Up loading new software

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Broken Bob
Posts: 42
Joined: Sat Jan 21, 2012 9:25 am
Location: louisville ky

Up loading new software

Post by Broken Bob » Thu Nov 06, 2014 6:07 am

I have been plugging along, fat, dumb and happy, downloading my machine to 0.9.3 software that I installed way back when I first learned of Sleepyhead and this chat room. I have only recently returned to the chat room after a long absence. I saw that Sleepy head is now at 0.9.8.1. I tried downloading this without success. Will I need to load previous upgrades first? Will I be able to overwrite the 9.3 and retain my history? I am using Windows 7 Home Premium. Before you all scold me to much, I have updated my profile to be notified of future upgrades. Thanks in advance for the responses I know will come. This is a great site!

_________________
Mask

User avatar
Pugsy
Posts: 64012
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Up loading new software

Post by Pugsy » Thu Nov 06, 2014 9:17 am

Version 0.9.8-1 won't change 0.9.3 at all. It won't overwrite anything because it is a totally different version.

You don't need prior versions.

0.9.8-1 doesn't have an installer though. So you can't do a regular install like you did with 0.9.3.
Instead you have to unzip and then go into the unzipped folder and find the application.exe file (it's either the last one or the next to the last one in the list) and click on it to open SleepyHead. I usually suggest to people that while there that they right click their mouse on that file and create a shortcut so they don't have to go into the folder each time they want to open SleepyHead.

Oh....you need a different zip/unzip program called 7 zip to unzip the newest SH version.
It's mentioned at where you get the download.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Up loading new software

Post by palerider » Thu Nov 06, 2014 9:21 am

Broken Bob wrote:I have been plugging along, fat, dumb and happy, downloading my machine to 0.9.3 software that I installed way back when I first learned of Sleepyhead and this chat room. I have only recently returned to the chat room after a long absence. I saw that Sleepy head is now at 0.9.8.1. I tried downloading this without success. Will I need to load previous upgrades first? Will I be able to overwrite the 9.3 and retain my history? I am using Windows 7 Home Premium. Before you all scold me to much, I have updated my profile to be notified of future upgrades. Thanks in advance for the responses I know will come. This is a great site!
your card should have data for at least the last year to be imported into SH.

you can try getting 0.9.6 from sleepfiles.com/SH and upgrading to that, then upgrading to 0.9.8-1 if you want older history.

when you say you tried "downloading without success" what does that mean?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Broken Bob
Posts: 42
Joined: Sat Jan 21, 2012 9:25 am
Location: louisville ky

Re: Up loading new software

Post by Broken Bob » Thu Nov 06, 2014 9:52 am

Pugsy
Thanks for your reply-I will give it a try.
Palerider-The 9.8 downloaded to my desktop but I could not find the exe. file or any file to install. Hopefully with Pugsy's information I will be successful.

_________________
Mask

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Up loading new software

Post by palerider » Thu Nov 06, 2014 10:00 am

Broken Bob wrote:Pugsy
Thanks for your reply-I will give it a try.
Palerider-The 9.8 downloaded to my desktop but I could not find the exe. file or any file to install. Hopefully with Pugsy's information I will be successful.
it's distributed as a 7zip archive (ends in 7z) because it's a good bit smaller than a zip file... so it saves on mark's upload time, and your download time

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

kwikwater
Posts: 183
Joined: Fri Jun 12, 2009 8:32 pm
Location: NW Colorado 50 miles from nowhere
Contact:

Re: Up loading new software

Post by kwikwater » Tue Nov 18, 2014 6:38 pm

Thanks for the tip Pugsy!! I was about to pull my hair out!

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Swift™ LT Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: PulseOx CMS-50D+, Papcap, SleepyHead

User avatar
saltydawg2
Posts: 48
Joined: Thu Sep 18, 2014 5:03 pm

Re: Up loading new software

Post by saltydawg2 » Sun Jan 25, 2015 7:29 pm

Only post this guy has made is to hawk an app, and his little dog toto, too!

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pressure 11.2 min. to 14max EPR 1

User avatar
Julie
Posts: 19908
Joined: Tue Feb 28, 2006 12:58 pm

Re: Up loading new software

Post by Julie » Mon Feb 02, 2015 7:13 am

Read the first note on the main page (... "Pugsy's Pointers - Tutorial").

