Reservoir Stuck

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Beeeej
Posts: 7
Joined: Thu Dec 18, 2014 8:06 pm

Reservoir Stuck

Post by Beeeej » Thu Dec 18, 2014 8:12 pm

I've been on my first CPAP for about a week - it's not going very well in general just yet, but that's a story for another day. More serious problem for the moment:

It's a ResMed AirSense 10 Autoset, and it's got a removable reservoir for the humidifier feature. Or at least, it's supposed to be removable. As of last night, it wouldn't come out. I was able to use the machine last night because there was still enough water left in it for six hours, but there's definitely not enough left now - and I still can't get the reservoir out of the machine. Basically, at this point I have no hope of using the CPAP without drying out my throat quite badly - and I'm still in the trial phase where I believe my insurance company wants to be able to verify I've used it for six hours a night or else they won't cover it going forward.

Has anybody else experienced this reservoir jam before? I've tried to search for similar problems in the hopes that someone would have found a solution, but I've come up empty. Really hoping you can help. Thanks.

_________________
Machine

User avatar
Nozzelnut
Posts: 188
Joined: Thu Sep 19, 2013 2:29 pm
Location: Western, NY

Re: Reservoir Stuck

Post by Nozzelnut » Thu Dec 18, 2014 8:56 pm

What about using a rubber/silicone jar opener to get a better grip on the humidifier tub and giving it a solid tug?

Is it a snap in fit or is there some type of release button to remove it?

_________________
Mask: Simplus Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: PRS1 60 series for back-up; Sleepyhead and Rescan 4.3

Beeeej
Posts: 7
Joined: Thu Dec 18, 2014 8:06 pm

Re: Reservoir Stuck

Post by Beeeej » Thu Dec 18, 2014 9:00 pm

Nozzelnut wrote:What about using a rubber/silicone jar opener to get a better grip on the humidifier tub and giving it a solid tug?

Is it a snap in fit or is there some type of release button to remove it?
Pressing on the middle area on the top of the reservoir is supposed to "release" it, as it does have a minimal snap-in kind of mechanism. If you click through to the machine (should be below in this post), you'll see the sixth photo of the machine pretty clearly displays the clear plastic reservoir with the "button" in the middle of the top.

That's why the jar opener suggestion (while appreciated) won't work. It's not at all that I can't get the proper leverage.

_________________
Machine

User avatar
englandsf
Posts: 1016
Joined: Thu Jul 10, 2014 6:17 pm
Location: Austin TX

Re: Reservoir Stuck

Post by englandsf » Thu Dec 18, 2014 9:05 pm

Are the two tank halves locked together correctly? That might prevent the latch from working as per the photos

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Sleepyhead SW. NeilMed and Alkalol Nasal rinses. Veramyst. AutoPAP 11-20 cms. Started June '14, untreated AHI 31-38, with PAP around 1.

Beeeej
Posts: 7
Joined: Thu Dec 18, 2014 8:06 pm

Re: Reservoir Stuck

Post by Beeeej » Thu Dec 18, 2014 9:08 pm

englandsf wrote:Are the two tank halves locked together correctly? That might prevent the latch from working as per the photos
As near as I am able to tell, yes, they are locked together correctly.

_________________
Machine

User avatar
Julie
Posts: 19932
Joined: Tue Feb 28, 2006 12:58 pm

Re: Reservoir Stuck

Post by Julie » Thu Dec 18, 2014 9:34 pm

And I thought mine were too... until water was spilling out and I couldn't get the thing back in the machine... and I found out I hadn't put it together properly (diff. machine to yours, but similar problem). It's sometimes hard to tell (mine clicked, but hadn't seated properly).

User avatar
TangledHose
Posts: 862
Joined: Wed Mar 13, 2013 11:13 am
Location: Centennial, CO

Re: Reservoir Stuck

Post by TangledHose » Thu Dec 18, 2014 11:22 pm

The tub is only held in by small plastic protrusions in the plastic, really not much holding it in, and that's just a traction pad on the top to give you some grip. It's posible that something like maybe water got between the tub and heat plate and as it evaporated it is causing a bit more adhesion. I am using this machine, and if it was me I'd just give it a stronger tug, it should break free. If you get it out make sure the bottom and heat plate are clean and dry before you put it back in.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: Also own F&P Icon Auto for backup and travel; and a Transcend when packing space is limited
Last edited by TangledHose on Fri Dec 19, 2014 12:19 am, edited 2 times in total.
Link to Sleepyhead:
http://sleepfiles.com/SH2/

