Pressure paradox

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Snorkler
Posts: 62
Joined: Thu Nov 26, 2009 11:58 am

Pressure paradox

Post by Snorkler » Sat Feb 27, 2010 9:01 am

I am using an auto-pap I have been tracking my AHI data. I have set the auto-pap in a range from 7 to 12. The pressure very rarely goes above 9 and when it does the results are worse. Reviewing the last week few week’s data I noticed that as my pressure increases from a minimum of 7 to 9 that the number of apneas increase (from 0.5 to 1.2), while the number of hypopneas decrease (from 3.3 to 0.4) and the overall AHI decreases (from 3.8 to 1.8 ). I have listed the effects of going from a psi of 7 to 9 below:

PSI: 7
AI: 0.5
HI: 3.3
AHI: 3.8

PSI: 8
AI: 1.0
HI: 3.1
AHI: 4.1

PSI: 9
AI: 1.2
HI: 0.4
AHI: 1.6

As pressure goes past 9 the numbers tend to get a little bit worse, so for me the best results are in a range from about 7 to 9.

My question is whether it is more desirable to pursue a lower overall apnea index which produces a somewhat higher hypopnea index (and a higher AHI index), or is it better to achieve a lower overall AHI index (even though this means a higher apnea index)?

By the way, I feel pretty good after a night’s sleep whether the pressure is at 7, 8 or 9. But I want to know if the more experienced people here have any guidance on this matter and whether this type of result is common.

Thanks,

Jeff
Last edited by Snorkler on Sat Feb 27, 2010 10:20 am, edited 1 time in total.

User avatar
BlackSpinner
Posts: 9745
Joined: Sat Apr 25, 2009 5:44 pm
Location: Edmonton Alberta
Contact:

Re: Pressure paradox

Post by BlackSpinner » Sat Feb 27, 2010 9:57 am

The question is whether the raise in pressure is due to stubborn apneas or vice versa.
Doesn't an apap keep rising until you start breathing again?

So basically you have some apneas that do not respond to the lower pressure and require the pressure to be jacked up until you do while the HI's seem to respond to lower pressures.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: Hybrid Full Face CPAP Mask with Nasal Pillows and Headgear
Additional Comments: Quatro mask for colds & flus S8 elite for back up
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal

jnk
Posts: 5787
Joined: Mon Jun 30, 2008 3:03 pm

Re: Pressure paradox

Post by jnk » Sat Feb 27, 2010 10:05 am

Are those the results of running at a straight pressure all night or in a range of pressures?

In general, with a Respironics machine, many consider the goal to be shooting for lower machine-estimated AHI overall. With a ResMed machine, many shoot for lower machine-estimated AI before paying any attention to HI or AHI. At least, that's how I understand it.

Snorkler
Posts: 62
Joined: Thu Nov 26, 2009 11:58 am

Re: Pressure paradox

Post by Snorkler » Sat Feb 27, 2010 10:23 am

jnk wrote:Are those the results of running at a straight pressure all night or in a range of pressures?

In general, with a Respironics machine, many consider the goal to be shooting for lower machine-estimated AHI overall. With a ResMed machine, many shoot for lower machine-estimated AI before paying any attention to HI or AHI. At least, that's how I understand it.
To answer your question, these are the result of letting the apap run in a range of pressures. Thanks for the answer that with a Respironics, many consider the goal to be shooting for the lower AHI overall. Given that whether the pressure is 7, 8 or 9 my overall AHI is below 5 I am not sure that there is much of a difference.

john5757
Posts: 341
Joined: Fri Jun 10, 2005 7:48 pm

Re: Pressure paradox

Post by john5757 » Sat Feb 27, 2010 10:30 am

Are you talking about events per hours? The slight increase of AI with pressure could be central in nature. Looks like pressure of 9 is your sweet spot. Just my opinion not medical.

_________________
MachineMask

Snorkler
Posts: 62
Joined: Thu Nov 26, 2009 11:58 am

Re: Pressure paradox

Post by Snorkler » Sat Feb 27, 2010 10:45 am

john5757 wrote:Are you talking about events per hours? The slight increase of AI with pressure could be central in nature. Looks like pressure of 9 is your sweet spot. Just my opinion not medical.
The figures represent average apneas, hypopneas and AHI per hour.

