backup cpap

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
splicer76
Posts: 2
Joined: Sat Sep 27, 2014 2:16 pm

backup cpap

Post by splicer76 » Sat Sep 27, 2014 2:37 pm

I have an older remstar pro with c-flex I would like to bring on camping trips or for when I'm not sleeping at home. I've have a newer cpap at home. I just found my backup machine. I've changed the pressure on it in the past using online directions with no problem. Now I'm getting an e-8 error. The machine still works at the pressure it's been set at but the error code won't let me adjust to my prescribed pressure. I'm trying to go from a 10 to a 12. How can I fix this? Will my home health provider fix it even though I'm using a newer machine regularly? Any help on this would be greatly appreciated

User avatar
Nick Danger
Posts: 621
Joined: Mon Jun 09, 2014 3:13 pm

Re: backup cpap

Post by Nick Danger » Sat Sep 27, 2014 4:07 pm

Link to the error codes on another forum: http://www.apneaboard.com/wiki/index.ph ... rror_codes - I'm not sure if these are the correct codes for your machine. The clinician's manual should list the error codes and their causes.

E-8 doesn't look good. Wulfman, in another post, talked about it indicating a control button is stuck as another possibility for that code - you might check that.

Let us know if you don't have the clinician's manual - somebody probably has it. If not, you should be able to get it from apneaboard.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: APAP mode, minimum pressure = 9. No ramp, EPR = 3, medium. Soft cervical collar. Sleepyhead software.

Guest

Re: backup cpap

Post by Guest » Sun Sep 28, 2014 11:34 am

I don't have the clinicians manual unfortunately. .

Guest

Re: backup cpap

Post by Guest » Sun Sep 28, 2014 11:51 am

splicer76 wrote:I have an older remstar pro with c-flex I would like to bring on camping trips or for when I'm not sleeping at home. I've have a newer cpap at home. I just found my backup machine. I've changed the pressure on it in the past using online directions with no problem. Now I'm getting an e-8 error. The machine still works at the pressure it's been set at but the error code won't let me adjust to my prescribed pressure. I'm trying to go from a 10 to a 12. How can I fix this? Will my home health provider fix it even though I'm using a newer machine regularly? Any help on this would be greatly appreciated
Could you clear up exactly which model you are having trouble with.

User avatar
Julie
Posts: 19907
Joined: Tue Feb 28, 2006 12:58 pm

Re: backup cpap

Post by Julie » Sun Sep 28, 2014 12:30 pm

If you have the Legacy ("Tank") which I have as a back-up, this is the error code you referred to I think -

"E-8 ERR_IRQ_STACK_OVERFLOW •Defective RAM chip •Replace Main PCA"

which may be fatal (at least as far as you're concerned... but I guess you could try calling Respironics 1-800-345-6443.
Last edited by Julie on Mon Sep 29, 2014 5:00 am, edited 1 time in total.

Guest

Re: backup cpap

Post by Guest » Sun Sep 28, 2014 6:07 pm

    The machine with the error code is a respironics remstar pro with c-flex. It has two gray arrow buttons under the digital screen. To the left of those buttons are two black buttons. One says heat. And the the ramp button is right under it. To the right of the gray arrow buttons is the power button. The machine has a slot for a credit card size memory card. Model number 1005961.

    User avatar
    Wulfman...
    Posts: 6688
    Joined: Mon Sep 01, 2014 6:41 pm
    Location: Nearest fishing spot

    Re: backup cpap

    Post by Wulfman... » Sun Sep 28, 2014 6:16 pm

    splicer76 wrote:I have an older remstar pro with c-flex I would like to bring on camping trips or for when I'm not sleeping at home. I've have a newer cpap at home. I just found my backup machine. I've changed the pressure on it in the past using online directions with no problem. Now I'm getting an e-8 error. The machine still works at the pressure it's been set at but the error code won't let me adjust to my prescribed pressure. I'm trying to go from a 10 to a 12. How can I fix this? Will my home health provider fix it even though I'm using a newer machine regularly? Any help on this would be greatly appreciated
    That alone doesn't tell us which generation of model of machine you have. Respironics used the "Pro" model designation for the last dozen years or so.
    I checked my service manuals for various models of the "Pro" and they say that the E-8 error could be anything from a stuck key to the PCA board (motherboard) could be fried.

