Leaks and Mouth Breathing

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
milesleroy
Posts: 22
Joined: Fri Oct 10, 2014 7:18 am

Leaks and Mouth Breathing

Post by milesleroy » Wed Oct 29, 2014 9:55 am

I use a nasal mask. Lately I have been waking up with a very dry mouth. But I don't see any indicator of large leaks.
Can you have mouth breathing without leaks showing up?

User avatar
Jeff241
Posts: 49
Joined: Thu Jul 17, 2014 6:58 am
Location: MN

Re: Leaks and Mouth Breathing

Post by Jeff241 » Wed Oct 29, 2014 10:02 am

I assume you've checked your data and can't see any large leaks. If your data is showing no leak then I'm not sure what is going on.

You can have mouth leaks that aren't large enough to cause a problem. You may just have your mouth open enough for some air to leak and cause a dry mouth. Your machine can handle some leakage, it's when it gets into the large range that the machine can't compensate.

_________________
Mask: Pilairo Q Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Straight CPAP @ 12 cmH2O, EPR 2, Sleepyhead Software v0.9.8-1
Last edited by Jeff241 on Wed Oct 29, 2014 10:07 am, edited 2 times in total.

User avatar
Julie
Posts: 19908
Joined: Tue Feb 28, 2006 12:58 pm

Re: Leaks and Mouth Breathing

Post by Julie » Wed Oct 29, 2014 10:03 am

Yes you can, especially if they're particularly large and register as clear airway events... but I would also check your humidity levels as climate and heating can use more (or less) water. If you are losing air from mouth breathing, your therapy won't be as effective and could explain your higher AHIs.

milesleroy
Posts: 22
Joined: Fri Oct 10, 2014 7:18 am

Re: Leaks and Mouth Breathing

Post by milesleroy » Wed Oct 29, 2014 10:06 am

I posted in another thread about how my AHI numbers have been going up recently. I assumed it wasn't mouth breathing b/c the leaks are not large on my reports.

Could mouth breathing artificially increase my AHI?

User avatar
Julie
Posts: 19908
Joined: Tue Feb 28, 2006 12:58 pm

Re: Leaks and Mouth Breathing

Post by Julie » Wed Oct 29, 2014 11:29 am

Not artificially... if you mouth breathe, you lose therapy air and so your machine records lowered effectiveness and the AHI goes up. Lots of people purposely raise their pressure hoping for better results, but they just make things worse because more pressure pushes more air out instead of down their airway.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Leaks and Mouth Breathing

Post by Wulfman... » Wed Oct 29, 2014 1:03 pm

milesleroy wrote:I use a nasal mask. Lately I have been waking up with a very dry mouth. But I don't see any indicator of large leaks.
Can you have mouth breathing without leaks showing up?
"Dry mouth" is pretty much an indicator of mouth-breathing or mouth-leaking......and your therapy is being compromised.
You can still have dry mouth with a full face mask by mouth-breathing, but your therapy isn't compromised.

The "leak numbers" don't have to be significant to indicate that you're losing your therapy air.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Pugsy
Posts: 64014
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Leaks and Mouth Breathing

Post by Pugsy » Wed Oct 29, 2014 1:16 pm

Any leak..no matter what the cause... will show up on the reports.
The machine can compensate for a good bit of leak...no matter what the cause.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

Noctuary
Posts: 444
Joined: Sat Oct 04, 2014 1:10 pm

Re: Leaks and Mouth Breathing

Post by Noctuary » Wed Oct 29, 2014 4:56 pm

I don't mouth-breathe but several times a week I wake up with a dry mouth, and sometimes I hear a hissing that wakes me up.

I'm willing to try a full face mask, but the more mask you have the more leaks you fight.

_________________
Mask
My machine: ResMed S9 Auto/H5i humidifier; in APAP mode- pressure range 8-13

Mask: Flexifit 407

User avatar
Julie
Posts: 19908
Joined: Tue Feb 28, 2006 12:58 pm

Re: Leaks and Mouth Breathing

Post by Julie » Wed Oct 29, 2014 7:02 pm

That's just not true. If your FFM fits well there's no reason to get more leaks, though some would have you believe it.

And how do you know you don't mouth breathe at 4 a.m. or 1 am, etc?

User avatar
Sheriff Buford
Posts: 4082
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: Leaks and Mouth Breathing

Post by Sheriff Buford » Wed Oct 29, 2014 7:16 pm

I may be wrong... but mouth-breathing will not show up as a leak. The issue is air escaping thru the mouth instead if entering your body for needed therapy. Once the air enters your body (in your case thru the nose) the machine not detect a leak rate. If your ahi is in line then you will have time to work on resolving the dry mouth. The machine will monitor the ahi rate which is a result of improper therapy.

Sheriff

User avatar
Pugsy
Posts: 64014
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Leaks and Mouth Breathing

Post by Pugsy » Wed Oct 29, 2014 7:24 pm

Sheriff Buford wrote:I may be wrong... but mouth-breathing will not show up as a leak.
Sorry...you are wrong. Mouth breathing will show up as a leak if air is exiting the mouth in an amount to alter the pressure the machine thinks you should have.
There's no way to have massive air leaks to the point of compromising therapy without them showing up on the leak graphs.
Now not all mouth breathing causes massive exodus of air out the mouth...I proved that experiment myself and have documented proof.

If someone is mouth breathing to the point that there's a massive amount of air exiting the mouth and not going down the airway....it's going to show up on the leak reports.
Even ResMed documents what it looks like in their ResScan interpretation guide.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Leaks and Mouth Breathing

Post by palerider » Wed Oct 29, 2014 7:33 pm

Sheriff Buford wrote:I may be wrong... but mouth-breathing will not show up as a leak. The issue is air escaping thru the mouth instead if entering your body for needed therapy. Once the air enters your body (in your case thru the nose) the machine not detect a leak rate. If your ahi is in line then you will have time to work on resolving the dry mouth. The machine will monitor the ahi rate which is a result of improper therapy.

Sheriff
if you're using a nasal mask, then air leaking out your mouth is the same as air leaking out anywhere else.

it reduces the pressure available to keep your throat open, so the machine has to work harder to compensate, and if too much is leaking, then it can't keep the pressure up.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Sheriff Buford
Posts: 4082
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: Leaks and Mouth Breathing

Post by Sheriff Buford » Thu Oct 30, 2014 5:54 am

Pugsy wrote:sorry...you are wrong.
I kinda knew I was, but my logic told me when wearing pillows, and the air has left the machine, it doesn't know what happens.... gotta go.. I hear the dog barkin' in the back yard... gotta' go check...

Sheriff

User avatar
Starlette
Posts: 1725
Joined: Mon Feb 08, 2010 9:30 pm
Location: Denver, CO

Re: Leaks and Mouth Breathing

Post by Starlette » Thu Oct 30, 2014 9:48 am

So OA's are the same as leaks?
141030 SH

Starlette

_________________
MachineMask
Additional Comments: Silent Nite Dental Appliance

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Leaks and Mouth Breathing

Post by palerider » Thu Oct 30, 2014 10:10 am

Starlette wrote:So OA's are the same as leaks?
not in the least.

completely different, what made you think that?

leaks are a loss of air, oa's are a blockage in your airway.

enough leaks can cause more oa's because you don't have enough pressure to prevent them.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.