ResMed Shutting Off and not turn back on when power goes out
ResMed Shutting Off and not turn back on when power goes out
It seems silly that Resmed or any cpap/bipap has a soft button. My oxygen has a hard on/off button so if there is a flicker in the power (it's on a surge protector) it doesn't shut off and stay off. Anyways living out here in hillybilly hell, the summers we have a lot of power issues. We rarely lose power but the lights will flicker or be off for 30 sec then come back on. This is really bad when it happens at night though with my bipap since it has a soft power button. So when lose that power for a sec or two the unit shuts off but doesn't come back on when the power does.
To me this seems like a really bad design flaw for medical equipment. I am ordering new batteries for my old battery backup to put my unit on that, to hopefully resolve waking up with my machine off and feeling like I cannot breathe! Last night I went 2 1/2hrs with my mask on but without my bipap on and I felt it all day today.
Is there any reason these expensive machines don't have the ability to restart after losing power? Is there a setting on the machine to restart after power and nobody set it to do so? Seems like machines that do so much for our health should stay on all night, right?
Sorry for my vent. I am a little frustrated with this machines faults...lol. Probably because I was doing so well with it but then with the power going out, not getting a good nights sleep I am a bit cranky and worried it's going to happen again tonight
To me this seems like a really bad design flaw for medical equipment. I am ordering new batteries for my old battery backup to put my unit on that, to hopefully resolve waking up with my machine off and feeling like I cannot breathe! Last night I went 2 1/2hrs with my mask on but without my bipap on and I felt it all day today.
Is there any reason these expensive machines don't have the ability to restart after losing power? Is there a setting on the machine to restart after power and nobody set it to do so? Seems like machines that do so much for our health should stay on all night, right?
Sorry for my vent. I am a little frustrated with this machines faults...lol. Probably because I was doing so well with it but then with the power going out, not getting a good nights sleep I am a bit cranky and worried it's going to happen again tonight
_________________
Mask: Amara View Full Face CPAP Mask with Headgear |
Additional Comments: SleepyHead Software and CMS50F pulse Ox |
Since my machine doesn't appear to be listed in the selection:
Resmed Aircurve 10 ASV w/ humidifier and climate line tubing w/ 2L O2. Just restarted therapy after stopping bipap for 1 yr due to complications w/ increased centrals
Resmed Aircurve 10 ASV w/ humidifier and climate line tubing w/ 2L O2. Just restarted therapy after stopping bipap for 1 yr due to complications w/ increased centrals
Re: ResMed Shutting Off and not turn back on when power goes out
Use a multi socket to plug your VPAP and also plug into it this:
http://www.amazon.com/PowerOut-Power-Fa ... _pr_sims_i
it costs $16 and during a power stoppage (even a momentary one) it will sound a soft alarm sound and put its own light on.
when the power returns its light will go off and then you use your finger to turn the VPAP machine on manually. IMO, you should be able to continue respirating while wearing the mask by using the vent holes in the mask and/or also back into the stopped VPAP if there is no antiaspyxia valve, for hours without suffocating, during power outage. Try it now.
http://www.amazon.com/PowerOut-Power-Fa ... _pr_sims_i
it costs $16 and during a power stoppage (even a momentary one) it will sound a soft alarm sound and put its own light on.
when the power returns its light will go off and then you use your finger to turn the VPAP machine on manually. IMO, you should be able to continue respirating while wearing the mask by using the vent holes in the mask and/or also back into the stopped VPAP if there is no antiaspyxia valve, for hours without suffocating, during power outage. Try it now.
_________________
Mask: Mirage™ SoftGel Nasal CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6 |
Last edited by avi123 on Sun Jul 06, 2014 11:12 pm, edited 1 time in total.
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png
- chunkyfrog
- Posts: 34452
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: ResMed Shutting Off and not turn back on when power goes out
The auto-start feature may depend on instant, full power return to turn back on.
If power interrupts, the voltage may fluctuate, or come on gradually,
and the machine is unable to process as normal. Manual operation is always an option,
Especially when unreliable power is in the picture.
If power interrupts, the voltage may fluctuate, or come on gradually,
and the machine is unable to process as normal. Manual operation is always an option,
Especially when unreliable power is in the picture.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: ResMed Shutting Off and not turn back on when power goes out
I thought that factory default was for these machines to return to whatever operation they were performing before the power went out when power is restored. You should be able to simulate this by simply unplugging from power then plugging back in.
