Translate Please

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
EMTunit31
Posts: 95
Joined: Wed Oct 01, 2014 2:49 pm
Location: Ohio

Translate Please

Post by EMTunit31 » Thu Oct 02, 2014 11:48 pm

My AHI started at 84.7
64.5 central
3.2 obstructive
1.8 mixed
15.2 hypopneas

65 lowest Spo2 % during sleep
Can someone explain all this to me, please

User avatar
kaiasgram
Posts: 3568
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: Translate Please

Post by kaiasgram » Thu Oct 02, 2014 11:55 pm

Is this from your sleep study with no PAP machine being used?

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Mask: Aloha Nasal Pillow CPAP Mask with Headgear
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura

User avatar
EMTunit31
Posts: 95
Joined: Wed Oct 01, 2014 2:49 pm
Location: Ohio

Re: Translate Please

Post by EMTunit31 » Fri Oct 03, 2014 12:08 am

Yes. The study was done about 3 weeks ago. I need a better understanding of what all this means. Layman style

User avatar
kaiasgram
Posts: 3568
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: Translate Please

Post by kaiasgram » Fri Oct 03, 2014 12:24 am

OK, then, the short version

AHI = per hour average of apnea events. At 84.7 events per hour average, you fall in the "Severe" category.

The numbers below the AHI line show a breakdown of the types of apnea events that make up that 84.7 AHI:

central (64.5 per hour) - this is where you stopped breathing but your airway was open, meaning it's not an "obstructive" event

obstructive (3.2 per hour) - you stopped breathing and your airway was closed/obstructed

mixed (1.8 per hour) - you stopped breathing and the event has both central and obstructive characteristics

hypopnea (15.2 per hour) - you had restricted breathing with a partial collapse (obstruction) of your airway

The SpO2 number - When you stop breathing for a period of time your blood oxygen level starts to drop. Your biggest drop brought your blood oxygen level down to 65%. That is serious.

I hope this helps as a starter. Ask more questions when you're ready, we're here to help each other.

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Mask: Aloha Nasal Pillow CPAP Mask with Headgear
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura
Last edited by kaiasgram on Fri Oct 03, 2014 12:35 am, edited 1 time in total.

User avatar
EMTunit31
Posts: 95
Joined: Wed Oct 01, 2014 2:49 pm
Location: Ohio

Re: Translate Please

Post by EMTunit31 » Fri Oct 03, 2014 12:34 am

Thank You so much. I not only understand better but I'm a little concerned now. I was given a Res Mart machine but I'm working on turning it back in for a ResMed. I need to correct theses problems and I need to find out whats going on at night. ResMart won't work in Sleepyhead. I need something that does. I'm sure I'll be asking more questions.
Oh yeah, them readings are from a sitting position. I don't dare lay down and sleep.

User avatar
Pugsy
Posts: 64014
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Translate Please

Post by Pugsy » Fri Oct 03, 2014 6:21 am

I am going to add some fuel to the fire....
With having a LOT more centrals than obstructives.....that particular machine cpap/apap may not even be the right type of machine for you.
What did your RX say...EXACTLY? Do you know?
With this diagnostic sleep study showing that many centrals did you get a sleep study with a machine at all? Those are called titration sleep studies.
Did your doctor discuss the centrals with you? If not, he/she should have. A few centrals are normal but 60 plus per hour we don't ignore and that means your diagnosis should be Central Sleep Apnea and not Obstructive Sleep Apnea.
CPAP/APAP is not the machine of usual choice for a diagnosis of Central Sleep Apnea.
Most of the time people with that many centrals jump to the front of the line and get one of those high dollar machines that treat centrals.
Either the ResMed S9 Adapt or the PR S1 Auto Advanced ...both are ASV types of machines
Adapto Servo Ventilation which means that they breathe for you when those centrals pop up.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
EMTunit31
Posts: 95
Joined: Wed Oct 01, 2014 2:49 pm
Location: Ohio

Re: Translate Please

Post by EMTunit31 » Fri Oct 03, 2014 1:05 pm

Pugsy,
I didn't talk to a sleep doctor at all. I took the Home sleep study and thats it, because my insurance won't pay for me to go in to the center for an overnight 2nd part. After the "in home" we went straight to the machine. When I went to pick up my machine I thought I was going to get set up for another sleep study (2nd) part at home. When we got there she had the machine and equipment setting out on the desk and ready to go over. At that point I was so confused that I told her I didn't even know why I was there and she casually said "because you stop breathing at night". Ummm, I was aware of that and I explained that I thought I was coming in for equipment for the second part of study. I have already had a stroke that has altered my vision so when I seen ResMed on the head gear and Res---- on the machine, I assumed it was ALL ResMed. A few days ago I realized it's not. I am now on a mission to get the machine exchanged. With your permission, I would like to print out your reply and take it with me. You guys know more about this than I do and with ammo she is less likely to try and feed me some line of crap. Don't mistake me wanting to take this post with me as coward because I'm not. I would like for them to realize that I have checked into this and want the best for myself. My aunt wanted to go with me when I got my equipment so she could make sure I got the right stuff but like I said, I didn't know I was going in to get the equipment. I thought I was getting ready for the 2nd part of the study. She wouldn't have allowed this to happen.

