Thankful for this forum - Issues with the SD card
Thankful for this forum - Issues with the SD card
I am so thankful for this forum. I started having problems with my cpap machine last Aug 18, 2014. Every time I turned on the machine it beeped endlessly until i pressed the button. Also there was an error on the screen.
I installed Encore Pro on my new Windows 8.1 PC because I could not log into Encore Por on my Windows 7 laptop. I forgot the password.
So, after so long, I logged back in. Got the link for Sleepyhead. Downloaded and Installed. Then tried to Import data from the SD.
Yes! I got all my data from Aug 31, 2011 to Aug 17, 2014. That means there were days I did not get to use the machine. My bad.
Then I copied the SD files to a folder on my PC and deleted the files from the SD Card
So... now is the time to see if the machine will be able to write to the SD Card....
Will let you guys know what happens...
Just in case please leave some tips here for me to see tomorrow. Gotta sleep...
I installed Encore Pro on my new Windows 8.1 PC because I could not log into Encore Por on my Windows 7 laptop. I forgot the password.
So, after so long, I logged back in. Got the link for Sleepyhead. Downloaded and Installed. Then tried to Import data from the SD.
Yes! I got all my data from Aug 31, 2011 to Aug 17, 2014. That means there were days I did not get to use the machine. My bad.
Then I copied the SD files to a folder on my PC and deleted the files from the SD Card
So... now is the time to see if the machine will be able to write to the SD Card....
Will let you guys know what happens...
Just in case please leave some tips here for me to see tomorrow. Gotta sleep...
Last edited by Rogue on Thu Aug 21, 2014 6:47 pm, edited 2 times in total.
- deerslayer
- Posts: 1193
- Joined: Tue Dec 11, 2007 10:46 am
- Location: Mid Tn
Re: Just so thankful for this forum -
Ditto ! & The Caring People on Here From Around the Globe. Awesome Preventive Medicine Tool.....Should be subsidized by the Federal Govt. Because of Health Care SavingsRogue wrote:I am so thankful for this forum.
RESMED AUTOSET 10 / P 10 NASAL PILLOWS /straight cpap 12.20 / AHI 0.79 avg /
thanks to Nick sleep HQ
Feeling Blessed & firmly believe in The Holy Trinity
thanks to Nick sleep HQ
Feeling Blessed & firmly believe in The Holy Trinity
-
- Posts: 1005
- Joined: Sun Sep 11, 2011 7:55 am
- Location: North of Philadelphia, PA
Re: Just so thankful for this forum -
Not sure why you would delete the files from the SD card? However, most of the data should be available on the machine (unless you changed machines). If you need to copy the files from the PC back to the SD card, make sure you are not duplicating the files because of different folder names. The new set of data may have a different folder name.Rogue wrote: Then I copied the SD files to a folder on my PC and deleted the files from the SD Card
So... now is the time to see if the machine will be able to write to the SD Card....
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Use SleepyHead software. |
Not a medical professional - just a patient who has done a lot of reading
Thankful for this forum - Having issues with my SDcard
Hmmm...
Encore Pro 2
When I go to Data Card Utilities and select Preview card - message is 'The media belongs to the current patient.'
When I select Download card and click 'Download'' - message is 'Media read successfully - The card data is not valid and was not saved. Contact Product Support for assistance.'
in SleepyHead the card data is imported and I can see the report.
What is going on???
Seems similar to another post ''Frustration with Encore Viewer 2.0.23.0 on 64 bit computer''
My computer has Windows 8.1 64-bit
Encore Pro 2
When I go to Data Card Utilities and select Preview card - message is 'The media belongs to the current patient.'
When I select Download card and click 'Download'' - message is 'Media read successfully - The card data is not valid and was not saved. Contact Product Support for assistance.'
in SleepyHead the card data is imported and I can see the report.
What is going on???
Seems similar to another post ''Frustration with Encore Viewer 2.0.23.0 on 64 bit computer''
My computer has Windows 8.1 64-bit
Re: Thankful for this forum - Issues with the SD card
Exactly which version of Pro are you using....all the numbers.
Exactly which model machine are you using?
Exactly which model machine are you using?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Thankful for this forum - Issues with the SD card
EncorePro2Pugsy wrote:Exactly which version of Pro are you using....all the numbers.
Exactly which model machine are you using?
