[Note: if you live at low altitude then go to high altitude I believe the effects and contraindications mentioned in the study would be accentuated. Many experiments regarding the cause of OSA are done at high altitude.]
BRIEF SUMMARY
Current Knowledge/Study Rationale: This study indicates that altitude associated central apneas increase exponentially with increasing
altitude affecting the results attained from diagnostic polysomnography
and leading to increasing difficulty with PAP titrations for OSA. ...
Study Impact: The higher frequency of central apnea at altitudes above 6000 ft. indicates that such altitudes may be a .. contraindication
for the routine use of … automatic
technologies that are unable to clearly differentiate between central and
obstructive apneas. The difficulty of treating OSA due to central apneas
occurring on treatment with PAP at altitude is likely to lead to a higher
level of untreated OSA among these patients.
Link: http://www.ncbi.nlm.nih.gov/pmc/articles/PMC3227706/
[Note: if you live at low altitude then go to high altitude I believe the effects and contraindications mentioned in the study would be accentuated. Many experiments regarding the cause of OSA are done at high altitude.]
BRIEF SUMMARY
Current Knowledge/Study Rationale: This study indicates that altitude associated central apneas increase exponentially with increasing
altitude affecting the results attained from diagnostic polysomnography
and leading to increasing difficulty with PAP titrations for OSA. ...
Study Impact: The higher frequency of central apnea at altitudes above 6000 ft. indicates that such altitudes may be a .. contraindication
for the routine use of … automatic
technologies that are unable to clearly differentiate between central and
obstructive apneas. The difficulty of treating OSA due to central apneas
occurring on treatment with PAP at altitude is likely to lead to a higher
level of untreated OSA among these patients.
Link: http://www.ncbi.nlm.nih.gov/pmc/articles/PMC3227706/
May any shills trolls sockpuppets or astroturfers at cpaptalk.com be like chaff before the wind!
Re: Contraindication : AutoPAP at High Altitude
The article you referenced does not say anything about the use of AutoPAP machines at high altitude. It talks about difficulties with in-lab titration caused by centrals increasing with altitude. It also suggests that some patients (in the minority) who can be successfully treated at low altitude with "noninvasive PAP" (not AutoPAP) may need to use an ASV machine at high altitude.
The article suggests to me that if travelling to a high altitude, you should keep an eye on your AHI and your detailed graphs in case you develop centrals, regardless of whether you use AutoPAP or fixed-pressure CPAP.
The article suggests to me that if travelling to a high altitude, you should keep an eye on your AHI and your detailed graphs in case you develop centrals, regardless of whether you use AutoPAP or fixed-pressure CPAP.
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: Pilairo Q Nasal Pillow CPAP Mask with Headgear |
Additional Comments: PR DS-560 System One; former S8 Autoset II user; Sleepyhead (Mac); I sometimes use the Mirage Activa |
Re: Contraindication : AutoPAP at High Altitude
"automatic technologies that are unable to clearly differentiate between central and obstructive apneas"big_dave wrote:The article you referenced does not say anything about the use of AutoPAP machines at high altitude. It talks about difficulties with in-lab titration caused by centrals increasing with altitude. It also suggests that some patients (in the minority) who can be successfully treated at low altitude with "noninvasive PAP" (not AutoPAP) may need to use an ASV machine at high altitude.
The article suggests to me that if travelling to a high altitude, you should keep an eye on your AHI and your detailed graphs in case you develop centrals, regardless of whether you use AutoPAP or fixed-pressure CPAP.
May any shills trolls sockpuppets or astroturfers at cpaptalk.com be like chaff before the wind!
Re: Contraindication : AutoPAP at High Altitude
The article says:
"The higher frequency of central apneas at altitudes above 1800 meters (6000 ft.) indicates that such altitudes may be a relative contraindication for the use of limited channel OSA screening devices and automatic technologies that are unable to clearly differentiate between central and obstructive apneas.
The contraindication is for the use of AutoPAP for titration. I believe that the contraindication would also apply to in-home sleep studies. It does not refer to the use of AutoPAP's by patients who have already been diagnosed.
I'm sorry I missed the quote you cited. It is only a minor part of an article focused on titration.
"The higher frequency of central apneas at altitudes above 1800 meters (6000 ft.) indicates that such altitudes may be a relative contraindication for the use of limited channel OSA screening devices and automatic technologies that are unable to clearly differentiate between central and obstructive apneas.