That is to say your machine may be defective. Do you have a DME you could question on this operation?
That is to say your machine may be defective. Do you have a DME you could question on this operation?
Re: ResMed Shutting Off and not turn back on when power goes out
Turn on the smartstart function on your S9 unit if you haven't already done so. Do you know how to get into the clinician's menu?
However, since you have a FFM, it may or may not work. Actually, some older models of the S9 won't let you turn smartstart on if you have the machine set to FFM. Even if the machine does let you turn it on, with an FFM, it may or may not work.
I forget whether it was the S9 or PRS1 machine, but one of them would automatically restart if the power was off for less than 5 minutes or so, but would not restart if the power was off longer than that.
Edit: it was the PRS1 machine that had the 5 minute break. My S9 seems to come back on automatically if I pull the power for a few seconds and then plug it back in. I haven't experimented with longer power off periods.
A small UPS could help your S9 unit work through short power outages. Some older CPAP machines like S8 units might be damaged by the MSW waveform from some UPS units, but the S9 should be fine.
However, since you have a FFM, it may or may not work. Actually, some older models of the S9 won't let you turn smartstart on if you have the machine set to FFM. Even if the machine does let you turn it on, with an FFM, it may or may not work.
I forget whether it was the S9 or PRS1 machine, but one of them would automatically restart if the power was off for less than 5 minutes or so, but would not restart if the power was off longer than that.
Edit: it was the PRS1 machine that had the 5 minute break. My S9 seems to come back on automatically if I pull the power for a few seconds and then plug it back in. I haven't experimented with longer power off periods.
A small UPS could help your S9 unit work through short power outages. Some older CPAP machines like S8 units might be damaged by the MSW waveform from some UPS units, but the S9 should be fine.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Re: ResMed Shutting Off and not turn back on when power goes out
SmartStart is a feature on the S9 machines that when turn on should restart the machine after a power outage and power has returned because the machine turns on when it senses breathing.
When you first put the mask on at night do you have to push the button to start the machine blowing or does it come on with just a couple of breaths?
Smart start sometimes doesn't work so great at the lower pressures though...like 4 to 6 because sometimes not enough air is moving.
When you first put the mask on at night do you have to push the button to start the machine blowing or does it come on with just a couple of breaths?
Smart start sometimes doesn't work so great at the lower pressures though...like 4 to 6 because sometimes not enough air is moving.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: ResMed Shutting Off and not turn back on when power goes out
You might want to put an "uninterruptible power supply" (UPS) between the house power and your machine. If your power problems are usually short, you would not need too much battery capacity to carry you through the fluctuations and short outages. They are the same things computer users use and are readily available at office supply stores and maybe places like WalMart and CostCo, too. They also monitor spikes and brown-outs and smooth out most fluctuations from faulty power supplies.
Just remember to set the UPS to silent alarm, or it will wake you up with an alarm that the power has gone to the battery, defeating the purpose of a sound night's sleep.
Don
Just remember to set the UPS to silent alarm, or it will wake you up with an alarm that the power has gone to the battery, defeating the purpose of a sound night's sleep.
Don
_________________
Mask: Simplus Full Face CPAP Mask with Headgear |
Additional Comments: I only look at little screen on machine, let doc analyse details. Now machine set up for me. |
Using Tapatalk
Re: ResMed Shutting Off and not turn back on when power goes out
That is the feature I am talkin about.archangle wrote:My S9 seems to come back on automatically if I pull the power for a few seconds and then plug it back in. I haven't experimented with longer power off periods.
- chunkyfrog
- Posts: 34452
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: ResMed Shutting Off and not turn back on when power goes out
Sometimes Smart Start will have trouble with the smallest size of nasal pillows,
But this has no connection with power outages--I think.
But this has no connection with power outages--I think.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: ResMed Shutting Off and not turn back on when power goes out
chunkyfrog wrote:Sometimes Smart Start will have trouble with the smallest size of nasal pillows,
But this has no connection with power outages--I think.
Anyone know how it works with oral masks? I am considering switching to the Oracle 452 mask since I have had 4 sinus surgeries. I have had a lot of sinus pain and even bruising (around the eyes) from the pap therapy. I can only assume it is from all the reconstructive/sinus windows and stuff not doing well with the pressures. It isn't from the mask being too tight or wrong size. Knowing my luck it probably won't work well with that mask either...lol.
