Can't stop air escaping from my mouth

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
VictorC
Posts: 62
Joined: Thu Oct 16, 2014 3:54 pm
Location: Belleville, IL

Can't stop air escaping from my mouth

Post by VictorC » Fri Dec 31, 2021 9:46 am

Hello everyone! I have been using my cpap successfully for years now. I have always been majorly slack-jawed when I sleep and always use a chin strap. In the past month the combination of my P10 and a chin strap has stopped working for me. I keep opening my mouth and letting air escape no matter how tight I cinch the chin strap down. Since the chin strap stopped working I tried switching to mouth taping. The tape prevented me from opening my mouth, but air was still escaping from the corners of my mouth.

I don't know what to do or why I suddenly have this problem. The only thing that has changed is that I have lost some weight recently: approximately 20 pounds. I don't know why that would make a difference. I'm including my therapy details below. I appreciate any suggestions or insight you might have.

Machine: Resmed Airsense 10
Mask: Resmed P10
Chin Strap: Phillips Respironics Premium
Pressure: 10-14 cm/h2o
EPR: 2

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Settings: APAP at 10.5-14cm H2O, AFLEX @ 2, Humidifier @ 2, Hose Temperature @ 3, No Ramp

User avatar
ChicagoGranny
Posts: 14569
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Can't stop air escaping from my mouth

Post by ChicagoGranny » Fri Dec 31, 2021 11:05 am

It's common for chin straps to be ineffective.
VictorC wrote:
Fri Dec 31, 2021 9:46 am
The tape prevented me from opening my mouth, but air was still escaping from the corners of my mouth.
Which type and width of tape are you using? Are you cutting it long enough to extend 0.5 inches beyond each corner of your lips?

One tip that helps me a lot is to turn the lips in slightly and then place the tape.

User avatar
zonker
Posts: 11048
Joined: Fri Jun 19, 2015 4:36 pm

Re: Can't stop air escaping from my mouth

Post by zonker » Fri Dec 31, 2021 11:44 am

VictorC wrote:
Fri Dec 31, 2021 9:46 am
Hello everyone!
howdy!

here's what i use instead of chinstrap and tape-

https://www.cpaptalk.com/viewtopic.php? ... s#p1086296

i have modified it a bit for my skull shape and to make it tighter. this plus a firm foam cervical collar, which is found in your local drugstore, has eliminated mouth leaks for me.

look for the collar in the section where they keep bedpans or crutches or walkers; things like that.

good luck!
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

VictorC
Posts: 62
Joined: Thu Oct 16, 2014 3:54 pm
Location: Belleville, IL

Re: Can't stop air escaping from my mouth

Post by VictorC » Fri Dec 31, 2021 11:50 am

ChicagoGranny wrote:
Fri Dec 31, 2021 11:05 am
Which type and width of tape are you using? Are you cutting it long enough to extend 0.5 inches beyond each corner of your lips?

One tip that helps me a lot is to turn the lips in slightly and then place the tape.
I think the tape is long enough, though I could try making it a little longer. I use 2" wide micropore tape. It's wide enough that I can just tape over my whole mouth. I'll try turning in my lips and see if that helps. Thanks!

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Settings: APAP at 10.5-14cm H2O, AFLEX @ 2, Humidifier @ 2, Hose Temperature @ 3, No Ramp

User avatar
LSAT
Posts: 13238
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Can't stop air escaping from my mouth

Post by LSAT » Fri Dec 31, 2021 11:52 am

Learn to keep your tongue at the roof of your mouth behind your front teeth. You will not lose any air out your mouth...even if you open your mouth a little.

VictorC
Posts: 62
Joined: Thu Oct 16, 2014 3:54 pm
Location: Belleville, IL

Re: Can't stop air escaping from my mouth

Post by VictorC » Fri Dec 31, 2021 11:55 am

zonker wrote:
Fri Dec 31, 2021 11:44 am
here's what i use instead of chinstrap and tape-

https://www.cpaptalk.com/viewtopic.php? ... s#p1086296

i have modified it a bit for my skull shape and to make it tighter. this plus a firm foam cervical collar, which is found in your local drugstore, has eliminated mouth leaks for me.

look for the collar in the section where they keep bedpans or crutches or walkers; things like that.
I'll have to try that if I can't get anything else to work. Between the chin strap and the mask I already feel like I have to prepare for bed by performing some rigorous ritual. I can't imagine what it would be like to add a cervical collar to the mix. Still if that's what it takes... But I might hold off on that as a last resort. Thanks!

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Settings: APAP at 10.5-14cm H2O, AFLEX @ 2, Humidifier @ 2, Hose Temperature @ 3, No Ramp

VictorC
Posts: 62
Joined: Thu Oct 16, 2014 3:54 pm
Location: Belleville, IL

Re: Can't stop air escaping from my mouth

Post by VictorC » Fri Dec 31, 2021 11:57 am

LSAT wrote:
Fri Dec 31, 2021 11:52 am
Learn to keep your tongue at the roof of your mouth behind your front teeth. You will not lose any air out your mouth...even if you open your mouth a little.
Hi! Thanks! I did this for years and it worked. It's only recently that this technique has stopped working for me.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Settings: APAP at 10.5-14cm H2O, AFLEX @ 2, Humidifier @ 2, Hose Temperature @ 3, No Ramp

