Woke Up Gasping
Re: Woke Up Gasping
Thanks Pugsy! Good to know. I actually have a Resmed Aircurve 10S not an auto despite my tag line. I couldn't find a plain 10S on the equipment drop down when I registered.
Re: Woke Up Gasping
Ahhhhh.
Then you simply won't have the flow limitation graph to evaluate since you don't have auto mode.
Then you simply won't have the flow limitation graph to evaluate since you don't have auto mode.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Woke Up Gasping
You might add your machine model to your comments section of your profile since it isn't available from the equipment menu. Will avoid confusion later.
Just don't choose a machine from the drop down menu.
Just don't choose a machine from the drop down menu.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
- vandownbytheriver
- Posts: 405
- Joined: Sat Feb 03, 2024 11:42 pm
Re: Woke Up Gasping
As Pugsy noted, you don't get FL data with the S. I can see flat-topping on your breaths, you probably need to go up in pressure some more. You're still getting CA's... keep working that pressure up. Zoom into Breathing at 10:45:40 to :50, see the flat tops? Those are FL's. Bi-level is meant to punch those up... but at these lower pressures you'll get more CA's if you increase iPap diff to ePap. You're not really close to what I consider bi-level cm's... I paid big bux for a lab bi-level titration, entirely worth it.Airyhead wrote: ↑Wed May 15, 2024 9:05 amJust a FYI. Troubling shooting the odd data that SleepHQ reported, I downloaded the SD card to the SleepHQ site from my PC instead of from my iPhone and the anomalies went away. But also just now seeing that there is no flow limit data from the Resmed 10S. The Resmed 11 did report that data. Does anyone know if this is normal for the 10s data set?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: I use O2Ring, Oscar, SleepHQ, and Cover Roll Stretch mouth tape. |
Re: Woke Up Gasping
I do see several "Flat-tops" throughout the night as well as some PS ripples with exhalations. What's interesting to see is in the past when I would get these disruptions my breathing would become very unstable, as in periodic breathing, (reminds of the Tacoma Narrows Bridge oscillations videos from the 1940's) and eventually end in a 20 second CA,..,,and then repeat. I seem to recover better with these new settings and have fewer CA's following these events.vandownbytheriver wrote: ↑Wed May 15, 2024 9:28 pmAs Pugsy noted, you don't get FL data with the S. I can see flat-topping on your breaths, you probably need to go up in pressure some more. You're still getting CA's... keep working that pressure up. Zoom into Breathing at 10:45:40 to :50, see the flat tops? Those are FL's. Bi-level is meant to punch those up... but at these lower pressures you'll get more CA's if you increase iPap diff to ePap. You're not really close to what I consider bi-level cm's... I paid big bux for a lab bi-level titration, entirely worth it.
I have had Aerophagia at pressures much above where I am now so am slow to go higher. I'm still having leak and rain-out problems and would like to get those settled before moving up. Over-all I'm encouraged by the improvements and very much appreciate the interest and help. Today's SleepHQ https://sleephq.com/public/d364deeb-ea9 ... 6172771f38
Pugsy, thanks for the advice on the equipment. I've changed the settings so will see how this looks now.