“AutoPAP, Activa, UMFF and Mouth Taping” found at:
viewtopic.php?f=1&t=37954
There I described my methods of data collection and analysis, using a ResMed S8 AutoSet Vantage, presenting and comparing the performance of a ResMed Activa mask and a ResMed Ultra-Mirage Full Face (UMFF) mask at a variety of machine settings.
Reviewing the above post will provide a context for what I now begin to describe here. This post will describe the same sort of “experiments”, but this time using a Respironics REMstar M-Series Auto machine. Since I am using similar protocols with respect to the Respironics data, it will also be possible to very carefully compare this new information to the previous ResMed data.
Efficacy Statistics
This posting, beginning an anticipated series, will look at the efficacy and pressure patterns of the Respironics REMstar M-Series Auto (with A-Flex and C-Flex, but not using these functions; also no ramp) set in AutoPAP mode, with pressures from 6.0 to 20.0 cmH2O. (For those of you who do not “like” such a “wide” pressure range, please see my previously cited ResMed posting, which demonstrated no statistically significant difference between “wide” and “narrow” AutoPAP pressure ranges. In addition, I intentionally now use the same settings on the Respironics as I did with the ResMed, to facilitate comparisons.) The following table presents the average of nightly values, the range of nightly values, and the standard deviation. Percent time in apnea is calculated nightly from the EncoreViewer report “Therapy Data Summary – All Data” (last page) “Average Time in Apnea Per Day” divided by “Total Blower Time”.

(My one night using mouth taping reflects a triumph of curiosity over wisdom. Based on my ResMed findings, I simply had to determine whether the previously-noted degraded efficacy was or was not machine specific. I intended to spend 5 nights using tape; after reviewing the data from a single night I abandoned this plan. Clearly, all efficacy parameters are markedly worse when tape is used, particularly those related to apnea. No explanation for this phenomenon is available.)
There is virtually no difference between the masks as regards AHI. The Activa is slightly superior for reducing apneas (AI and percent-time-in-apnea), while the UMFF is slightly superior for reducing hypopneas (HI), though none of these differences appear to be statistically significant.
Still, viewing a scatter plot of the efficacy parameters is informative:

On four of the twenty nights (20%), Activa values fall within the arbitrary target box (AHI ≤ 10, AI ≤ 1), while the UMFF never achieves this goal. While the values for each mask overlap, and despite the almost identical AHI averages, visually the “clusters” for each mask appear distinct. This is due to the higher AI values for the UMFF. The near-parallelism of the two linear regression lines emphasize the similarities between the two masks; their separation emphasizes the differences. The temptation to remove the “outlier” values for the Activa had to be strongly resisted. Were the single entry for the Activa WITH mouth taping to be included on this chart (AHI=14.1, AI=8.8), the AI axis would need to be extended.
I would also call attention to results for two of the three “process” indices (parameters which drive the Respironics AutoPAP algorithm but which are not conventional “therapeutic outcome” factors):

Non-Responsive Apneas appear only sporadically in isolated occurrences with both masks, and no association with higher pressure is noted. The likelihood that these reflect true central apneas, and the relevance of the reported occurrences, are highly doubtful. Whether the slightly higher frequency during the night using the Activa WITH mouth taping is of any relevance is also doubtful.
Similarly, the small difference in Flow Limitation is of no clinical nor statistical relevance.
Vibratory Snore, however, is massively higher for the UMFF than for the Activa, and the difference is statistically significant. That this factor should differ so markedly with the selection of mask is a VERY surprising finding. Exactly what the Respironics sensors are detecting and calling snore is unclear; whether the mask itself is vibrating at critical frequencies, or whether the mask is affecting underlying physiology (as mouth-taping arguably does) and “creating” actual snoring, cannot be resolved here.
As mentioned previously, it appears that the occurrence of whatever the machine is reporting as “Vibratory Snore” may be producing “runaway” pressure increases. An example follows:

Pressure Profiles
The Respironics AutoPAP algorithm runs at higher pressures with the UMFF than it does with the Activa. Although these differences may not achieve statistical significance, they are arguably clinically relevant. The 90th centile “Titration pressure” for the UMFF mask is almost 3.5 cmH2O higher than that for the Activa mask, a surprisingly large value. In addition, the maximum pressure encountered with the Activa was 16cm, while the maximum pressure with the UMFF was 20 (although only for one minute on one night). Further, with the UMFF, the maximum pressure reached 18cm or higher on five of the twenty nights (25%), while with the Activa the maximum pressure exceeded 13cm on only one night. In addition, there is evidence of “runaway” pressure increases with the UMFF, which will be further reviewed in the discussion of “Vibratory Snore” below. Whether such high-pressure interventions were “necessary,” and/or whether setting a lower “maximum pressure” when using the UMFF mask might be desirable, are unclear.
As suggested by the pressure statistics presented above, the pressure profiles differ for the two masks. The amount of time spent at each pressure range for each of the masks appears below:

As is demonstrated above, with the Activa the machine spends 90% of its time at pressures of 9 and below. With the UMFF, the machine spends 29% of its time at pressures of 10 and higher.
The pressures at which apneas are reported can be similarly displayed:

While the modal pressure for apneas is at 8 cmH2O for both masks, with the Activa 35% of apneas occur above that pressure, while with the UMFF 56% of apneas occur above that pressure. With the Activa the highest pressure at which apnea occurred was 15cm (2 apneas on one night), while with the UMFF the highest pressure at apnea was 19cm (only once; there were also 2 apneas on one night at 18cm).
The correspondence of these two attributes, the distribution of pressure over time, and the distribution of apnea at various pressures, can also be displayed for each mask:


With both masks, there is a close correspondence of the two curves. The significance or desirability of this situation is unclear: does it indicate that the machine is successfully working toward its intended goal, or not? Are pressures rising to follow the apneas, or are apneas following rising pressures?
This question is of particular interest because the one night use of the Activa WITH mouth taping produces the following graph, with unknown statistical reliability, but with a clearly different pattern than displayed above for each curve, and for the relationship between the two curves:

Leak Information
Leak patterns for the masks were examined by doing with Excel spreadsheets what ResMed machines do within their firmware: the pressure-specific venting rates (as provided in manufacturer data) are subtracted from each leak datum based on the pressure at each time point. Respironics machine literature does not quantify its leak specifications, so the categories I have adopted are based on ResMed specifications. With this adjustment and classification, the results are as follows:

It is seen that the Activa mask, with mouth leaks eliminated by taping, exhibits very few higher-level leaks. Differences between the Activa without tape, and the UMFF, are minimal, and leaks are not excessive. This similarity is somewhat surprising, since mouth leaks are expected with the Activa without tape, while only skin-seal leaks are expected with the UMFF. At no time was any “Large Leak” reported by the Respironics machine. Issues regarding the methodology used in this analysis will be revisited below.
Conclusion
Different performance of the two masks (Activa and UMFF) have been described. Under identical conditions with a Respironics AutoPAP, there are differences in efficacy, pressure patterns and leak characteristics. Overall AHI is virtually identical for both masks, though control of apneas is better with the Activa. The Activa also runs at lower overall pressures than does the UMFF. Leak rates are acceptable for both masks.
Most surprising is the detection of significant occurrence of Vibratory Snore with the UMFF, but virtually none with the Activa. Vibratory Snore, as a driver of the Respironics AutoPAP algorithm, appears to contribute to the higher pressures encountered with the UMFF mask, and possibly contributes to “runaway” pressure increases.
Velbor