Mouth breathing as clear events
Mouth breathing as clear events
Last two nights I've had AHIs of 2.45 of which 1.5 was clear airway apneas. With a nasal pillow and some allergy congestion could I be mouth breathing at times and triggering false clear airway events? Obstructive and hypopneas make up the other 1, no unclassifieds.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Sleepyhead SW. NeilMed and Alkalol Nasal rinses. Veramyst. AutoPAP 11-20 cms. Started June '14, untreated AHI 31-38, with PAP around 1. |
Re: Mouth breathing as clear events
What does you leak rate look like around the time of a CA event?
_________________
Mask: Oracle HC452 Oral CPAP Mask |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV |
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
Re: Mouth breathing as clear events
Sub 20 but flow rate is peaking. What does that imply?
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Sleepyhead SW. NeilMed and Alkalol Nasal rinses. Veramyst. AutoPAP 11-20 cms. Started June '14, untreated AHI 31-38, with PAP around 1. |
Re: Mouth breathing as clear events
Why do you assume that the Clear Airway events are false? IMO..you are nit picking. If your AHI is low and you are sleeping and feeling well, don't worry about the details.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is S9 Autoset...... |
Re: Mouth breathing as clear events
Remember that a few centrals are considered normal to have show up anyway. Like sleep onset centrals or sleep stage transition centrals.
Could mouth breathing alone get flagged by mistake? Not unless you pause your breathing itself. Hold your breath type of thing.
You can look at the flow rate and see if the flow is reduced.
Could you be mouth breathing and be holding your breath inadvertently while awake or semi awake ? Sure and then the machine would likely give that pause a central flag.
Mouth breathing doesn't interrupt the flow of air and the machine looks for interruptions/reduction of air flow. If the mouth breathing isn't huge then it may not even show up on the leak graphs. I have tested this myself while awake and have proof.
Could mouth breathing alone get flagged by mistake? Not unless you pause your breathing itself. Hold your breath type of thing.
You can look at the flow rate and see if the flow is reduced.
Could you be mouth breathing and be holding your breath inadvertently while awake or semi awake ? Sure and then the machine would likely give that pause a central flag.
Mouth breathing doesn't interrupt the flow of air and the machine looks for interruptions/reduction of air flow. If the mouth breathing isn't huge then it may not even show up on the leak graphs. I have tested this myself while awake and have proof.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
- BlackSpinner
- Posts: 9745
- Joined: Sat Apr 25, 2009 5:44 pm
- Location: Edmonton Alberta
- Contact:
Re: Mouth breathing as clear events
Are you taking allergy meds? If you are that could be causing the problem. I sleep lighter and have more disruptions when I use them.
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: Hybrid Full Face CPAP Mask with Nasal Pillows and Headgear |
Additional Comments: Quatro mask for colds & flus S8 elite for back up |
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal
Re: Mouth breathing as clear events
No allergy meds but not sleeping as well this week. More congestion with seasonal allergies I think.
So, Clear Airway and centrals are the same thing?
If I get a congested nose and then stop breathing until my mouth opens reflexively will that look like this?
Not hugely worried LSAT but annoying that I've gone from 1 to 2+ on AHI...
So, Clear Airway and centrals are the same thing?
If I get a congested nose and then stop breathing until my mouth opens reflexively will that look like this?
Not hugely worried LSAT but annoying that I've gone from 1 to 2+ on AHI...
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Sleepyhead SW. NeilMed and Alkalol Nasal rinses. Veramyst. AutoPAP 11-20 cms. Started June '14, untreated AHI 31-38, with PAP around 1. |
Re: Mouth breathing as clear events
Yes.englandsf wrote:So, Clear Airway and centrals are the same thing?
When SleepyHead was first written it was written for a Respironics machine which uses the term "Clear Airway" and the ResMed machines were added later and ResMed uses the term Central. They are both open airway cessation of breathing.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Mouth breathing as clear events
Thanks
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Sleepyhead SW. NeilMed and Alkalol Nasal rinses. Veramyst. AutoPAP 11-20 cms. Started June '14, untreated AHI 31-38, with PAP around 1. |
Re: Mouth breathing as clear events
just a point of order, if your nose was congested to the point where you'd have to breath through your mouth, and the machine scored an apnea, it would score an *obstructive* apnea, because the airway to your lungs would be blocked.englandsf wrote: With a nasal pillow and some allergy congestion could I be mouth breathing at times and triggering false clear airway events?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Mouth breathing as clear events
Good point pr, but I do get some weird air flow when my mouth is open and my nose is partially blocked.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Sleepyhead SW. NeilMed and Alkalol Nasal rinses. Veramyst. AutoPAP 11-20 cms. Started June '14, untreated AHI 31-38, with PAP around 1. |
Re: Mouth breathing as clear events
Maybe I'm adjusting the pillows and creating another confusing or odd flow
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Sleepyhead SW. NeilMed and Alkalol Nasal rinses. Veramyst. AutoPAP 11-20 cms. Started June '14, untreated AHI 31-38, with PAP around 1. |
Re: Mouth breathing as clear events
you'd have high leaks if that were it... did you?englandsf wrote:Maybe I'm adjusting the pillows and creating another confusing or odd flow
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Mouth breathing as clear events
Nope, just high flow.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Sleepyhead SW. NeilMed and Alkalol Nasal rinses. Veramyst. AutoPAP 11-20 cms. Started June '14, untreated AHI 31-38, with PAP around 1. |
Re: Mouth breathing as clear events
and do you know what that means?englandsf wrote:Nope, just high flow.
Get OSCAR