E-15 error code
E-15 error code
HELP HAVE A REMstar c-FLEX REMPlus cpap showing an error code of E-15 can anyone tell me what this is and help me get a good nights sleep.Thank You
Re: E-15 error code
The list I have shows E-15 as:
E-15 ERR_ARM_UNDEFINED_INSTRUCTION_EXCEPTION •Defective FLASH chip •Replace Main PCA
I am not sure what a flash chip but the fix of replacing the main circuit board does not sound good. Sorry!
Jerry
E-15 ERR_ARM_UNDEFINED_INSTRUCTION_EXCEPTION •Defective FLASH chip •Replace Main PCA
I am not sure what a flash chip but the fix of replacing the main circuit board does not sound good. Sorry!
Jerry
_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
Additional Comments: 11cm/H2O, Encore Pro 1.8i, Pro Analyzer, Encore Viewer1.0 - 3 Remstar Pro2's, 1 Remstar Auto |
Life's journey is not to arrive at the grave safely in a well preserved body, but rather to skid in sideways, totally worn out, shouting: "Wow what a ride!"
I still play Cowboys and Bad Guys but now I use real bullets. CAS
I still play Cowboys and Bad Guys but now I use real bullets. CAS
Re: E-15 error code
I think that is the code I had on mine. It was under warrenty so they fixed it free. It appears that they changed much of the insides. The DME said that often this problem is caused by moving it with water in it. If your is out of warrenty or your insurance won't replace you might be better off to get a new one on line then get it fixed I would think it would be cheaper.
Take it into your DME when you can and they can tell you in about five minutes if they have to send it in or not.
Jack
Take it into your DME when you can and they can tell you in about five minutes if they have to send it in or not.
Jack
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
Re: E-15 error code
Actually, the E-15 error code on the REMstar Plus w/C-Flex shows this in the Service Manual:
Error code: E-15
Problem: Rotor locked. The motor could not be restarted.
Corrective Action: Check wiring. Replace blower assembly. If error persists, replace main PCA
Den
Error code: E-15
Problem: Rotor locked. The motor could not be restarted.
Corrective Action: Check wiring. Replace blower assembly. If error persists, replace main PCA
Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.