Program Version : 2.13.6.0
Database Version: 2.27.0.7
Database Server: (local)\ENCOREPRO2
Copyright 2014 Koninklijke Philips N.V. and its affiliates.
CPAP Machine Model : DS550HS
Last edited by Rogue on Thu Aug 21, 2014 7:14 pm, edited 1 time in total.
Re: Thankful for this forum - Issues with the SD card
Rogue wrote:Program Version : 2.13.6.0
This normally shows up when the software is not compatible with the machine itself. Pro 2.13 is a long ago version. Not sure why it would have worked in the past with your machine and not now unless it broke or maybe you got a newer machine.Rogue wrote:The card data is not valid and was not saved. Contact Product Support for assistance.'
Anyway, check your private messages as I sent you some information on getting the latest Pro version which works on all the machines and even Windows 8.1 (and older).
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Thankful for this forum - Issues with the SD card
Could it be because the Software Version on my machine?
SoftwareVersion=V1.24
SoftwareVersion=V1.24
Re: Thankful for this forum - Issues with the SD card
No I don't think so.Rogue wrote:Could it be because the Software Version on my machine?
SoftwareVersion=V1.24
Crap,,,,I just realized that you already have the latest version...I just sent you the same thing again.
I don't know why you are getting that message. I have 5 different PR S1 machines in my Encore Pro.
You might try creating a new user profile and see what happens.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Thankful for this forum - Issues with the SD card
I sent you Encore Basic. See what happens with it.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Thankful for this forum - Issues with the SD card
Installed Encore Basic 2.4 and tried to download the data.
Errors.
Media read successfully - The card data is not valid and not saved. Contact Product Support for assistance.
also
The media does not belong to an EncoreBasic patient.
Errors.
Media read successfully - The card data is not valid and not saved. Contact Product Support for assistance.
also
The media does not belong to an EncoreBasic patient.
Re: Thankful for this forum - Issues with the SD card
Do you have a new/blank SD card that you can use and try again?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Thankful for this forum - Issues with the SD card
Surprisingly though it can be read by Sleepyhead.
Just some other observations. The SD card only has 216KB left free space.
I deleted the files yesterday, but they are all back plus yesterday's data just by deleting in windows and also erasing using EncorePro.
Does the machine keep data from way back 2011 and writing it all back to the SD card?
I will have to buy a new SD Card.
Just some other observations. The SD card only has 216KB left free space.
I deleted the files yesterday, but they are all back plus yesterday's data just by deleting in windows and also erasing using EncorePro.
Does the machine keep data from way back 2011 and writing it all back to the SD card?
I will have to buy a new SD Card.
Re: Thankful for this forum - Issues with the SD card
I thought the machine only stored for 365 days the data that it stores long term but I very well could be wrong.Rogue wrote:Does the machine keep data from way back 2011 and writing it all back to the SD card?
Long term data would be hours of use and AHI and maybe some pressure or leak averages. The machine never stores the flow rate (wave form in Encore) which are the xxx.005 files and that's why the SD card has to be in the machine to be able to see flow rate.
You can go into the folder where all the files are (it's a sub folder under the machine serial number folder) and look for the oldest files. 3 main file types for data. xxx.001, xxx.002 and xxx.005. I thought that some of the older xxx.001 and xxx.002 files were written over after 30 days but maybe not if your SD card is nearly full.
I have no explanation as to why Encore wants to spit out your SD card. We have eliminated the known possible culprits except the SD card itself. That would be the next step if it were me. Get a new SD card (generic works fine) and put it in the machine and let it put whatever data it has stored in its internal memory to the SD card and then use it one night and the next day see what it happens with Encore.
When you get the new card and look at the files it puts on the SD card you should be able to see how far back the files go.
The modified date is for the night before. Like last night's data files will have a modified date of 8/22/14 on them.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Thankful for this forum - Issues with the SD card
There may be some residual on the card and just deleting the files didn't clear it off.
Since the card is not readable, try doing a complete format of the card and start over.
Then put the card in the cpap machine and let it re-write the data.
Encore should be able to read the card and report no data.
BTW what was the error that showed on the LCD screen? Perhaps that will be informative.
Since the card is not readable, try doing a complete format of the card and start over.
Then put the card in the cpap machine and let it re-write the data.
Encore should be able to read the card and report no data.
BTW what was the error that showed on the LCD screen? Perhaps that will be informative.
_________________
Mask: Oracle HC452 Oral CPAP Mask |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV |
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.