The contraindication is for the use of AutoPAP for titration. I believe that the contraindication would also apply to in-home sleep studies. It does not refer to the use of AutoPAP's by patients who have already been diagnosed.
I'm sorry I missed the quote you cited. It is only a minor part of an article focused on titration.
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: Pilairo Q Nasal Pillow CPAP Mask with Headgear |
Additional Comments: PR DS-560 System One; former S8 Autoset II user; Sleepyhead (Mac); I sometimes use the Mirage Activa |
Re: Contraindication : AutoPAP at High Altitude
Many AutoPAP users here often post symptoms of CO2 maintenance problems which are strongly associated with PAP induced hypocapnic central apneas. I am concerned for them.
I know that my own threshold for CO2 maintenance issues is strongly affected by stress. A couple of months ago I would see signs of unstable breathing in my data at 13 cm/h2o. The anniversary of an assault and robbery upon my person has dropped this below the 8 cm/h2o which still keeps AHI away now. Stress, I believe, has a great deal to do with the threshold at which PAP will induce unstable breathing and hypocapnic central apnea.
For someone under the stress of job or family emergency traveling to that job or emergency, who lives at low altitude but is traveling to high altitude I would think the danger great.
I simply want to make people aware of the issue of stress, altitude, OSA and PAP treatment. I would rather not be responsible for their harm.
I know that my own threshold for CO2 maintenance issues is strongly affected by stress. A couple of months ago I would see signs of unstable breathing in my data at 13 cm/h2o. The anniversary of an assault and robbery upon my person has dropped this below the 8 cm/h2o which still keeps AHI away now. Stress, I believe, has a great deal to do with the threshold at which PAP will induce unstable breathing and hypocapnic central apnea.
For someone under the stress of job or family emergency traveling to that job or emergency, who lives at low altitude but is traveling to high altitude I would think the danger great.
I simply want to make people aware of the issue of stress, altitude, OSA and PAP treatment. I would rather not be responsible for their harm.
May any shills trolls sockpuppets or astroturfers at cpaptalk.com be like chaff before the wind!
Re: Contraindication : AutoPAP at High Altitude
It would be helpful if you could post a summary of what you mean by "C02 maintenance," and what you look for in your data to avoid this problem. I noticed that you have posted about it several times in the past, but your previous posts are rather lengthy. Links to your previous posts may be helpful.
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: Pilairo Q Nasal Pillow CPAP Mask with Headgear |
Additional Comments: PR DS-560 System One; former S8 Autoset II user; Sleepyhead (Mac); I sometimes use the Mirage Activa |
- zoocrewphoto
- Posts: 3732
- Joined: Mon Apr 30, 2012 10:34 pm
- Location: Seatac, WA
Re: Contraindication : AutoPAP at High Altitude
I live in Seattle, pretty darned close to the water. On a hill that is about an altitude of 300 feet.
I did a 10 day vacation to Moab, Utah, which is at 4,000 feet. I had no problems at all. Great ahi every night.
I think I have only had one night above 5 since I started April 30 of this year. And most nights, it is below 2.
I did a 10 day vacation to Moab, Utah, which is at 4,000 feet. I had no problems at all. Great ahi every night.
I think I have only had one night above 5 since I started April 30 of this year. And most nights, it is below 2.
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Resmed S9 autoset pressure range 11-17 |
Who would have thought it would be this challenging to sleep and breathe at the same time?
Re: Contraindication : AutoPAP at High Altitude
How was your stress level?zoocrewphoto wrote:I live in Seattle, pretty darned close to the water. On a hill that is about an altitude of 300 feet.
I did a 10 day vacation to Moab, Utah, which is at 4,000 feet. I had no problems at all. Great ahi every night.
I think I have only had one night above 5 since I started April 30 of this year. And most nights, it is below 2.
May any shills trolls sockpuppets or astroturfers at cpaptalk.com be like chaff before the wind!
Re: Contraindication : AutoPAP at High Altitude
I would love to.big_dave wrote:It would be helpful if you could post a summary of what you mean by "C02 maintenance," and what you look for in your data to avoid this problem. I noticed that you have posted about it several times in the past, but your previous posts are rather lengthy. Links to your previous posts may be helpful.
But i am not up to that right now.