User avatar
sthnreb
Posts: 614
Joined: Wed Feb 09, 2005 9:28 am
Location: Metro Atlanta

Re: Up loading new software

Post by sthnreb » Mon Feb 02, 2015 8:44 am

I get "The program can't start because Qt5Core.dll is missing from your computer. Try reinstalling the program to fix this problem." So, what is my problem?

_________________
MachineMask
Additional Comments: Resmed AirCurve 10 VAuto BiLevel with HumidAir
Bi-Pap for 17 years now. Rx 12/8 and using a Resmed AirCurve 10 SAuto Bipap Auto.

User avatar
Pugsy
Posts: 64012
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Up loading new software

Post by Pugsy » Mon Feb 02, 2015 8:50 am

sthnreb wrote:I get "The program can't start because Qt5Core.dll is missing from your computer. Try reinstalling the program to fix this problem." So, what is my problem?
Exactly which version are you trying to install?
Did you get it unzipped? It has to be unzipped with 7Zip as the regular Windows unzipper won't do it. Then you click on the application.exe file after things are unzipped.
What operating system does your computer have and is it 32 or 64 bit?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
Last edited by Pugsy on Mon Feb 02, 2015 12:52 pm, edited 1 time in total.
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
squid13
Posts: 2299
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: Up loading new software

Post by squid13 » Mon Feb 02, 2015 9:19 am

sthnreb wrote:I get "The program can't start because Qt5Core.dll is missing from your computer. Try reinstalling the program to fix this problem." So, what is my problem?
See if this will help http://www.wikifixes.com/errors/dll/qt5 ... aQodCIkAEg

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV

User avatar
OkyDoky
Posts: 2870
Joined: Mon Aug 25, 2014 5:18 pm

Re: Up loading new software

Post by OkyDoky » Mon Feb 02, 2015 9:19 am

sthnreb wrote:I get "The program can't start because Qt5Core.dll is missing from your computer. Try reinstalling the program to fix this problem." So, what is my problem?
You may need this windows download it was needed for my computer. http://www.microsoft.com/en-gb/download ... x?id=40784
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Up loading new software

Post by palerider » Mon Feb 02, 2015 12:46 pm

squid13 wrote:
sthnreb wrote:I get "The program can't start because Qt5Core.dll is missing from your computer. Try reinstalling the program to fix this problem." So, what is my problem?
See if this will help http://www.wikifixes.com/errors/dll/qt5 ... aQodCIkAEg
I think his error is usually caused by trying to run things out of the zip file without extracting it first.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
OkyDoky
Posts: 2870
Joined: Mon Aug 25, 2014 5:18 pm

Re: Up loading new software

Post by OkyDoky » Mon Feb 02, 2015 12:57 pm

palerider wrote:
squid13 wrote:
sthnreb wrote:I get "The program can't start because Qt5Core.dll is missing from your computer. Try reinstalling the program to fix this problem." So, what is my problem?
See if this will help http://www.wikifixes.com/errors/dll/qt5 ... aQodCIkAEg
I think his error is usually caused by trying to run things out of the zip file without extracting it first.
Right, that was the first error I kept getting when I kept trying to open in the zip file. I could see it was in the file and I was stumped. But a helpful soul here pointed out that it had to be extracted to a different file before you could open it. Once I opened the extracted file that's when I found out I needed the other file downloaded from windows. Finally all was running fine.
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

User avatar
sthnreb
Posts: 614
Joined: Wed Feb 09, 2005 9:28 am
Location: Metro Atlanta

Re: Up loading new software

Post by sthnreb » Mon Feb 02, 2015 9:47 pm

"Exactly which version are you trying to install?" I have Windows 7 Ultimate and I was loading the newer beta version for the 64 bit system. It said for the AirCurve 10 VAuto machine. I did download the zipper you mentioned. It appears somewhere in the downloading I got the file Talika search engine which installed a lot of junk. I did get the file with the zip mentioned and got it loaded. However on a reboot, my screen would go black. I had to go back to safe mode and go in and remove all the junk which still didn't fix it. So, I had to do a system restore going back to Jan. 30, 2015. Don't suppose I lost to much but I did lose the SleepyHead software. It then booted ok. I went in and removed all the junk files loaded. Now, maybe I will go for the Resmed 5.3? Can you run the SD card info in it? Thanks for the responses. I wasn't able to read them until now.

_________________
MachineMask
Additional Comments: Resmed AirCurve 10 VAuto BiLevel with HumidAir
Bi-Pap for 17 years now. Rx 12/8 and using a Resmed AirCurve 10 SAuto Bipap Auto.