Large leak reported even though leak rate is small. (Huh??)

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
JSG
Posts: 15
Joined: Mon Sep 22, 2014 10:52 am

Large leak reported even though leak rate is small. (Huh??)

Post by JSG » Mon Sep 22, 2014 10:57 am

Hi there.

I'm a new BiPAP user and have been scouring the internet to find an answer to this, but to no avail. I'm using SleepyHead software to look at my data each night. It turns out that I am having large leaks for about 2 hours at a time, yet, for that same time period the leak rate shown (on the leak rate graph) is minimal. How is this possible? Isn't it a contradiction? Bear in mind that the machine does show some leak rate--that is, it's not showing a perfectly flat line--so the machine definitely is picking up leak data instead of turning off leak data collection (and reporting a value of 0 leakage during that time).

Thanks for your help!

Josh
****************************************
Philips Respironics BiPAP Auto Bi-Flex System One
Fisher & Paykel Simplus Full Face CPAP Mask
BiPAP Pressures: 20/16
Humidifier Level: 5

User avatar
Pugsy
Posts: 64026
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Large leak reported even though leak rate is small. (Huh??)

Post by Pugsy » Mon Sep 22, 2014 11:06 am

Please post an image of the detailed daily graphs showing what you are talking about.
Don't know how? We talk about it here.
Screen shot thread viewtopic.php?f=1&t=81072&p=737779#p737779

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
Sheriff Buford
Posts: 4082
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: Large leak reported even though leak rate is small. (Huh??)

Post by Sheriff Buford » Mon Sep 22, 2014 11:08 am

Please fill in your equipment data, so we can know what type of equipment you have. Since all machines are different and without knowing your brand... Resmed can compensate for mask vents and leak rates under 24 L/min. The report out is "unintentional" leak rates.

Sheriff

JSG
Posts: 15
Joined: Mon Sep 22, 2014 10:52 am

Re: Large leak reported even though leak rate is small. (Huh??)

Post by JSG » Mon Sep 22, 2014 1:43 pm

Thanks, guys. Being a technical writer, I definitely know how to take a screen shot. Now, if I only knew why they include that cup holder that mysteriously slides in and out of the computer... I'll take the screenshot and post it when I get home. Thanks again!
****************************************
Philips Respironics BiPAP Auto Bi-Flex System One
Fisher & Paykel Simplus Full Face CPAP Mask
BiPAP Pressures: 20/16
Humidifier Level: 5

JSG
Posts: 15
Joined: Mon Sep 22, 2014 10:52 am

Re: Large leak reported even though leak rate is small. (Huh??)

Post by JSG » Mon Sep 22, 2014 3:19 pm

OK, guys. I've posted my machine information in the signature. Here is a screenshot from SleepyHead of what I'm talking about. Notice how I have very large duration large leaks, and yet the leak rate shown is non-zero but minimal, which seems to be contradictory: http://joshgoldstein.org/cpap_pictures/LargeLeak.tiff
****************************************
Philips Respironics BiPAP Auto Bi-Flex System One
Fisher & Paykel Simplus Full Face CPAP Mask
BiPAP Pressures: 20/16
Humidifier Level: 5

JSG
Posts: 15
Joined: Mon Sep 22, 2014 10:52 am

Re: Large leak reported even though leak rate is small. (Huh??)

Post by JSG » Mon Sep 22, 2014 3:23 pm

I wasn't able to figure out how to insert the image inline in my previous post, but I was able to post it as a URL. Hope you guys are able to see it. Thanks in advance.
****************************************
Philips Respironics BiPAP Auto Bi-Flex System One
Fisher & Paykel Simplus Full Face CPAP Mask
BiPAP Pressures: 20/16
Humidifier Level: 5

User avatar
Krelvin
Posts: 1977
Joined: Tue Jun 06, 2006 5:23 pm
Location: Metro Phx Area - Dry Heat!

Re: Large leak reported even though leak rate is small. (Huh??)

Post by Krelvin » Mon Sep 22, 2014 3:31 pm

Image
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Large leak reported even though leak rate is small. (Huh??)

