SSSSSS
SSSSSS
I kept being awakened last night by a hissing sound. Every time I went back to sleep it would start again. I am a nose breather, but why is air forcing itself out of my closed mouth? I recently upped my low end pressure from 5 to 7; should I take it back down?
Last edited by Noctuary on Sun Nov 02, 2014 4:16 pm, edited 2 times in total.
My machine: ResMed S9 Auto/H5i humidifier; in APAP mode- pressure range 8-13
Mask: Flexifit 407
Mask: Flexifit 407
Re: SSSSSS
What kind of mask are you using - FF or nasal?
When you raise the pressure, more air will be forced out of whatever opening is handy, and if that's your mouth, higher pressure will exacerbate things. Lowering your pressure is not the answer unless you not only feel fine at 5, but your AHI and other factors are also good. You may want to look at a FF mask instead.
When you raise the pressure, more air will be forced out of whatever opening is handy, and if that's your mouth, higher pressure will exacerbate things. Lowering your pressure is not the answer unless you not only feel fine at 5, but your AHI and other factors are also good. You may want to look at a FF mask instead.
_________________
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear |
Humidifier: IntelliPAP Integrated Heated Humidifier |
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: SSSSSS
You need to keep your mouth SHUT to keep it from leaking out. The pressures you're using are practically nothing.Noctuary wrote:I kept being awoken last night by a hissing sound. Everytime I went back to sleep it would start again. I am a nose breather, but why is air forcing itself out of my closed mouth? I recently upped my low end pressure from 5 to 7; should I take it back down?
Since you have nothing in your profile, we don't know if your machine is an Auto and you're using a range of pressures or what kind of mask you're using.......or anything else. It's hard to "guess" at what answer may help.
You mention your "low end pressure" but what's the maximum pressure? What do your software reports show?
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: SSSSSS
am I the only one that can open their mouth and stick their tongue out while using a pillow mask and have no air coming out my mouth?Wulfman... wrote:You need to keep your mouth SHUT to keep it from leaking out. The pressures you're using are practically nothing.Noctuary wrote:I kept being awoken last night by a hissing sound. Everytime I went back to sleep it would start again. I am a nose breather, but why is air forcing itself out of my closed mouth? I recently upped my low end pressure from 5 to 7; should I take it back down?
Since you have nothing in your profile, we don't know if your machine is an Auto and you're using a range of pressures or what kind of mask you're using.......or anything else. It's hard to "guess" at what answer may help.
You mention your "low end pressure" but what's the maximum pressure? What do your software reports show?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: SSSSSS
The tongue is relaxing and not blocking the air that is going up the noise and down the airway.
When the tongue relaxes the air that is in the airway will try to exit out the first path it finds and that's into the mouth. If lips are closed we normally end up with chipmunk cheeks which eventually will force the lips to part...either a lot and cause a torrent of air out the mouth or maybe just a little parting and it's a slow leak or hsssss.
The tongue needs to do a better job of staying in place to keep the door closed. Easy to do while awake but when we go to sleep the tongue falls down on the job. The tongue needs to be retrained to stay in position and for some it can be done but for others the tongue wins.
Play with the tongue placement a bit while awake with the machine and mask on and you will see what I mean.
Normally if the tongue is placed (and stays there) in the roof of the mouth no air will enter the mouth so it won't ever try to leak out the lips.
No palerider...you aren't the only one. I can yawn, talk, open my mouth all I want with mask and machine on...and never have any air try to exit through the mouth. I can even stick out my tongue without the air exiting...there's something that I do involuntary that still keeps that back door closed. I can't describe what I do though.
When the tongue relaxes the air that is in the airway will try to exit out the first path it finds and that's into the mouth. If lips are closed we normally end up with chipmunk cheeks which eventually will force the lips to part...either a lot and cause a torrent of air out the mouth or maybe just a little parting and it's a slow leak or hsssss.
The tongue needs to do a better job of staying in place to keep the door closed. Easy to do while awake but when we go to sleep the tongue falls down on the job. The tongue needs to be retrained to stay in position and for some it can be done but for others the tongue wins.
Play with the tongue placement a bit while awake with the machine and mask on and you will see what I mean.
Normally if the tongue is placed (and stays there) in the roof of the mouth no air will enter the mouth so it won't ever try to leak out the lips.
No palerider...you aren't the only one. I can yawn, talk, open my mouth all I want with mask and machine on...and never have any air try to exit through the mouth. I can even stick out my tongue without the air exiting...there's something that I do involuntary that still keeps that back door closed. I can't describe what I do though.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
Last edited by Pugsy on Sun Nov 02, 2014 2:29 pm, edited 1 time in total.
