Congestion & Dryness

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
SleepyHead47

Congestion & Dryness

Post by SleepyHead47 » Sat Dec 13, 2014 11:37 am

Been meaning to do this for quite some time. Been on the CPAP for going on 2 years. Had hard time adjusting from the start. Wish I was like some folks I hear about that they take to it and sleep the whole night through with it on. The tech at the med supply business says if I just get 4 hours a night I should not worry about it. But I do, of course! I can do that for awhile then I suddenly either congest up to where I get to coughing or I dry out to the point of my lips and mouth sticking together. I try to pull the nasal mask away from my face and regroup so to speak sometimes that works but a lot of times it doesn't. They have the air setting at 4 and I have the air temp at 67 degrees, and humidity at 3.0. I have adjusted both the air temp and humidity to try to have it at the most comfortable settings but honestly I feel I don't have a clue what I am doing. I am so congested in the mornings, coughing and blowing my nose, it's awful. And sometimes that's only when I've maybe had it on 1-2 hours during the night. Have tried an antihistamine at bedtime but that doesn't seem to matter either. Does anyone have some advice. I will not quit! But surely if some can sleep all night with it I can't see any reason why I can't. And I do know how that it benefits my health!!!

User avatar
Pugsy
Posts: 64012
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Congestion & Dryness

Post by Pugsy » Sat Dec 13, 2014 11:43 am

SleepyHead47 wrote: They have the air setting at 4
I assume that is the pressure? Is this ramp or auto minimum or what?
SleepyHead47 wrote:air temp at 67 degrees, and humidity at 3.0
Sounds like heated hose...possibly you have a Resmed machine of some sort?
It would be really helpful if we knew exactly which machine and equipment you are using
SleepyHead47 wrote:I am so congested in the mornings, coughing and blowing my nose, it's awful.
Sounds like it could be humidity related....you say you have adjusted the humidity levels...what have you tried?
Some people need more humidity and some people need less to none humidity. There is no right or wrong..there's just what works for the individual and individual needs and preferences can vary widely.
A setting of 3 is pretty much middle of the road...which way have you tried?

Also what mask are you using?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
Julie
Posts: 19908
Joined: Tue Feb 28, 2006 12:58 pm

Re: Congestion & Dryness

Post by Julie » Sat Dec 13, 2014 12:10 pm

Unless you're in a very dry climate, try turning down, or even off, the humidity... it can definitely make you congested.

User avatar
mike1953
Posts: 154
Joined: Sun Aug 31, 2014 7:15 pm
Location: Kansas

Re: Congestion & Dryness

Post by mike1953 » Sat Dec 13, 2014 7:23 pm

Adjusting your humidifier up or down until you are comfortable. That is what I have done. I do not use heated hose and started out at 3 went to 3.5 then down to 2.5 which for me is the best.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Using a Buckwheat pillow its great

User avatar
chunkyfrog
Posts: 34390
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: Congestion & Dryness

Post by chunkyfrog » Sat Dec 13, 2014 7:44 pm

Picture yourself as Goldilocks. Feel entitled to adjust the temperature and humidity (separately) until it's just right.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

Sleeprider
Posts: 1562
Joined: Tue May 06, 2008 5:57 pm
Location: Murrysville, PA

Re: Congestion & Dryness

Post by Sleeprider » Sat Dec 13, 2014 9:27 pm

You say the air setting is at 4. While it would be helpful to know exactly what you are using in equipment, if your pressure setting is really fixed at 4 cmH2O, then that would suffocate most of us. That is the minimum pressure setting for most CPAP machines, and it will make it very difficult to move enough volume to feel comfortable, let alone treat apnea.

Fill out your profile on equipment and post some claification on your settings. Good luck!

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleepyhead software. Just changed from PRS1 BiPAP Auto DS760TS

SleepyHead47

Re: Congestion & Dryness

Post by SleepyHead47 » Sun Dec 14, 2014 4:23 pm

Thanks so much for all the replies, input and questions!!! Experience of others can be very helpful! Sorry, I didn't put more information into my original post. You are right, Pugsy, am using a ResMed S9 with the H5i heated humidifier. Tried a couple different masks but the best seemed to be the ResMed Swift FX Nano For Her nasal mask. I soon realized that I also needed to purchase a chin strap since I was air tunneling with the CPAP. Talk about getting dried out. Yes, they have the air setting at 4 and the ramp at 45 minutes. We reside in Iowa so there is plenty of humidity. And, the med supply tech did say the settings should be adjusted for different times of years like lower humidity in the winter while our furnace runs, and a little higher for in the summer while the AC is taking the humidity out of the air. I tried lowering AGAIN after reading the suggestions to lower the humidity level. At 2.5 I DID sleep longer with it on probably about 5 hours which is good for me. I have adjusted the temp from 65 to 70. 70 being way too warm. On the humidity I have adjusted it from 2 to 4. I have never tried turning the humidity to 0. Julia, I just might try turning it off to see how that feels. And you are right Mike1953 it's having patience with it and adjusting it to a person's own comfort. I'd like to just put it on at bedtime and take it off when I get up in the morning. The tech suggested setting the humidity level at 4 when I was drying out but I wake up with it dripping from the mask. You are right, too, chunkyfrog, I'll pretend I'm Goldilocks and be gutsy about my comfort!!! Good analogy!!! Thanks so much, everyone for your input!

