Thanks for the interesting info. I was looking at this in the Clinician Manual. The graphs in the Clinician Manual make this very evident. I didn't put "two and two" together though until reading your post. Thanks for this valuable information.papdad wrote:JNK I didn't say that she should reset Ti Max on her machine. I was merely trying to restate what is says in the manual about Ti Max and what it does and how it can affect the machine's operation. A Ti Max setting of 3.0 is, according to the manual, for a person with an average BPM of 10 breaths per minute. BlueBonnet's numbers were above that. I was just trying to shed a little light on her orginally posted question. I tried to copy the information from the manual but the tables in the manual would not copy. The following is what did copy from the manual.
Norm
Setting Ti Max
The majority of patients on bilevel therapy experience mouth leaks, especially
during sleep. During mouth leak there may be poor synchronization between
patient and machine breaths, which reduces both the comfort and effectiveness
of therapy.
The VPAP Auto allows clinicians to specify a maximum IPAP time (ie the time the
device spends delivering the inspiratory pressure during a spontaneous breath).
Thus, if a patient develops mouth leak, the VPAP Auto automatically cycles to
EPAP after reaching the maximum time set by the clinician.
In the following graph, Panel A demonstrates poor synchrony between the
diaphragm EMG (EMGdi—which is indicative of inspiratory effort) and
inspiration, resulting in poor chest wall movement and ineffective therapy.
Panel B demonstrates improved synchrony when Ti Max is set to profile the
patient’s intrinsic inspiratory time, as demonstrated by EMGdi, chest wall
movement, and pressure profile synchronization.
TiControl
Patient Flow
Ti
Min
Ti
Max
Cycle
Window
Operating Information 7
Use the equation or table below to calculate a value for Ti Max. This will generally
ensure that Ti Max is set slightly longer that the patient’s inspiratory time.
Ti Max = (60 ÷ patient’s respiratory rate) ÷ 2
Setting Ti Min
The Ti Min parameter allows the clinician to set a minimum time the patient
spends in IPAP. It allows adjustment of the minimum inspiratory time parameter
ranging from 0.1secs to Ti Max. Ti Min is usually set to its default setting of 0.1
second.
At this point, I don't want to make changes to the machine. I first want to (hopefully) get my insurance company's approval on the machine and calculate how much it will cost me out of pocket. I can't afford to purchase one without going through insurance until May 1st at the earliest. I'm HOPING my DME and/or my ENT will be familiar enough with the device to recommend the best settings for me. At the very least, I need to get through letting my DME sent my ENT a report, then adjust the machine according to my ENT's specification.
Thanks for your help though guys! I can't wait to try some adjustments (and hopefully start feeling better again)!