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: SSSSSS
We really don't know if you're "anatomically correct".......palerider wrote:am I the only one that can open their mouth and stick their tongue out while using a pillow mask and have no air coming out my mouth?Wulfman... wrote:You need to keep your mouth SHUT to keep it from leaking out. The pressures you're using are practically nothing.Noctuary wrote:I kept being awoken last night by a hissing sound. Everytime I went back to sleep it would start again. I am a nose breather, but why is air forcing itself out of my closed mouth? I recently upped my low end pressure from 5 to 7; should I take it back down?
Since you have nothing in your profile, we don't know if your machine is an Auto and you're using a range of pressures or what kind of mask you're using.......or anything else. It's hard to "guess" at what answer may help.
You mention your "low end pressure" but what's the maximum pressure? What do your software reports show?
I've never tried it because I've never tried to use anything other than a full face mask.
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: SSSSSS
I use a nasal mask. My unit is set to APAP, pressure 7-20. Weird though; I feel better today than I have in a long time. Just the slightest fatigue.
My machine: ResMed S9 Auto/H5i humidifier; in APAP mode- pressure range 8-13
Mask: Flexifit 407
Mask: Flexifit 407
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: SSSSSS
Are you using software to monitor your therapy? Using an APAP in a range of pressures without software is like "flying blind".Noctuary wrote:I use a nasal mask. My unit is set to APAP, pressure 7-20. Weird though; I feel better today than I have in a long time. Just the slightest fatigue.
Edit: The machine you just selected for your profile is an S9 Escape.......not data and no pressure range. Maybe try again.
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: SSSSSS
well, at least I'm in good company with pugsyWulfman... wrote:We really don't know if you're "anatomically correct"....... .palerider wrote:am I the only one that can open their mouth and stick their tongue out while using a pillow mask and have no air coming out my mouth?
I've done the same things as she indicated...
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- chunkyfrog
- Posts: 34416
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: SSSSSS
Pale rider, it's rude to stick your tongue out.
--just saying.
--just saying.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: SSSSSS
No software; I tried it with the Sleep Head, but it just crashes. I updated my machine; its a ResMed Escape Auto.Wulfman... wrote:Are you using software to monitor your therapy? Using an APAP in a range of pressures without software is like "flying blind".Noctuary wrote:I use a nasal mask. My unit is set to APAP, pressure 7-20. Weird though; I feel better today than I have in a long time. Just the slightest fatigue.
Edit: The machine you just selected for your profile is an S9 Escape.......not data and no pressure range. Maybe try again.
Den
.
My machine: ResMed S9 Auto/H5i humidifier; in APAP mode- pressure range 8-13
Mask: Flexifit 407
Mask: Flexifit 407
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: SSSSSS
Except the machine in your profile is the S9 Autoset. Still not the right one if you have the Escape Auto.Noctuary wrote:No software; I tried it with the Sleep Head, but it just crashes. I updated my machine; its a ResMed Escape Auto.Wulfman... wrote:Are you using software to monitor your therapy? Using an APAP in a range of pressures without software is like "flying blind".Noctuary wrote:I use a nasal mask. My unit is set to APAP, pressure 7-20. Weird though; I feel better today than I have in a long time. Just the slightest fatigue.
Edit: The machine you just selected for your profile is an S9 Escape.......not data and no pressure range. Maybe try again.
Den
.
Problem is, the Escape Auto is extremely limited in the data it will provide. I think about all it gives is AHI and hours used. No leak data or other pertinent information.
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: SSSSSS
the escape auto collects almost no data and is not currently supported by sleepyhead as of 0.9.8-1Noctuary wrote: No software; I tried it with the Sleep Head, but it just crashes. I updated my machine; its a ResMed Escape Auto.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: SSSSSS
Then I don't know what else to do.
My machine: ResMed S9 Auto/H5i humidifier; in APAP mode- pressure range 8-13
Mask: Flexifit 407
Mask: Flexifit 407
Re: SSSSSS
Same here, except I haven't tried the talking part. Retraining since on therapy, since it drove me crazy to open my mouth and have the windstorm. Man, was that disconcerting!No palerider...you aren't the only one. I can yawn, talk, open my mouth all I want with mask and machine on...and never have any air try to exit through the mouth. I can even stick out my tongue without the air exiting...there's something that I do involuntary that still keeps that back door closed. I can't describe what I do though.