User avatar
Tatooed Lady
Posts: 984
Joined: Sat Jun 07, 2014 6:18 pm
Location: Central Wisconsin

Re: Congestion & Dryness

Post by Tatooed Lady » Sun Dec 14, 2014 4:40 pm

SleepyHead47 wrote:Thanks so much for all the replies, input and questions!!! Experience of others can be very helpful! Sorry, I didn't put more information into my original post. You are right, Pugsy, am using a ResMed S9 with the H5i heated humidifier. Tried a couple different masks but the best seemed to be the ResMed Swift FX Nano For Her nasal mask. I soon realized that I also needed to purchase a chin strap since I was air tunneling with the CPAP. Talk about getting dried out. Yes, they have the air setting at 4 and the ramp at 45 minutes. We reside in Iowa so there is plenty of humidity. And, the med supply tech did say the settings should be adjusted for different times of years like lower humidity in the winter while our furnace runs, and a little higher for in the summer while the AC is taking the humidity out of the air. I tried lowering AGAIN after reading the suggestions to lower the humidity level. At 2.5 I DID sleep longer with it on probably about 5 hours which is good for me. I have adjusted the temp from 65 to 70. 70 being way too warm. On the humidity I have adjusted it from 2 to 4. I have never tried turning the humidity to 0. Julia, I just might try turning it off to see how that feels. And you are right Mike1953 it's having patience with it and adjusting it to a person's own comfort. I'd like to just put it on at bedtime and take it off when I get up in the morning. The tech suggested setting the humidity level at 4 when I was drying out but I wake up with it dripping from the mask. You are right, too, chunkyfrog, I'll pretend I'm Goldilocks and be gutsy about my comfort!!! Good analogy!!! Thanks so much, everyone for your input!
Using a ramp and having the pressure set at 4 is counterproductive. The ramp feature starts low (example 4cm), and ramps UP to higher pressures (some people use 16+ cm of pressure), to give the wearer time to acclimate to the higher pressures while falling asleep. So...if your low end (assuming an auto machine) is 4 cm to START with, the ramp isn't going to do anything. You will start at your base set pressure. My machine was set to start at 4cm as well when I got it. I got in an argument with the DME about upping the pressure so I could breathe more easily. You might be having the same type of issue...not uncommon.
Iowa has humidity, yes, HOWEVER...so does Wisconsin, usually. But as soon as you turn on the furnace or start using the wood burner, etc...humidity in the house plummets. You might get a cheap thermometer that also shows humidity, and use it in the bedroom so you can get an idea on how dry the air REALLY is. It could be that you would benefit from a small humidifier over the winter months.
With the nasal type masks, you can also use a small travel size pillow tucked under your chin to keep your mouth shut. I use one sometimes, plus it's a snuggle buddy.

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Precious and POW are very very good to me.
As Bette Davis famously said, “Old age ain’t for sissies.”
I'm with the band.
So.Many.TOYS!

tmarshl
Posts: 4
Joined: Sat Feb 11, 2006 12:42 pm

Re: Congestion & Dryness

Post by tmarshl » Mon Dec 15, 2014 9:14 am

I had similar problems, caused primarily by mouth breathing (even with a full-face mask). I solved that each night by using two strips of 3M Transpore surgical tape vertically to tape my mouth closed. Works like a charm, where chin strap failed.

_________________
Mask
Additional Comments: Pressure Titration 11

bjl
Posts: 26
Joined: Sat Oct 11, 2014 2:38 pm

Re: Congestion & Dryness

Post by bjl » Mon Dec 15, 2014 3:02 pm

I would definitely recommend experimenting with the humidity set to 0. I have a deviated septum and a generally snuffy nose. I was also having problems with congestion when I had the humidity set to 2. So I figured I needed more humidity because that's good for your sinuses right? Well, upping the humidity made things worse. On the advice of some folks here I turned off the humidity entirely and that made a world of difference. I sometimes set the humidity to 1 on very dry nights, but keep it at 0 most of the time.

Hope this helps.

-b

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sometimes use Quattro Air FFM, but pretty much sticking with the Swift FX at this point

User avatar
tireddaisey
Posts: 27
Joined: Sun Nov 23, 2014 7:12 pm

Re: Congestion & Dryness

Post by tireddaisey » Sun Dec 21, 2014 6:55 pm

Sorry to jump on this thread (it does relate) but the "new topic" function does not work for me on the updated mobile app.

Been on cpap for about 5 weeks now and experiencing major sneezing episodes and stuffiness. My house is generally quite dry and I've been comfortable at humidity lvl 3 or 4. But could the continued high lvl cause sustained sneezing, dizziness from the congestion, etc.? Should I lower the humidity lvl? And being on the nasal pillows would getting a backup ffm be advised for the drier months, will insurance consider 2 different masks?

Thanks!

User avatar
Julie
Posts: 19908
Joined: Tue Feb 28, 2006 12:58 pm

Re: Congestion & Dryness

Post by Julie » Sun Dec 21, 2014 7:17 pm

You have to try other settings - only you can judge what works there. And you should call your insurance co. and ask what they cover...

User avatar
tireddaisey
Posts: 27
Joined: Sun Nov 23, 2014 7:12 pm

Re: Congestion & Dryness

Post by tireddaisey » Mon Dec 22, 2014 8:57 am

Thanks Julie! Turned down humidity last night and slept well. Will see if decreased sneeziness happens today and will check in with insurance.