when they're in progress.
And, your "events" are almost non-existent when your pressure is lower.
Clusters could indicate that they're positional.
Your Flow Limitations are what's driving your pressures up.
I think it would be interesting if you actually LOWERED your max pressure by one centimeter......for a number of nights.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
OkyDoky
Posts: 2870
Joined: Mon Aug 25, 2014 5:18 pm

Re: Do I need to raise my Maximum Pressure (Clusters)

Post by OkyDoky » Tue Jan 13, 2015 3:49 pm

Den is much more experienced that I am. And that's the nice thing about monitoring Apap with Sleepyhead make a little change for a few days and see what results you get.
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

User avatar
Elle
Posts: 1229
Joined: Fri Nov 17, 2006 9:47 pm
Location: Canada

Re: Do I need to raise my Maximum Pressure (Clusters)

Post by Elle » Tue Jan 13, 2015 4:05 pm

i had started a different thread before seeing this but decreasing my pressure has lowered my AHI.

User avatar
LSAT
Posts: 13232
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Do I need to raise my Maximum Pressure (Clusters)

Post by LSAT » Tue Jan 13, 2015 4:17 pm

fdw wrote:I've been doing OK for a while but recently I've noticed cluster OA's in some of the OA's they've lasted as long as 47 seconds (according to SH). Anyway I now have a Pacemaker/Defibrillator implanted back in mid October 2014 which should not have anything to do with the clusters Im beginning to see, along with the long duration of the OA's (just thought I'd let you know about the implant). That's why I ask about should I increase my pressure.

I attached 2 different graphs one with a large cluster and the other with none but just last nights results.

Thanks for any insight.
https://lh4.googleusercontent.com/-roDw ... 44430.jpeg

https://lh6.googleusercontent.com/-0yeg ... 44353.jpeg

I don't see any reason to change your pressures...Your numbers look great. I don't see anything I would consider a "cluster".

User avatar
fdw
Posts: 260
Joined: Wed Aug 27, 2014 7:08 am
Location: D/FW TEXAS

Re: Do I need to raise my Maximum Pressure (Clusters)

Post by fdw » Tue Jan 13, 2015 4:26 pm

I see what Den is seeing and kinda agree with his assessment regarding lowering the maximum pressure 1cm, in my case from 14.0 to 13.0. My question to Den is my minimum is 12.8 should I lower or keep the same (if I should lower should it be 1cm as well?)

BTW, my last titration was 13.0 several (7) years ago, and I have EAP set at 1.0 therefore the 14.0 Maximum pressure setting.

Thanks
ResMed AirSense S10 Autoset (Ins. Pur. 3/16/2016)
AutoCPAP 10-15 EPR 1.0 (Latest Sleep Study 3-7-16)
Started CPAP March 1995 with a pressure of 11.0

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Do I need to raise my Maximum Pressure (Clusters)

Post by Wulfman... » Tue Jan 13, 2015 4:35 pm

fdw wrote:I see what Den is seeing and kinda agree with his assessment regarding lowering the maximum pressure 1cm, in my case from 14.0 to 13.0. My question to Den is my minimum is 12.8 should I lower or keep the same (if I should lower should it be 1cm as well?)

BTW, my last titration was 13.0 several (7) years ago, and I have EAP set at 1.0 therefore the 14.0 Maximum pressure setting.

Thanks
Your call. I was just throwing out some observations.
You could also try setting the minimum and maximum at 13 .......or 13 and 14 and see what things look like.
I assume you meant EPR? You could try leaving it at 1 or turning it off.
As I've said before, these machine have many variables/options and it's a matter of tweaking to find the "sweet spot".


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
fdw
Posts: 260
Joined: Wed Aug 27, 2014 7:08 am
Location: D/FW TEXAS

Re: Do I need to raise my Maximum Pressure (Clusters)

Post by fdw » Tue Jan 13, 2015 4:41 pm

Wulfman... wrote:
fdw wrote:I see what Den is seeing and kinda agree with his assessment regarding lowering the maximum pressure 1cm, in my case from 14.0 to 13.0. My question to Den is my minimum is 12.8 should I lower or keep the same (if I should lower should it be 1cm as well?)

BTW, my last titration was 13.0 several (7) years ago, and I have EAP set at 1.0 therefore the 14.0 Maximum pressure setting.

Thanks
Your call. I was just throwing out some observations.
You could also try setting the minimum and maximum at 13 .......or 13 and 14 and see what things look like.
I assume you meant EPR? You could try leaving it at 1 or turning it off.
As I've said before, these machine have many variables/options and it's a matter of tweaking to find the "sweet spot".


Den


.
Im lost.....or 13 and 14? Im already at 12.8 minimum and 14.0 maximum.
ResMed AirSense S10 Autoset (Ins. Pur. 3/16/2016)
AutoCPAP 10-15 EPR 1.0 (Latest Sleep Study 3-7-16)
Started CPAP March 1995 with a pressure of 11.0

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Do I need to raise my Maximum Pressure (Clusters)

Post by Wulfman... » Tue Jan 13, 2015 4:44 pm

fdw wrote:
Wulfman... wrote:
fdw wrote:I see what Den is seeing and kinda agree with his assessment regarding lowering the maximum pressure 1cm, in my case from 14.0 to 13.0. My question to Den is my minimum is 12.8 should I lower or keep the same (if I should lower should it be 1cm as well?)

BTW, my last titration was 13.0 several (7) years ago, and I have EAP set at 1.0 therefore the 14.0 Maximum pressure setting.

Thanks
Your call. I was just throwing out some observations.
You could also try setting the minimum and maximum at 13 .......or 13 and 14 and see what things look like.
I assume you meant EPR? You could try leaving it at 1 or turning it off.
As I've said before, these machine have many variables/options and it's a matter of tweaking to find the "sweet spot".


Den


.
Im lost.....or 13 minimum and 13 maximum?
Sorry.......I'm getting myself confused......your max is already at 14......DUH!


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Do I need to raise my Maximum Pressure (Clusters)

Post by Wulfman... » Tue Jan 13, 2015 4:49 pm

Bear with me.......

You might even be able to try 12 min and 12 max with EPR turned off.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
fdw
Posts: 260
Joined: Wed Aug 27, 2014 7:08 am
Location: D/FW TEXAS

Re: Do I need to raise my Maximum Pressure (Clusters)

Post by fdw » Tue Jan 13, 2015 4:56 pm

Wulfman... wrote:Bear with me.......

You might even be able to try 12 min and 12 max with EPR turned off.


Den

.
So your suggesting straight CPAP
ResMed AirSense S10 Autoset (Ins. Pur. 3/16/2016)
AutoCPAP 10-15 EPR 1.0 (Latest Sleep Study 3-7-16)
Started CPAP March 1995 with a pressure of 11.0

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Do I need to raise my Maximum Pressure (Clusters)

Post by Wulfman... » Tue Jan 13, 2015 5:06 pm

fdw wrote:
Wulfman... wrote:Bear with me.......

You might even be able to try 12 min and 12 max with EPR turned off.


Den

.
So your suggesting straight CPAP
Kinda sounds that way, doesn't it. (Yes, that's my suggestion.)
The problem (as I see it) is with your Flow Limitations. They're driving unnecessary pressure increases.

I see lots of spaces in your report where your events are pretty non-existent and your pressures are at their minimum.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05