I am not sure about centrals, but according to the reports I have zero "non-responsive" events. I think this means that I have no central apneas reported.

jnk
Posts: 5787
Joined: Mon Jun 30, 2008 3:03 pm

Re: Pressure paradox

Post by jnk » Sat Feb 27, 2010 10:58 am

Snorkler wrote: . . . Given that whether the pressure is 7, 8 or 9 my overall AHI is below 5 I am not sure that there is much of a difference.
There is a difference if you FEEL the difference. Some do, some don't. Depends on how sensitive your nervous system is. Many feel that they may as well get AHI as low as possible, just for the principle of the thing and to keep them focused on their therapy. Others would rather stop thinking about it, to settle their own stomach.

Some might suggest that you now try those three pressures without varying pressures in order to see what that does for your numbers. Some get better numbers nightly for treatment using non-varying pressure. That may be worth checking out if you enjoy the data and tweaking thing. Although, if you don't feel a difference either way, no one would fault you for simply being happy with success.

jeff

Snorkler
Posts: 62
Joined: Thu Nov 26, 2009 11:58 am

Re: Pressure paradox

Post by Snorkler » Sat Feb 27, 2010 11:34 am

jnk wrote:
Snorkler wrote: . . . Given that whether the pressure is 7, 8 or 9 my overall AHI is below 5 I am not sure that there is much of a difference.
There is a difference if you FEEL the difference. Some do, some don't. Depends on how sensitive your nervous system is. Many feel that they may as well get AHI as low as possible, just for the principle of the thing and to keep them focused on their therapy. Others would rather stop thinking about it, to settle their own stomach.

Some might suggest that you now try those three pressures without varying pressures in order to see what that does for your numbers. Some get better numbers nightly for treatment using non-varying pressure. That may be worth checking out if you enjoy the data and tweaking thing. Although, if you don't feel a difference either way, no one would fault you for simply being happy with success.

jeff
I do not feel much of a difference when running pressure at 7, 8 or 9. Maybe this is because I am not too sensitive to the differences.

I had a cpap before I got this apap, but it did not have full data capabilities. Maybe I will follow your suggestion at some point to try running the apap in cpap mode (non-varying pressure). But for now, I am just trying to adjust the minimum pressure on the apap in order to imrpove the AHI data.

User avatar
KatieW
Posts: 1672
Joined: Sun Sep 27, 2009 11:07 am
Location: southern AZ

Re: Pressure paradox

Post by KatieW » Sat Feb 27, 2010 5:11 pm

Snorkler wrote:
My question is whether it is more desirable to pursue a lower overall apnea index which produces a somewhat higher hypopnea index (and a higher AHI index), or is it better to achieve a lower overall AHI index (even though this means a higher apnea index)?
This may be a simplistic way of looking at it, but my choice would be to have a lower AI, since an apnea is when airflow stops for more than 10 seconds. Since oxygen levels, heart rate, and blood pressure can be affected, you want to minimize them as much as possible. The duration of apneas and degree of oxygen desaturation would be an indication of the severity of this.

On the other hand, HI, as measured on a xpap are a partial reduction of air flow. And while this can disturb your sleep, they are not as "dangerous".

The way to measure if oxygen desaturation is occurring is to use a recording pulse oximeter. If you doctor orders one, you might be able to borrow one from your DME for an overnight test. Or you can buy your own--you can do a search to find various models used by forum members. I have the CMS 50E, from SE Medical Supply.

My own experience is that I do better on cpap, and that both my AI and HI went down, once I found the right pressure.

_________________
Mask: Pico Nasal CPAP Mask with Headgear
Additional Comments: Resmed AirCurve 10 ASV and Humidifier, Oscar for Mac
KatieW

Snorkler
Posts: 62
Joined: Thu Nov 26, 2009 11:58 am

Re: Pressure paradox

Post by Snorkler » Sat Feb 27, 2010 9:02 pm

Thanks KatieW, I will take your suggestions about the oxyimeter.

By posting here I was hoping to find some suggestions of others and you have provided me with a good point of view.

For now I am not sure whether to set my minimum to 7, 8 or 9. Once I figure this out, then I may try using the apap in cpap mode.