    If it has 5 buttons on the top, you should be able to get into the setup by holding the two gray buttons below the LCD screen while you plug the power cord into the back. See instructions below.


    If it's one of the old Legacy models (pre-M Series) with the Ramp and Humidifier buttons on the left side (as you face it from the top), you could try these instructions.

    6.4 System Error codes

    This section contains information on extracting error codes that have
    occurred during use of the REMstar Auto with C-Flex or REMstar Pro 2.
    There are two ways to extract the information.

    6.4.1 Extracting Error Codes from the Unit

    Step 1 Hold the humidifier button and ramp button down while
    plugging in the power cord. Continue holding the buttons
    down until the unit beeps.

    Step 2 The most recent error code will be displayed. By pressing
    the humidifier key, the unit will show the previous error code,
    allowing the user to view the entire error log in a circular buffer
    fashion. Only the ten most recent errors will be shown. An E-0
    indicates no error.

    Step 3 By pressing the ramp key, the unit will display the blower hours
    when the error occurred. Pressing the humidifier key will allow the
    user to view the hours that each error occurred in a circular buffer
    fashion.

    Step 4 To erase the error logs, press the right user button. An X will
    appear in the lower left corner of the LCD. Then press the
    Pressure On / Off key. This will erase all errors and the blower
    hours at which the errors occurred, and return the unit to the
    system self test screen.


    These should work to get into the machine setup. They're for the Pro 2, but should work for a Legacy Pro model.


    When in the Setup Menu, the humidifier ^ and ramp v buttons operate as up and down keys to change the settings, the left/right user buttons < > allow you to go to the previous/next question or setting, and the pressure start/stop button is used to exit the Setup Menu. Holding the humidifier or ramp buttons down will cause the values to change more quickly.

    To enter the Therapy Setup Menu, hold the two top user buttons < > down while plugging in the power cord.
    Continue holding the buttons down until the REMstar Pro 2 with C-Flex beeps.

    Note: The word "setup" will appear on all of the screens indicating that you are in the Therapy Setup Menu.
    If you press the Pressure start/stop button, you will exit the Setup Menu.

    a. Compliance hours: (recommend leaving alone) Select next setting with >

    b. Therapy mode: (CPAP or CFLE) select with ^ or v Select next setting with >

    c. CPAP pressure setting: Select with ^ or v Select next setting with >

    d. C-Flex mode (if you chose CFLE mode): Select setting 1, 2 or 3 with ^ Select next setting with >

    e: Ramp time setting: Select with ^ or v (ramp time will be turned off with a setting of 0:00) Select next setting with >

    f. Ramp pressure setting: Select with ^ or v Select next setting with >

    g. Patient disconnect setting: 1 = on 0 = off Select with ^ or v Select next setting with >

    h. Buttons lights setting: 1 = on 0 = off Select with ^ or v (Last setting. Use On/Off button to exit Setup)


    .
    (5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
    "Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
    User since 05/14/05

    drbuck
    Posts: 4
    Joined: Sun Sep 28, 2014 6:49 pm
    Contact:

    Re: backup cpap

    Post by drbuck » Sun Sep 28, 2014 6:57 pm

    I generally have just a good dental mouthpiece for my backup. It is not as strong as the CPAP but eliminates my snoring and lowers my AHI.

    User avatar
    Wulfman...
    Posts: 6688
    Joined: Mon Sep 01, 2014 6:41 pm
    Location: Nearest fishing spot

    Re: backup cpap

    Post by Wulfman... » Sun Sep 28, 2014 7:16 pm

    drbuck wrote:I generally have just a good dental mouthpiece for my backup. It is not as strong as the CPAP but eliminates my snoring and lowers my AHI.
    I'm guessing you didn't even read the original post......did ya?
    Maybe just the "Subject" title and you went from there.