Post by palerider » Mon Sep 22, 2014 3:52 pm

JSG wrote:I wasn't able to figure out how to insert the image inline in my previous post, but I was able to post it as a URL. Hope you guys are able to see it. Thanks in advance.
I don't think web browsers deal well with tiffs... they prefer gifs, jpgs and pngs

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

JSG
Posts: 15
Joined: Mon Sep 22, 2014 10:52 am

Re: Large leak reported even though leak rate is small. (Huh??)

Post by JSG » Mon Sep 22, 2014 3:54 pm

Thanks, Krelvin and Palerider. I'm still hoping someone will eventually have an answer to my question. Any ideas?
****************************************
Philips Respironics BiPAP Auto Bi-Flex System One
Fisher & Paykel Simplus Full Face CPAP Mask
BiPAP Pressures: 20/16
Humidifier Level: 5

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Large leak reported even though leak rate is small. (Huh??)

Post by palerider » Mon Sep 22, 2014 3:54 pm

JSG wrote:OK, guys. I've posted my machine information in the signature. Here is a screenshot from SleepyHead of what I'm talking about. Notice how I have very large duration large leaks, and yet the leak rate shown is non-zero but minimal, which seems to be contradictory
actually, your leaks are literally off the chart. the bottom line is a guess from sleepyhead about excess leakage, the top line, that's going off the top of the chart, is the total leak line, it should be closer to 50lpm, not well over 120.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zoocrewphoto
Posts: 3732
Joined: Mon Apr 30, 2012 10:34 pm
Location: Seatac, WA

Re: Large leak reported even though leak rate is small. (Huh??)

Post by zoocrewphoto » Mon Sep 22, 2014 3:56 pm

I'm seeing two leak lines. One in the normal range and one at the too that sometimes goes off the chart. I don't understand why there are two lines like that, but I suspect that is what is being reported. Pugsy will probably know why.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 autoset pressure range 11-17
Who would have thought it would be this challenging to sleep and breathe at the same time?

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Large leak reported even though leak rate is small. (Huh??)

Post by palerider » Mon Sep 22, 2014 4:39 pm

zoocrewphoto wrote:I don't understand why there are two lines like that, but I suspect that is what is being reported.
read the post above yours

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 64026
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Large leak reported even though leak rate is small. (Huh??)

Post by Pugsy » Mon Sep 22, 2014 6:15 pm

The leaks are massive. Whatever is going on with your leaks you need to work on them...a lot. Way more prolonged and massive leaks than I would accept and I let a lot of stuff slide.

Palerider is correct. You should see 2 leak lines and much closer together but your top leak line is so high it is off the charts a good bit of the time. Top line is total leak....excess leak plus the mask's expected vent rate.
The bottom line is supposed to be excess leak only but it may not be totally accurate. Best to base evaluations on the total leak line.

Take a look at mine below....the worst I got to was a little shy of 80 L/min and then for not very long. You should be able to easily see the top leak line.

Image

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

Tango
Posts: 75
Joined: Thu Aug 28, 2014 8:54 pm

Re: Large leak reported even though leak rate is small. (Huh??)

Post by Tango » Mon Sep 22, 2014 6:27 pm

I'm new and love reading these graphs when you all post them. Learning something new everyday!
Anyway, I don't see any events flagged?
Do you not have any?

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: I couldn't find the humidifier I use. I think it's climate control by resmed

User avatar
zoocrewphoto
Posts: 3732
Joined: Mon Apr 30, 2012 10:34 pm
Location: Seatac, WA

Re: Large leak reported even though leak rate is small. (Huh??)

Post by zoocrewphoto » Mon Sep 22, 2014 8:31 pm

palerider wrote:
zoocrewphoto wrote:I don't understand why there are two lines like that, but I suspect that is what is being reported.
read the post above yours
If you will look at the posting times, you will see that mine posted right after yours. It wasn't there when I started typing. And when I post from my iPod, I don't see posts that are new since I started to type a reply until I go to read the thread again.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 autoset pressure range 11-17
Who would have thought it would be this challenging to sleep and breathe at the same time?