User avatar
TangledHose
Posts: 862
Joined: Wed Mar 13, 2013 11:13 am
Location: Centennial, CO

Re: Reservoir Stuck

Post by TangledHose » Thu Dec 18, 2014 11:29 pm

There are only two of those small plastic tabs holding it in, they are both on top so you might squeeze the top and bottom halves together while you pull at same time.....the little triangle in this picture, not much too it:

Image

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: Also own F&P Icon Auto for backup and travel; and a Transcend when packing space is limited
Link to Sleepyhead:
http://sleepfiles.com/SH2/

User avatar
TangledHose
Posts: 862
Joined: Wed Mar 13, 2013 11:13 am
Location: Centennial, CO

Re: Reservoir Stuck

Post by TangledHose » Thu Dec 18, 2014 11:51 pm

If you manage to get the tub unstuck it might be worth looking in the machine to make sure the outlet connector is in the proper position as the diagram below shows, I am thinking that possibly it's not secured properly and maybe jamming the tub in place, anyway it's worth checking if you get the tub out before you reinsert it:

Image

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: Also own F&P Icon Auto for backup and travel; and a Transcend when packing space is limited
Link to Sleepyhead:
http://sleepfiles.com/SH2/

User avatar
TangledHose
Posts: 862
Joined: Wed Mar 13, 2013 11:13 am
Location: Centennial, CO

Re: Reservoir Stuck

Post by TangledHose » Fri Dec 19, 2014 10:24 am

I wonder if he got it unstuck? Curiosity is getting to me

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: Also own F&P Icon Auto for backup and travel; and a Transcend when packing space is limited
Link to Sleepyhead:
http://sleepfiles.com/SH2/

Beeeej
Posts: 7
Joined: Thu Dec 18, 2014 8:06 pm

Re: Reservoir Stuck

Post by Beeeej » Fri Dec 19, 2014 10:31 am

TangledHose wrote:I wonder if he got it unstuck? Curiosity is getting to me
I didn't. I'll be calling my sleep clinic today for advice. Thanks for caring!

_________________
Machine

User avatar
TangledHose
Posts: 862
Joined: Wed Mar 13, 2013 11:13 am
Location: Centennial, CO

Re: Reservoir Stuck

Post by TangledHose » Fri Dec 19, 2014 10:48 am

That's too bad.

Well, if you get a chance later shoot us a report here.......you are the first person that has had this problem and we would like to keep track in case this issue pops up again.

One final thought, because the two plastic tabs are on top of the tub the manual does say to gently squeeze the two halves together as you pull it out.........you might try a little downward pressure on the tub as you're pulling since it's only held in at the top.

Good Luck with it.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: Also own F&P Icon Auto for backup and travel; and a Transcend when packing space is limited
Link to Sleepyhead:
http://sleepfiles.com/SH2/

User avatar
SGearhart
Posts: 434
Joined: Fri Nov 07, 2014 8:54 am
Location: Georgia

Re: Reservoir Stuck

Post by SGearhart » Fri Dec 19, 2014 10:54 am

You could always use the Universal Computer/CPAP Repair Tool . . .

Image

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50F & 50IW - SleepyHead 1.0.0-Beta
Last edited by SGearhart on Fri Dec 19, 2014 12:03 pm, edited 1 time in total.

Beeeej
Posts: 7
Joined: Thu Dec 18, 2014 8:06 pm

Re: Reservoir Stuck

Post by Beeeej » Fri Dec 19, 2014 10:57 am

TangledHose wrote:That's too bad.

Well, if you get a chance later shoot us a report here.......you are the first person that has had this problem and we would like to keep track in case this issue pops up again.

One final thought, because the two plastic tabs are on top of the tub the manual does say to gently squeeze the two halves together as you pull it out.........you might try a little downward pressure on the tub as you're pulling since it's only held in at the top.

Good Luck with it.
Thanks, but dude, squeezing and downward pressure are basically all I've been doing for about 48 hours now.

_________________
Machine

User avatar
chunkyfrog
Posts: 34459
Joined: Mon Jul 12, 2010 5:10 pm
Location: In the abyss that is Nebraska--wish me luck!

Re: Reservoir Stuck

Post by chunkyfrog » Fri Dec 19, 2014 11:14 am

I found that I have to squeeze the end of the tank considerably harder than I would expect;
however, it seems to get better with daily operation.
I also need to place my other hand firmly on top of the unit and wiggle the tank firmly.
When re-inserting the tank, the "click" is barely audible and quite dull.
The strength and acute hearing required may be need to be considered against future use of the line.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her