User avatar
kaiasgram
Posts: 3568
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: Translate Please

Post by kaiasgram » Fri Oct 03, 2014 1:42 pm

EMT, I'm glad Pugsy spotted you today -- I knew she would! What we post becomes public material so of course you can print out a copy of whatever you want here and take it with you. But be aware that the equipment provider can only dispense what's written on the prescription so you really need to see a copy of that prescription. So I think the doctor should be your first stop.

What was involved in your home sleep test -- besides a thing on your finger, did you have a couple of belts around your chest and abdomen? What about EEG leads attached to your face or scalp?

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Mask: Aloha Nasal Pillow CPAP Mask with Headgear
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura

User avatar
EMTunit31
Posts: 95
Joined: Wed Oct 01, 2014 2:49 pm
Location: Ohio

Re: Translate Please

Post by EMTunit31 » Fri Oct 03, 2014 2:14 pm

I'm glad Pugsy found me too. Its time for me to be aggressive and take matters into my own hands instead of relying on someone else to do it. Now that I'm clear on the meanings of all this I feel confident enough to do just that. Thank you.
The 1 and only sleep study I did was at home. I had a belt around my belly and one around my chest. I had the O2 on my finger, nasal cannulas, and a probe hanging in front of my mouth. No wires attached to my head. I don't have the script but I do have the doctors interpret report.
1. Obstructive sleep apnea/hypopnea syndrome
ICD-9 327.23
Severity Severe, overall AHI at 85 events per hour sleep hour
Minimum oxygen saturation: 70% (I don't understand this because it clearly says on the report 65%)
I will try and figure out how to post my report on here.

User avatar
kaiasgram
Posts: 3568
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: Translate Please

Post by kaiasgram » Fri Oct 03, 2014 2:42 pm

To post your report (good idea): White out your name first. If the report is already in digital format you can upload it to a free site like Photobucket or imgur. When you upload you will be shown an image code that starts with [IMG]. You copy and then paste that code right into your post here. It will just be text, but if you click the Preview button you should see your image -- which means we'll see it too when you Submit your post.

If your report is on paper only, you could take photos of the page(s), then get the photo onto your computer and upload it to the Photobucket or imgur, etc.

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Mask: Aloha Nasal Pillow CPAP Mask with Headgear
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura

User avatar
Bill44133
Posts: 1087
Joined: Tue Jul 09, 2013 8:34 pm
Location: North Royalton, OH

Re: Translate Please

Post by Bill44133 » Fri Oct 03, 2014 3:03 pm

Welcome to the forum You are in a good place here. Work any issues you have out here. Lots of smart and helpful people here.

Take 15 minutes and review this youtube video it will help you understand what is going on with you:
https://www.youtube.com/watch?v=-gie2dhqP2c

This video put the whole thing in perspective for me. I was able to show this video to my family to help them understand what is going on me.


I wish you well

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Zzz-Mask Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Settings are IPap 23 EPap 19

User avatar
EMTunit31
Posts: 95
Joined: Wed Oct 01, 2014 2:49 pm
Location: Ohio

Re: Translate Please

Post by EMTunit31 » Fri Oct 03, 2014 3:25 pm

Thank you Bill for that link. It was very informative. I will do the same as you did and show my family so they understand also whats going on.

User avatar
EMTunit31
Posts: 95
Joined: Wed Oct 01, 2014 2:49 pm
Location: Ohio

Re: Translate Please

Post by EMTunit31 » Fri Oct 03, 2014 3:44 pm

I hope this link works. It's to my Sleep Study Report

http://s249.photobucket.com/user/emtuni ... sort=3&o=0

User avatar
kaiasgram
Posts: 3568
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: Translate Please

Post by kaiasgram » Fri Oct 03, 2014 4:04 pm

EMTunit31 wrote:I hope this link works. It's to my Sleep Study Report

http://s249.photobucket.com/user/emtuni ... sort=3&o=0
The link works.

Honestly if I had a report like this I'd take it to my own doctor and insist on some serious attention. All those centrals need to be examined more closely with a full sleep study (or at least an in-lab titration). The home test you did was not measuring your actual sleep (brain waves) because there were no EEG leads. Hopefully Sludge will correct me if I'm wrong but I think the positive identification of central apnea requires positive identification of sleep as evidenced by the EEG recordings.

The inadequacy of home sleep tests is evident here. For example the report shows your sleep position as supine, though you said you slept in a sitting position all night. There's also some wonkiness in the report where it says # of desats was zero.

If indeed you do have central apnea it would be a travesty to make you suffer through weeks and weeks on the wrong machine especially with your level of severity.

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Mask: Aloha Nasal Pillow CPAP Mask with Headgear
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura

User avatar
Sludge
Posts: 953
Joined: Wed Dec 04, 2013 4:36 am

Re: Translate Please

Post by Sludge » Fri Oct 03, 2014 5:24 pm

EMTunit31 wrote: I have already had a stroke ...
That would be a very good reason to have all those centrals.

brb...
You Kids Have Fun!!