_________________
Mask: Amara View Full Face CPAP Mask with Headgear |
Additional Comments: SleepyHead Software and CMS50F pulse Ox |
Since my machine doesn't appear to be listed in the selection:
Resmed Aircurve 10 ASV w/ humidifier and climate line tubing w/ 2L O2. Just restarted therapy after stopping bipap for 1 yr due to complications w/ increased centrals
Resmed Aircurve 10 ASV w/ humidifier and climate line tubing w/ 2L O2. Just restarted therapy after stopping bipap for 1 yr due to complications w/ increased centrals
Re: ResMed Shutting Off and not turn back on when power goes out
Sorry that's what I meant, not backup but battery for UPS. I have a couple really big ones my husband got years ago but they need new batteries. I still feel these machines should come back on all by themselves for as much as we pay for them...lol. I am going to give the smartstart a try tonight and see what happens.OboeVet wrote:You might want to put an "uninterruptible power supply" (UPS) between the house power and your machine. If your power problems are usually short, you would not need too much battery capacity to carry you through the fluctuations and short outages. They are the same things computer users use and are readily available at office supply stores and maybe places like WalMart and CostCo, too. They also monitor spikes and brown-outs and smooth out most fluctuations from faulty power supplies.
Just remember to set the UPS to silent alarm, or it will wake you up with an alarm that the power has gone to the battery, defeating the purpose of a sound night's sleep.
Don
Thanks to everyone for all the replies!
_________________
Mask: Amara View Full Face CPAP Mask with Headgear |
Additional Comments: SleepyHead Software and CMS50F pulse Ox |
Since my machine doesn't appear to be listed in the selection:
Resmed Aircurve 10 ASV w/ humidifier and climate line tubing w/ 2L O2. Just restarted therapy after stopping bipap for 1 yr due to complications w/ increased centrals
Resmed Aircurve 10 ASV w/ humidifier and climate line tubing w/ 2L O2. Just restarted therapy after stopping bipap for 1 yr due to complications w/ increased centrals
Re: ResMed Shutting Off and not turn back on when power goes out
I have just tried my S9 Autoset. I turned off the switch on the 6-Outlet Power Strip to which the S9 power supply is connected while the S9 was running. Then I waited about a minute and switched the Power Strip back ON. The S9 started to work as usual and blow air without my need to push the Start button. I then changed the Power Off waiting time from right away to about 5 minutes. It was the same. The S9 always started on its own when the voltage came back. I do NOT use the Smartstart.
About the Oracle mask 452 enter it into to search box above, there are a few posters who use it.
Since when avi is a bug?
About the Oracle mask 452 enter it into to search box above, there are a few posters who use it.
Since when avi is a bug?
_________________
Mask: Mirage™ SoftGel Nasal CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6 |
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png
- chunkyfrog
- Posts: 34452
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: ResMed Shutting Off and not turn back on when power goes out
Avi could be an abbreviated avatar--but a bug?
Something new every day. Doggone it, now I need a snack.
Something new every day. Doggone it, now I need a snack.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: ResMed Shutting Off and not turn back on when power goes out
My smartstart seems to be working fine with my mask. We had a couple power outages last night as we had a bad storm. So all is well there.
Yes, avi short for avatar. Before I got sick I was a macro photographer (easing back into it once again ) and so my avi might be something creepy crawly or something strange looking. Matter of fact I will give you all something new to look at here in a moment
Yes, avi short for avatar. Before I got sick I was a macro photographer (easing back into it once again ) and so my avi might be something creepy crawly or something strange looking. Matter of fact I will give you all something new to look at here in a moment
_________________
Mask: Amara View Full Face CPAP Mask with Headgear |
Additional Comments: SleepyHead Software and CMS50F pulse Ox |
Since my machine doesn't appear to be listed in the selection:
Resmed Aircurve 10 ASV w/ humidifier and climate line tubing w/ 2L O2. Just restarted therapy after stopping bipap for 1 yr due to complications w/ increased centrals
Resmed Aircurve 10 ASV w/ humidifier and climate line tubing w/ 2L O2. Just restarted therapy after stopping bipap for 1 yr due to complications w/ increased centrals