    Den

    .
    (5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
    "Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
    User since 05/14/05

    Guest

    Re: backup cpap

    Post by Guest » Sun Sep 28, 2014 7:25 pm

    splicer76 wrote:The machine with the error code is a respironics remstar pro with c-flex. It has two gray arrow buttons under the digital screen. To the left of those buttons are two black buttons. One says heat. And the the ramp button is right under it. To the right of the gray arrow buttons is the power button. The machine has a slot for a credit card size memory card. Model number 1005961.
    Check the memphis craigslist

    User avatar
    archangle
    Posts: 9294
    Joined: Sun Mar 27, 2011 11:55 am

    Re: backup cpap

    Post by archangle » Sun Sep 28, 2014 11:23 pm

    Guest wrote:
      The machine with the error code is a respironics remstar pro with c-flex. It has two gray arrow buttons under the digital screen. To the left of those buttons are two black buttons. One says heat. And the the ramp button is right under it. To the right of the gray arrow buttons is the power button. The machine has a slot for a credit card size memory card. Model number 1005961.
      1005961 is a "Legacy" REMstar PRO with C-Flex.

      Be sure the filters are clean. Run it for a few seconds with the filters out, and with the humidifier tank out and see if the noise goes away.

      Is the noise a grinding bearing type noise, or a whistling or air leak type of noise?

      I've done well with craigslist buys, but I've often had to look for months to find a good machine at a good price. You'll see a lot of crazy high price offers of old junk at higher than new online prices.

      This is probably a good reminder to everyone to watch for a backup machine before you need one.

      _________________
      Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
      Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
      Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
      Please enter your equipment in your profile so we can help you.
      Click here for information on the most common alternative to CPAP.
      If it's midnight and a DME tells you it's dark outside, go and check for yourself.

      Useful Links.

      User avatar
      Drowsy Dancer
      Posts: 1271
      Joined: Mon Feb 28, 2011 9:35 am
      Location: here

      Re: backup cpap

      Post by Drowsy Dancer » Mon Sep 29, 2014 12:35 am

      archangle wrote:This is probably a good reminder to everyone to watch for a backup machine before you need one.
      +1

      Buying a machine via Craigslist was one of the ickiest things I've ever done (hey, I've led a sheltered life), but also one of the smartest (the other being buying real estate in Seattle in 1988, but I digress).

      _________________
      Machine: PR System One REMStar 60 Series Auto CPAP Machine
      Mask: Swift™ FX Bella Nasal Pillow CPAP Mask with Headgears
      Additional Comments: Software: SleepyHead. Pressure: APAP 9.5 min/11 max, A-Flex x2
      How we squander our hours of pain. -- Rilke

      Guest

      Re: backup cpap

      Post by Guest » Tue Sep 30, 2014 2:08 pm

      I never said it was making a noise. It still works fine. I just can't change the pressure setting. But I will try running it without the filters in it for a few minutes. Maybe by some chance it will clear out the error code. This machine doesn't have a humidifier. I had to use a separate humidifier connected via hose when I used to use it

      Guest

      Re: backup cpap

      Post by Guest » Tue Sep 30, 2014 2:17 pm

      Guest wrote:
      splicer76 wrote:The machine with the error code is a respironics remstar pro with c-flex. It has two gray arrow buttons under the digital screen. To the left of those buttons are two black buttons. One says heat. And the the ramp button is right under it. To the right of the gray arrow buttons is the power button. The machine has a slot for a credit card size memory card. Model number 1005961.
      Check the memphis craigslist
      I can't post a link sooo ...
      Check the memphis craigslist

      Guest

      Re: backup cpap

      Post by Guest » Tue Sep 30, 2014 2:25 pm

      Thanks wulfman. Your post helped me figure it out. Greatly appreciated!