User avatar
zonker
Posts: 11048
Joined: Fri Jun 19, 2015 4:36 pm

Re: Can't stop air escaping from my mouth

Post by zonker » Fri Dec 31, 2021 12:00 pm

VictorC wrote:
Fri Dec 31, 2021 11:55 am

I'll have to try that if I can't get anything else to work. Between the chin strap and the mask I already feel like I have to prepare for bed by performing some rigorous ritual. I can't imagine what it would be like to add a cervical collar to the mix. Still if that's what it takes... But I might hold off on that as a last resort. Thanks!
:lol: :lol: :lol: i know;i feel like an astronaut getting suited up for a spacewalk.

here's an early pic of me in my rig-

Image

i've since added a bit of "drool catcher" and an extra band around the top of my gob. it looks as dorky as hell, but no leaks and it helped drop my ahi dramatically.
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Jlfinkels
Posts: 468
Joined: Sun Jun 20, 2021 3:51 pm

Re: Can't stop air escaping from my mouth

Post by Jlfinkels » Fri Dec 31, 2021 12:45 pm

Here’s the important thing about air escaping through your mouth, you can likely stop it from happening with practice. When I started I tried to use nasal pillows (P10) and after a few days of frustration with little sleep I swore I could never do it. I almost packed it all in, but decided to post here for advice. As a result I started to work on it a bit at a time as follows:
  • Switched to a full face mask so I could continue therapy (F20)
  • Started to practice with P10 while awake watching TV, reading, practicing accordion (don’t judge)
  • Took naps with P10 and realized I needed more help so bought some chinstraps
  • Hated the chinstraps and Zonker pointed me to cervical collars
  • Added a collar and tried napping again - it worked for short intervals
  • After 6 months of practice switched over to the nasal pillows - et voila!
The key thing was not giving up or giving in to frustration, practicing often, and not being too hard on myself for having problems with it. As I could use the F20 fairly successfully I knew I had a failsafe when I had issue with the pillows (and I did quite often so would switch mid sleep to the F20), but I kept at it. I think you and others can do it too, just take your time and forgive yourself when you have problems.
Sometimes it is the very people who no one imagines anything of who do the things that no one imagines

User avatar
Pugsy
Posts: 64168
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Can't stop air escaping from my mouth

Post by Pugsy » Fri Dec 31, 2021 1:40 pm

If you are otherwise okay with taping just use a bigger piece of tape to cover the mouth better so the corners don't come loose.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

VictorC
Posts: 62
Joined: Thu Oct 16, 2014 3:54 pm
Location: Belleville, IL

Re: Can't stop air escaping from my mouth

Post by VictorC » Sat Jan 01, 2022 10:13 am

I tried a bigger piece of tape last night, and I think it held better. Now the big issue is that air escaping tends to be loud and wakes my wife up. She in turn wakes me up. I did finally put an SD card in my machine so I have data now. I see several leaks throughout the night. They seem to last about 15 minutes apiece and then resolve.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Settings: APAP at 10.5-14cm H2O, AFLEX @ 2, Humidifier @ 2, Hose Temperature @ 3, No Ramp

jeepdoctor
Posts: 173
Joined: Sun Jul 09, 2006 5:23 pm
Location: "Green Country" Northeastern OK

Re: Can't stop air escaping from my mouth

Post by jeepdoctor » Sat Jan 01, 2022 3:48 pm

Try this. Use a Leader alcohol swab and clean the oils from the skin around your mouth. Tear off a 6 to 8” piece of transpore tape and place over mouth. Tear off a 4” piece of 3M durapore tape and place over the transpore tape. I have been doing this for about 20 years and it works better than anything else I have tried. I buy the tape 12 rolls at a time from Amazon. 😀

User avatar
Respirator99
Posts: 344
Joined: Mon Jul 05, 2021 12:39 am
Location: Australia

Re: Can't stop air escaping from my mouth

Post by Respirator99 » Sat Jan 01, 2022 9:31 pm

You could try nasal pillows + chin strap + mouth tape + collar + ....

or you could just change to a full face mask and your problems will go away overnight. I never could understand the lengths people go to in order to persist with a nasal mask when an FFM is such a simple solution.
* Download Oscar
* Oscar help
* An alternative to Oscar - try SleepHQ

I have no medical training or qualifications. Take my advice for what it's worth.

User avatar
zonker
Posts: 11048
Joined: Fri Jun 19, 2015 4:36 pm

Re: Can't stop air escaping from my mouth

Post by zonker » Sat Jan 01, 2022 10:47 pm

Respirator99 wrote:
Sat Jan 01, 2022 9:31 pm
You could try nasal pillows + chin strap + mouth tape + collar + ....

or you could just change to a full face mask and your problems will go away overnight. I never could understand the lengths people go to in order to persist with a nasal mask when an FFM is such a simple solution.
because it simply won't work for me.

different people have different needs. if not, there wouldn't BE nasal pillows or nasal masks.

note that i never recommend ANY kind of mask because they all come with a large ymmv sticker.
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

VictorC
Posts: 62
Joined: Thu Oct 16, 2014 3:54 pm
Location: Belleville, IL

Re: Can't stop air escaping from my mouth

Post by VictorC » Sun Jan 02, 2022 7:49 am

From looking at my data it looks like I had a pretty good night last night. Still see several leaks, but nothing major. As for moving to a full face mask, I have a F10, but I've never been able to adjust to it. The longest I've ever been able to use it is for 3 days, and then I switched back to pillows, because I was so exhausted I could no longer function otherwise. I know a full face mask would solve my issue once and for all, it just brings with it other problems. Oh and even with a full face mask I'd still need a chin strap. My jaw has dropped under it in the past.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Settings: APAP at 10.5-14cm H2O, AFLEX @ 2, Humidifier @ 2, Hose Temperature @ 3, No Ramp