This figure is a perfect example of your minimum IPAP pressure being set too low. Your machine was set to a minimum ipap of 6 and a maximum ipap of 20. As your data shows, the automatic algorithm lets the pressure go too low and then you get a lot of issues so the machine raises the pressure up to handle the problem. Then the machine allows pressure to drift down again causing problems, then you get a lot of issues before the pressure is high enough to help you. You need to set the minimum IPAP pressure higher - closer to where your needs are.SolisQuaesitor71 wrote:Hi, thanks for all of the advice so far, as i had very little sleep the other night i thought i'd post last nights result before i venture to the hospital in 10 mins. I had more sleep, only waking briefly at 3am before going back to sleep to awake properly at about 6am. I'd value your opinions.
You subsequently mentioned the ramp being raised to 10. Ramp is different from the minimum IPAP pressure. Ramp is a comfort feature that increases pressure from zero to the IPAP minimum over time to help you ease into treatment. I suspect that your tech raised the minimum IPAP pressure to 10, closer to what you need, so you will now ramp up to a MINIMUM IPAP pressure of 10.
Post your data tomorrow so we will know what the new setting is and how you responded. Until the minimum IPAP pressure is set close enough to where it is needed, your therapy will be off. Then you will have turned the corner on your treatment.