User avatar
DreamStalker
Posts: 7509
Joined: Mon Aug 07, 2006 9:58 am
Location: Nowhere & Everywhere At Once

Re: Pressure paradox

Post by DreamStalker » Sun Feb 28, 2010 9:24 am

Snorkler wrote:I am using an auto-pap I have been tracking my AHI data. I have set the auto-pap in a range from 7 to 12. The pressure very rarely goes above 9 and when it does the results are worse. Reviewing the last week few week’s data I noticed that as my pressure increases from a minimum of 7 to 9 that the number of apneas increase (from 0.5 to 1.2), while the number of hypopneas decrease (from 3.3 to 0.4) and the overall AHI decreases (from 3.8 to 1.8 ). I have listed the effects of going from a psi of 7 to 9 below:

PSI: 7
AI: 0.5
HI: 3.3
AHI: 3.8

PSI: 8
AI: 1.0
HI: 3.1
AHI: 4.1

PSI: 9
AI: 1.2
HI: 0.4
AHI: 1.6

As pressure goes past 9 the numbers tend to get a little bit worse, so for me the best results are in a range from about 7 to 9.

My question is whether it is more desirable to pursue a lower overall apnea index which produces a somewhat higher hypopnea index (and a higher AHI index), or is it better to achieve a lower overall AHI index (even though this means a higher apnea index)?

By the way, I feel pretty good after a night’s sleep whether the pressure is at 7, 8 or 9. But I want to know if the more experienced people here have any guidance on this matter and whether this type of result is common.

Thanks,

Jeff
If those are data from 3 individual nights, the data is meaningless. If those are long-term averages from several weeks worth of data, you may be sensitive to central or mixed apnea.

It is generally more desirable to maximize O2. Therefore you need to monitor O2 with an oximeter to determine what apnea/hypopnea ratio is best for you.
President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.

Snorkler
Posts: 62
Joined: Thu Nov 26, 2009 11:58 am

Re: Pressure paradox

Post by Snorkler » Sat Mar 06, 2010 8:47 am

OK, well I changed the pressure on my apap and raised the minimum from 7 to 9. After one week using the apap with the minimum pressure set at 9 my week-long average AHI decreased from an average of around 2.5 down to 1.0. The weekly averages look as follows:

PSI: 9
AI: 0.6
HI: 0.4
AHI: 1.0

So, I am happy to say that I have found a very good pressure. During the past week I had one night with a low AHI of 0.5 while my high was an AHI of 1.8. The 0.5 AHI was the lowest that I have recorded to date.

I have yet to get an oximeter to check O2 levels, but I am feeling great!!!

Thanks for all the help from everyone here

User avatar
KatieW
Posts: 1672
Joined: Sun Sep 27, 2009 11:07 am
Location: southern AZ

Re: Pressure paradox

Post by KatieW » Sat Mar 06, 2010 9:09 am

That's good news. Congratulations.

_________________
Mask: Pico Nasal CPAP Mask with Headgear
Additional Comments: Resmed AirCurve 10 ASV and Humidifier, Oscar for Mac
KatieW

mindy
Posts: 1753
Joined: Sun Aug 12, 2007 9:36 am

Re: Pressure paradox

Post by mindy » Sat Mar 06, 2010 10:35 am

DreamStalker wrote:
If those are data from 3 individual nights, the data is meaningless. If those are long-term averages from several weeks worth of data, you may be sensitive to central or mixed apnea.

It is generally more desirable to maximize O2. Therefore you need to monitor O2 with an oximeter to determine what apnea/hypopnea ratio is best for you.
I agree with Dreamstalker based on my own experience. I used a Respironics M-Series auto and for me, it worked better with a single pressure. When I had a range, it would inevitably bump up against the top end no matter what I did. As a result, I did better with a single pressure. Eventually I went with the Sandman Auto which, like it's predecessor the Puritan Bennett auto, has a different algorithm to detect central apneas. Once I get the settings adjusted properly for me, I could use a pressure range again and my AHI dropped lower than it had ever been on the Respironics.

I also got a recording pulse oximeter and use that to periodically verify that my therapy is going well.

This is just what worked for me ... I'm not in any way saying that I think others should do this. I do think that if results of using a pressure range aren't giving you adequate therapy, you might want to consider trying a single pressure.

Mindy

_________________
Mask: Swift™ FX Bella Nasal Pillow CPAP Mask with Headgears
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pressure 7-11. Padacheek
"Life isn't about waiting for the storm to pass, it's about learning how to dance in the rain."
--- Author unknown