In a nut shell breathing control issues make you blow off too much carbon dioxide which causes physiological changes which damage cells and also appear to tend you toward blowing off more carbon dioxide. PAP pressure is a breathing control issue due to the constant pressure (the equivalent of making the drive to the muscles stronger or the muscles themselves stronger).
Google Scolar: PubMed Search:
Sleep Apnea CO2 reserve
Central Sleep Apnea
Hypocapnic Central Apnea
Sleep Apnea unstable breathing
Sleep Apnea altitude
for details -
- good hunting!
Todzo
May any shills trolls sockpuppets or astroturfers at cpaptalk.com be like chaff before the wind!
Re: Contraindication : AutoPAP at High Altitude
Todzo
I have read and reread several of your threads and still can't understand a thing you say. I'm beginning to wonder if you are indeed a patient or are you just pulling our chain so to speak. You seem to really enjoy posting a reply to every post that is made...me thinks you just like to see your jumbled up words in print and get a real good giggle at the scam I think you are pulling of the patients here (IMHO). I am really beginning to agree with DeltaDave, he is a professional in the field of sleep apnea.....as opposed to you ??????????????
Cheers and a merry christmas to you.......
Nan
I have read and reread several of your threads and still can't understand a thing you say. I'm beginning to wonder if you are indeed a patient or are you just pulling our chain so to speak. You seem to really enjoy posting a reply to every post that is made...me thinks you just like to see your jumbled up words in print and get a real good giggle at the scam I think you are pulling of the patients here (IMHO). I am really beginning to agree with DeltaDave, he is a professional in the field of sleep apnea.....as opposed to you ??????????????
Cheers and a merry christmas to you.......
Nan
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Started cpap in 2010.. still at it with great results.
- Lizistired
- Posts: 2835
- Joined: Tue Dec 14, 2010 10:47 pm
- Location: Indiana
Re: Contraindication : AutoPAP at High Altitude
I think Todzo has been watching those Buteyko videos.
_________________
Mask: Swift™ LT Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Swift FX sometimes, CMS-50F, Cervical collar sometimes, White noise, Zeo... I'm not well, but I'm better. |
ResScan: http://www.resmed.com/int/assets/html/s ... c=patients
ResScan Tutorial- http://montfordhouse.com/cpap/resscan_tutorial/
Machines Video: http://www.cpaplibrary.com/machine-education
ResScan Tutorial- http://montfordhouse.com/cpap/resscan_tutorial/
Machines Video: http://www.cpaplibrary.com/machine-education
Re: Contraindication : AutoPAP at High Altitude
simply - the blood is not on my hands
May any shills trolls sockpuppets or astroturfers at cpaptalk.com be like chaff before the wind!
Re: Contraindication : AutoPAP at High Altitude
Todzo wrote:simply - the blood is not on my hands
When you say something like you just said, people who might have previously been giving you the benefit of the doubt no longer will. That was so childish and more proof you don't know what you are talking about.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: P: 6/10 |
If only the folks with sawdust for brains were as sweet and obliging and innocent as The Scarecrow! ~a friend~
- zoocrewphoto
- Posts: 3732
- Joined: Mon Apr 30, 2012 10:34 pm
- Location: Seatac, WA
Re: Contraindication : AutoPAP at High Altitude
nanwilson wrote:Todzo
I have read and reread several of your threads and still can't understand a thing you say. I'm beginning to wonder if you are indeed a patient or are you just pulling our chain so to speak. You seem to really enjoy posting a reply to every post that is made...me thinks you just like to see your jumbled up words in print and get a real good giggle at the scam I think you are pulling of the patients here (IMHO). I am really beginning to agree with DeltaDave, he is a professional in the field of sleep apnea.....as opposed to you ??????????????
Cheers and a merry christmas to you.......
Nan
Me too. I have never understood any of the stuff about CO2, unstable breathing, overbreathing, etc. My machine does well, and I feel better, so I don't worry about it.
But I have wondered if Todzo is really a sleep apnea patient, or somebody trying to do marketing research, or just messing with us.
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Resmed S9 autoset pressure range 11-17 |
Who would have thought it would be this challenging to sleep and breathe at the same time?
Re: Contraindication : AutoPAP at High Altitude
The sky is falling! The sky is falling!Todzo wrote:simply - the blood is not on my hands
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: Pilairo Q Nasal Pillow CPAP Mask with Headgear |
Additional Comments: PR DS-560 System One; former S8 Autoset II user; Sleepyhead (Mac); I sometimes use the Mirage Activa |